[labnetwork] HEPA Filters
John Shott
shott at stanford.edu
Fri Sep 28 12:46:42 EDT 2012
Labnetwork:
We (Stanford Nanofabrication Facility) recently changed our 1985 HEPA
filters for the first time since they were installed. As a part of a
facility upgrade that included some improved humidity control in the
lithography area, we had to also add fire sprinklers above the HEPA
filters. Interestingly, as a part of the bid process, one of the
general contractors commissioned a clean room survey of the existing
condition ... expecting to find that our clean room would be terribly
dirty due to our "antique" HEPA filters. They were disappointed to
learn that our 27-year-old HEPA filters were still sufficiently robust
to pass Class 100 clean room certification standards. As other have
commented, however, we were likely paying higher energy costs to push
air through these filters.
In terms of our pre-filtration ... which is am important part of this
system .... we have the following three stages of pre-filtration.
1. All air coming into the building passed through banks of 2" deep
Flanders MERV8 pleated filters. Those are, effectively, high-end
furnace filters and are rated, I believe, at an overall efficiency of
35% (about 85% for particles in the 3-10 micron range).
2. Each of our AC units, which pre-treats all of the incoming make up
air, has two sets of filters:
a. A 95% efficiency bag filter that is about 15" deep. We used to
change these on a timed basis, but now change them when they have a
differential pressure of 0.75" WC.
b. A 99.97% HEPA filter! These are changed every 2-5 years depending on
pressure differential across them. That insures that the "real" clean
room ceiling HEPA filters don't see too many "big chunks" hitting them
on a regular basis.
Finally, our campus facilities group monitors, pays for, and changes
these three sets of pre-filters.
Let me know if you have any questions,
John
On 9/27/2012 7:03 AM, Iulian Codreanu wrote:
> Good Morning.
>
> Where are the prefilters some of you mentioned located? At the
> discharge of the makeup air units and recirc units, right?
>
> Thanks for the feedback.
>
> Iulian