[labnetwork] Spot issue in Unaxis-750 PECVD

Vibhor Kumar vk409 at soe.rutgers.edu
Mon Aug 23 09:40:43 EDT 2021


Hi David,
Nothing is changed recently.
We are using ULSI grade CF4.
For leak check, I have evacuated the silane gas line and found no leak. Is there any other method you would like to suggest?
Sure, I will discuss with my superiors and contact Plasma-Therm.

Thanks,
-Vibhor-




________________________________
From: Lishan, David (Plasma-Therm LLC) <david.lishan at plasmatherm.com>
Sent: Monday, August 23, 2021 8:29 AM
To: April, Mark R. <mapril at draper.com>; Paolini, Steven <spaolini at cns.fas.harvard.edu>; Vibhor Kumar <vk409 at soe.rutgers.edu>; labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: RE: Spot issue in Unaxis-750 PECVD


Hi Vibhor,

The imprint of the showerhead hole pattern is often an indication that silane is reacting with residual oxygen somewhere behind the showerhead.  The source of the oxygen could be a leak anywhere from behind the showerhead all the back to the main valves of the gas bottles.  If the CF4 clean gas is a premix of CF4/O2, it is possible that the oxygen is coming from there.  If there is small leak and the system has been left idle overnight, it is generally the first run of the day that shows the most pronounced imprint pattern.  Subsequent runs can show a less pronounced pattern.  Perhaps this is what has been described in the original problem statement.



Has anything been recently changed on the system e.g. silane gas bottle, mass flow controller or gas line?



Suggest contacting Plasma Therm Tech Support if the problem persists.



BR,

David



David Lishan, Ph.D., Sr. Member IEEE

Principal Scientist, Director – Technical Marketing



[cid:image002.png at 01D79801.4E3A45B0]


mobile +1.727.251.9056

David.Lishan at plasmatherm.com<mailto:David.Lishan at plasmatherm.com> | plasmatherm.com<https://nam02.safelinks.protection.outlook.com/?url=http%3A%2F%2Fplasmatherm.com%2F&data=04%7C01%7Cvk409%40soe.rutgers.edu%7Cab24b92e084247df089f08d96639faaa%7Cb92d2b234d35447093ff69aca6632ffe%7C1%7C1%7C637653221471771066%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=ZyT19HNNtR6kCAzQx7eEVule86p%2F5%2FcpsqqPTFrz4JU%3D&reserved=0>

10050 16th St. North | St. Petersburg, FL 33716 USA



From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of April, Mark R.
Sent: Monday, August 23, 2021 6:48 AM
To: Paolini, Steven <spaolini at cns.fas.harvard.edu>; Vibhor Kumar <vk409 at soe.rutgers.edu>; labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Spot issue in Unaxis-750 PECVD



CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe.



Good Morning Vibhor,

I concur with Mr. Paolini.  I had a similar issue on a PECVD tool.  After a thorough shower head clean, including the diffuser plate, the shower head pattern was eliminated.  This routinely gets done each year now, during a major PM.  You may need to perform more due to your usage.

Hope this helps.



Mark R. April
Senior  Equipment Engineer

Microfabrication Laboratory
[color_logo_small4]
555 Technology Square

Cambridge, MA  02139

mapril at draper.com<mailto:mapril at draper.com>

O # 617-258-1613 C # 617-455-1596

www.draper.com<https://nam02.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.draper.com%2F&data=04%7C01%7Cvk409%40soe.rutgers.edu%7Cab24b92e084247df089f08d96639faaa%7Cb92d2b234d35447093ff69aca6632ffe%7C1%7C1%7C637653221471771066%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=JcRSxo2MK3Sj0oWrgg0VeJsuoHX21OZEZxHkL4BERh0%3D&reserved=0>







From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Paolini, Steven
Sent: Friday, August 20, 2021 4:02 PM
To: Vibhor Kumar <vk409 at soe.rutgers.edu<mailto:vk409 at soe.rutgers.edu>>; labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] Spot issue in Unaxis-750 PECVD



Vibhor,

  You are most likely seeing the effect of the showerhead being in need of cleaning, particularly the backside. The Deposition of dielectric films creates a powdery substance just outside of the intended plasma concentrate. This debris, over time, will allow small reactions to happen under the showerhead and show up as spots on your wafer, most likely in the pattern of the holes in the showerhead. I would suggest removing the showerhead and diffuser plate and performing a very thorough clean on all the parts. In my experience, and depending on the usage, this should be performed routinely at an interval that keeps the effect at bay.

  Hope this helps,

    Steve Paolini

   Equipment Dood



Steve Paolini

Principal Equipment Engineer

Harvard University Center for Nanoscale Systems

11 Oxford St.

Cambridge, MA 02138

617- 496- 9816

spaolini at cns.fas.harvard.edu<mailto:spaolini at cns.fas.harvard.edu>

www.cns.fas.harvard.edu<https://nam02.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.cns.fas.harvard.edu%2F&data=04%7C01%7Cvk409%40soe.rutgers.edu%7Cab24b92e084247df089f08d96639faaa%7Cb92d2b234d35447093ff69aca6632ffe%7C1%7C1%7C637653221471781060%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=8TusfKGQs7itySqAbVBk3CxvdM7sOSb5Aj7irbPrKDM%3D&reserved=0>



From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Vibhor Kumar
Sent: Friday, August 20, 2021 1:49 PM
To: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: [labnetwork] Spot issue in Unaxis-750 PECVD



Hello Everyone,



During nitride/oxide deposition with Unaxis-750 PECVD, we are getting a mirror image of shower head on the wafer, i.e., spots. We cleaned the chamber with CF4 for 2 consecutive days, each day 4:30 hrs. The conditions were



CF4 = 100 sccm, Temp = 250 C, chamber pressure = 600 mT, time = 4:30 hrs



We also evacuated the silane gas line, but there is no leakage.



After completing the cleaning, when we conditioned the chamber and deposited nitride, the spots were lightened. But on the second day, the spots look enlarging again.



Can anyone please suggest solution to this issue?



Thanks,

-Vibhor Kumar-

Postdoctoral Associate

Rutgers, The State University of New Jersey.



________________________________

Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email.

________________________________
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210823/25be56e3/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 4637 bytes
Desc: image002.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210823/25be56e3/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.png
Type: image/png
Size: 1560 bytes
Desc: image003.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210823/25be56e3/attachment-0001.png>


More information about the labnetwork mailing list