[labnetwork] CDA and copper piping
John Boyle
johnboyle at fast.net
Thu Dec 8 19:56:28 EST 2022
Copper type K, silver soldered with argon purge by experienced mechanical
piping contractor. This is durable and meets all piping codes. I usually use
a tubing that is cleaned for oxygen service. It is ab it more expensive but
takes some uncertainty out of the installation.
CDA should be dried (-100 F dew point) and filtered. One "trick" to use if
particulates are an issue---after tubing is installed run the line and vent
it to a safe location.while air is flowing tap the tubing with a length of
PVC pipe (2-3 ft.) staring at the compressor and following it down the line.
This will help remove any particulate in the line and is low cost and easy
to do. Install filters after purging out the line.
DO NOT use PVC for compressed air. It can burst or explode. PVC in air is
specifically forbidden by ASME Piping codes. Using PVC is a safety issue and
a liability. It can also present insurance issues with your insurer.
John Boyle
John Boyle Consulting, LLC
4848 Canterbury Drive
Emmaus, PA 18049
484-432-7596 Cell
610-965-3208 Office
From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of
Paolini, Steven
Sent: Thursday, December 08, 2022 11:07 AM
To: Denis, Kevin L. (GSFC-5530); labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] CDA and copper piping
Kevin,
Copper is OK for CDA use but I would recommend having the joints brazed
while a purge gas (argon) is flowing. Traditional soldering leaves flux
residue that could migrate downstream. The new style "Propress" fittings are
quite good and very clean as well.
In any event, do NOT use PVC!!!!! When PVC is overpressurized, it explodes
in shards!
Equipment Dood.
Steve Paolini
Principal Equipment Engineer
Harvard University Center for Nanoscale Systems
11 Oxford St.
Cambridge, MA 02138
617- 496- 9816
spaolini at cns.fas.harvard.edu
www.cns.fas.harvard.edu
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Denis, Kevin
L. (GSFC-5530)
Sent: Wednesday, December 07, 2022 4:46 PM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] CDA and copper piping
Hello All,
Has anyone used copper piping for clean dry air lines. If not what kind of
problems would one run into if going this route? Thanks for any insight
that can be provided.
Regards,
Kevin