[labnetwork] Suss MA6 Question

James Mitchell jtmitch5 at ncsu.edu
Thu Mar 23 08:50:05 EDT 2023


Dave, Do you have a part number for CB?
I think I will try this first.

Jim

On Wed, Mar 22, 2023 at 7:17 PM Dave Terry <dave101260 at gmail.com> wrote:

> I had a similar issue several years ago, and turned out to be the CB on
> the back of the system.
>
> Swapped it out, and all is fine since.
>
>
>
> On Wed, Mar 22, 2023 at 3:48 PM James Mitchell <jtmitch5 at ncsu.edu> wrote:
>
>> Lamp power supply will not turn on after lamp failure.
>> We had a lamp (350 W) pop on our system.
>> Cleaned up and installed a new lamp.
>> When we start the power supply the breaker in the back of the system
>> trips.
>> Specifically, it trips when we press CP. System starts at STAND-BY.
>> Press "ON" and the display shows "READY". Press CP and it displays
>> "Wait..." for a second and then trips.
>> We checked for shorts and all looked good. We swapped to the other 220
>> Volt plug in the back and it trips as well.
>> Any ideas on this? We are not sure if the start box or the supply
>> itself is the issue.
>>
>>
>> *Thank you, Jim*
>>
>> *James Mitchell*
>> *Specialty Trades Technician*
>> *Dept. of Electrical and Computer Engineering*
>> *North Carolina State University Nanofabrication Facility (NNF)*
>> *MRC RM243A **Box 7911*
>> *2410 Campus Shore Dr., Raleigh, NC 27606
>> <https://www.google.com/maps/search/2410+Campus+Shore+Dr.,+Raleigh,+NC+27606?entry=gmail&source=g>*
>> *jtmitch5 at ncsu.edu* <jtmitch5 at ncsu.edu>
>> *Desk: 919-515-5394*
>> *Cell: 919-717-7325*
>>
>>
>> _______________________________________________
>> labnetwork mailing list
>> labnetwork at mtl.mit.edu
>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>>
> --
> Best Regards,
> Dave Terry
> •••••••••••••••••••
> Sent from Gmail Mobile
> 617 784 7942
>


-- 
*Thank you, Jim*

*James Mitchell*
*Specialty Trades Technician*
*Dept. of Electrical and Computer Engineering*
*North Carolina State University Nanofabrication Facility (NNF)*
*MRC RM243A **Box 7911*
*2410 Campus Shore Dr., Raleigh, NC 27606*
*jtmitch5 at ncsu.edu* <jtmitch5 at ncsu.edu>
*Desk: 919-515-5394*
*Cell: 919-717-7325*