Aamer,
  The way I understand your question is that you don't want to run at higher chamber pressures than the increased oxygen flow will produce right? If so, you need to install a larger pump or increase conductance if possible. Since you are running without a throttle valve, the chamber and pump are essentially acting in a manner as if the throttle valve were wide open. In this scenario, a throttle valve would only be useful to increase the chamber pressure.

Steve Paolini
Principal Equipment Engineer
Harvard University Center for Nanoscale Systems
11 Oxford St.
Cambridge, MA 02138
617- 496- 9816
spaolini at cns.fas.harvard.edu
www.cns.fas.harvard.edu

From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu> On Behalf Of Dr. Aamer Mahmood
Sent: Thursday, February 07, 2019 3:10 PM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Chamber pressure during reactive e-beam evaporation

Hello all,
We have been doing reactive oxide depositions with e-beam evaporation of metals and a few sccm oxygen flowing through our system. We do not have a variable position gate valve and can only control the oxygen MFC settings. The corresponding chamber pressure has been less than 2e-4 Torr.
I have recently received a request to deposit reactive metal oxide films at higher oxygen flow rates that'll correspond to higher chamber pressure during deposition.
I am hesitant to increase the deposition chamber pressure due to the increased flow rate because the manufacturer mentioned the aforementioned chamber pressure as an acceptable upper limit for the e-gun.
I'd appreciate any insights/experiences on this matter particularly 1) how does increased oxygen pressure in the chamber affect the system hardware, 2) any safety concerns etc.
Thanks in advance.

-Aamer Mahmood
Qatar Environment and Energy Research Institute
Doha, Qatar

This email message, including any attachments is intended solely for the addressee(s), and may contain information that is confidential or legally protected. Any use, disclosure, copying, printing or distribution of the information contained herein by persons other than the designated addressee is unauthorized and may be unlawful. If you believe that you have received this email in error, please notify the sender immediately and delete this email and any attachments from your system and discard any printout thereof. Qatar Foundation makes no warranty that this email is error- or virus-free .The views, opinions, conclusions, and other information expressed in this email are not necessarily endorsed by Qatar Foundation unless otherwise indicated by an authorized representative. As such, Qatar Foundation does not accept any liability for any emails and attachments that may violate the relevant applicable laws.