I agree with Gheorghe. Lithium can contaminate your system and can also
interfere with other processes if it's a multi-user evaporator. It reacts
with water and with Nitrogen (that's why it's normally used in a glove box
with Ar).
Just be careful.
Thanks,
Nava
Nava Ariel-Sternberg, Ph.D.
Director of CNI Shared Labs
CEPSR/MC 8903
530 west 120th st. NY
NY 10027
Office: 212-8549927
Cell: 201-5627600
From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu> On
Behalf Of Gheorghe Iordache
Sent: Tuesday, September 17, 2019 10:31 AM
To: Morrison, Richard H., Jr <rmorrison at draper.com>; labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Lithium evaporation
Dear Rick,
This material deposition chart is a good source of information:
https://www.lesker.com/newweb/deposition_materials/materialdepositionchart.c
fm?pgid=0
Lithium may contaminate your vacuum system and special precautions shall be
taken for exhaust and/or decontamination. It shall be avoided the contact
with water or humid air as they react violently with Lithium dust (see for
example page. 9-62 of Laser Program Annual Report, Volume 3, 1980 ! J)
Kind regards,
=======================
Gheorghe IORDACHE, PhD Physics, MBA
Director of the Nanofabrication Core Lab
KAUST
King Abdullah University of Science and Technology
Al Khawarizmi Building 1 East, Level 3, Office 3415
Thuwal 23955-6900
Kingdom of Saudi Arabia
Office: +966 (0)12 8080447
Mobile: +966 (0)540381154
Email: <mailto:gheorghe.iordache at kaust.edu.sa>
gheorghe.iordache at kaust.edu.sa
Linkedin: <https://www.linkedin.com/in/gheorghe-iordache-phd-physics-mba/>
https://www.linkedin.com/in/gheorghe-iordache-phd-physics-mba/
Follow the Core Labs: <https://corelabs.kaust.edu.sa/>
corelabs.kaust.edu.sa; <https://twitter.com/kaust_corelabs>
https://twitter.com/kaust_corelabs;
<https://www.instagram.com/kaustcorelabs/>
https://www.instagram.com/kaustcorelabs/;
<https://www.linkedin.com/showcase/12179999>
https://www.linkedin.com/showcase/12179999
Visit the Core Labs: <https://corelabs.kaust.edu.sa/visitor-information>
https://corelabs.kaust.edu.sa/visitor-information
================
Please note that the weekend in Saudi Arabia is Friday and Saturday
========
What you do for yourself vanishes once you're gone.
What you do for others stays for others as your legacy.
From: labnetwork-bounces at mtl.mit.edu <mailto:labnetwork-bounces at mtl.mit.edu>
[mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H.,
Jr
Sent: 17 September 2019 12:33
To: Clark O.D. <odc1n08 at soton.ac.uk <mailto:odc1n08 at soton.ac.uk> >;
labnetwork at mtl.mit.edu <mailto:labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] Lithium evaporation
Thanks Clark excellent information.
Rick
From: Clark O.D. [mailto:odc1n08 at soton.ac.uk]
Sent: Tuesday, September 17, 2019 3:43 AM
To: Morrison, Richard H., Jr <rmorrison at draper.com
<mailto:rmorrison at draper.com> >; labnetwork at mtl.mit.edu
<mailto:labnetwork at mtl.mit.edu>
Subject: RE: Lithium evaporation
Can be done and evaporates fairly reliably. It takes very little power to
melt. For us it helped to keep the walls of the crucible cold to stop it
creeping up the sides and out onto the hearth. Can't remember what crucible
we used but a TiN coating is probably not a bad idea to try and stop Li
interdiffusion. There is various literature which discusses it being an
effective barrier to Li. Pyrolytic Boron Nitride is another good
alternative.
It also evaporates well from a k-cell at ~500C but eventually it will creep
around the end of your crucible and attack heating filaments if you run it
too long. PBN crucibles again.
From: labnetwork-bounces at mtl.mit.edu <mailto:labnetwork-bounces at mtl.mit.edu>
<labnetwork-bounces at mtl.mit.edu <mailto:labnetwork-bounces at mtl.mit.edu> > On
Behalf Of Morrison, Richard H., Jr
Sent: 13 September 2019 16:55
To: labnetwork at mtl.mit.edu <mailto:labnetwork at mtl.mit.edu>
Subject: [labnetwork] Lithium evaporation
Hi Folks,
Has anybody allowed Lithium into an E-Beam evaporator, if so any tips?
Thanks
Rick
Richard H. Morrison
Principal Member of the Technical Staff
Draper
555 Technology Square
Cambridge, MA
02139-3573
Work 617-258-3420
Cell 508-930-3461
www.draper.com <http://www.draper.com>
_____
Notice: This email and any attachments may contain proprietary (Draper
non-public) and/or export-controlled information of Draper. If you are not
the intended recipient of this email, please immediately notify the sender
by replying to this email and immediately destroy all copies of this email.
_____
_____
Notice: This email and any attachments may contain proprietary (Draper
non-public) and/or export-controlled information of Draper. If you are not
the intended recipient of this email, please immediately notify the sender
by replying to this email and immediately destroy all copies of this email.
_____
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20190919/39791058/attachment.html>
More information about the labnetwork
mailing list