Dear Aaron,
I used to work on Mass Spectrometers for VG Analytical, which became
part of Thermo Instruments eventually.They had a low accelerating
voltage of 8KV, but we had some tools with a FIB option installed (Fast
Ion Bombardment), and that operated at up to 16 KV.We had a viewport
above the source area, so as to have a reality check on what was really
going as one poked around inside the vacuum chamber. The viewport had
leaded glass. This might be erring on the side of excessive caution,
but such view port glasses are readily available for purchase in the
market place, so that's what I'd suggest.
On Thu, 2020-05-28 at 13:48 +0000, Robert Vandusen wrote:
> Hi Aaron.
> Back in 2015 we had similar concerns with our Balzers BA 510
> thermal/ebeam evaporator. We had reintroduced thermal capability to
> our system and had modified our internal SS shielding to allow us
> to have a direct line of sight to the thermal melt through a glass
> slide mounted on the shield and through one of the window viewports.
> Prior to this we only used ebeam evaporation and the shielding
> blocked the viewport. As you have found, I was unable
> to find any literature on x-ray exposure concerns outside the
> chamber area.
>
> I contacted our Health and Safety department who helped conduct x-ray
> testing during ebeam evaporations of a few different materials.
> They found no measurable x-rays coming through the viewport
> or elsewhere on the chamber. This satisfied our safety concerns.
> However since the system was now identified as an potential x-ray
> producing source we had to get an Ontario ministry of labor x-ray
> permit and post it at the room doors and on the system.
> It basically requires us to get a new assessment if we make any
> changes or modifications to the system that could result in increased
> exposure.
> So we have piece of mind, but now a lot more questions when users see
> the x-ray signage.
>
> Hope this helps.
> Rob
> Robert Vandusen
>
> Microfabrication Cleanlab Facility Manager
>
> Electronics Department
>
> Carleton University
>
> room: 4184 Mackenzie Building
>
> 613-520-2600 ext 5761
>
> Robert_vandusen at cunet.carleton.ca
>
>
>
>
>
> From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu>
> On Behalf Of Aaron Hryciw
>
> Sent: May 27, 2020 11:21 PM
>
> To: labnetwork at mtl.mit.edu
>
> Subject: [labnetwork] Hazard due to x-ray production in e-beam
> evaporators?
>
>
>
>
>
>
>
> [External Email]
>
>
>
>
>
>
>
> Dear colleagues,
>
>
>
> One of the users in our open-access facility recently expressed
> concern about x-ray production in electron-beam evaporation systems.
> Since the typical acceleration voltage in our e-beam systems is 7–10
> kV, he was concerned that bremsstrahlung and characteristic
> x-rays would be generated during deposition, with a maximum energy
> of 7–10 keV (soft to hard x-rays), and that these x-rays would pose a
> health and safety hazard to an operator standing next to the viewport
> for ~30 minutes.
>
>
>
> A literature search yielded a few reports describing radiation damage
> to sensitive devices from x-rays produced during an e-beam
> metallization step, but I did not find any mention of related health
> and safety considerations. While some x-rays are undoubtedly
> produced, presumably there are reasons why they are not hazardous to
> an operator (e.g., perhaps total x-ray output is very small, majority
> of x-ray spectrum is low energy, x-rays do not penetrate stainless
> steel chamber walls or viewport windows, etc.). I
> would like to answer this user with specific physical arguments as
> to why the hazard is insignificant, however, so any advice you could
> offer to this end would be greatly appreciated. Many thanks.
>
>
>
> Cheers,
>
>
>
> – Aaron
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
> Aaron Hryciw, PhD, PEng
> Fabrication Group Manager
> University of Alberta - nanoFAB
> W1-060 ECERF Building
> 9107 - 116 Street
> Edmonton, Alberta
> Canada T6G 2V4 Ph: 780-940-7938
> www.nanofab.ualberta.ca
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
> This email contains links to content or websites. Always be cautious
> when clicking on external links or attachments. If in doubt, please
> forward suspicious emails to
> phishing at carleton.ca.
>
>
>
>
>
> -----End of Disclaimer-----
>
>
>
>
>
>
> _______________________________________________labnetwork mailing
> listlabnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
--
SDI Fabsurplus LLC
Stephen Howe
Company Owner
email: info at fabsurplus.com
Mobile: +1 830-388-1071
Skype: stephencshowe
WWW.FABSURPLUS.COM
Linkedin: https://www.linkedin.com/company/sdi-fabsurplus
Twitter: https://twitter.com/fabsurplus
Instagram: www.instagram.com/fabsurplus
Facebook: www.facebook.com/fabsurplus
Featured Sales Items:-
Fabsurplus.com Closed-Bid Warehouse Clearance Sale May 2020
Applied Materials Centura 5200 EPI HTF-EPI Deposition, 3 CH ATM 200 mm
Axcelis NV-GSD-200E2 Medium Current Ion Implanter, 180kV, 150 mm
Important Notice
The information contained in this email is confidential. If you are not the intended recipient,you must not disclose or use the information in this email in any way.If you received it in error, please tell us immediately by return email and delete this message.SDI-Fabsurplus, LLC does not guarantee the integrity of any emails or attached files. It is also not responsible for any changes made to them by any other person.
Fraud Alert - The strategically important semiconductor equipment sector is currently under attack by fraudsters.Many false and mis-spelled versions of common websites are being registered for use in"Business E-mail Compromise" schemes. False companies have been set up in all parts of the world, including the USA, to do wire fraud. Avoid e-mail fraud by defining an I.T. security policy and carefully monitoring any bank transfers for false information.Check any e-mails sent to you are legitimate by looking at the e-mail headers, and scan any attachments for viruses before opening them.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20200529/27a77bc6/attachment.html>
More information about the labnetwork
mailing list