Vito Logiudice vito.logiudice at uwaterloo.ca
Thu Nov 7 11:17:25 EST 2013


Hi Iulian,

We will soon be installing two parallel 0.003um filters on our particular
bulk N2 installation. All related copper pipes from the aluminum
evaporator are being installed per oxygen service installation practices
(cleaned & degreased and silver-soldered under N2 purge) so we should be
OK from a particle perspective, at least given past experience with a very
similar setup. In the event of any filter clogging issues, the redundant
filters will allow for easy replacement of the affected unit without
interrupting flow.

Best regards,
Vito
-- 
Vito Logiudice  M.A.Sc., P.Eng.
Director of Operations, Quantum NanoFab
University of Waterloo
200 University Avenue West
Waterloo, ON      Canada   N2L 3G1
Tel:  1-519-888-4567  ext. 38703
Email: vlogiudi at uwaterloo.ca
Website: https://qncfab.uwaterloo.ca/






-----Original Message-----
From: Iulian Codreanu <codreanu at udel.edu>
Date: Wednesday, 6 November, 2013 6:08 PM
To: Fab Network <labnetwork at mtl.mit.edu>
Subject: [labnetwork] Particle filters on bulk N2

>Esteemed Colleagues,
>
>What is the micron rating of the filters you use after the vaporizers on
>your bulk N2 system?  Would a 0.003 um filter clog too fast in a system
>with brazed Cu piping and aluminum vaporizers?
>
>Thank you very much.
>
>Iulian
>
>-- 
>iulian Codreanu, Ph.D.
>Director of Operations, UD NanoFab
>University of Delaware
>149 Evans Hall
>Newark, DE 19716
>302-831-2784
>
>
>_______________________________________________
>labnetwork mailing list
>labnetwork at mtl.mit.edu
>https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork





More information about the labnetwork mailing list