[labnetwork] Jupiter Callisto abatement system N2 usage

Rob Breisch rob.breisch at trilliumus.com
Fri Sep 15 16:18:18 EDT 2023


Hi Joseph -

That does seem like a lot of N2.  Are you supposed to include any N2 purge from the pump in those numbers.  For example, an Edwards iQDP80 will typically have a ~ 50SLM purge, so in the case of the 3 CFM purge requirement, it's only an additional 30-40 SLM (1-1.5 CFM).  All that N2 is going to flow through into the Callisto.

Also, if you are abating a process that requires passivation of the abatement column, that does often require extra N2, so perhaps that where the higher levels are coming from.  To be clear, that should process dependent and not always necessary to the best of my knowledge.

My experience with dry abatement using granulate (I believe this how Callisto work) is that there is also an upper limit on N2 as well where too much Nitrogen reduces the residence time of the effluent within the cannister, making the abatement less effective.

There are other options for dry abatement if you are looking for something with lower N2 consumption.

Regards,


Rob Breisch
Vice President, Sales and Marketing
[email logo 3]
     Trillium US
     My Mobile: 801-726-5035
     Trillium Toll Free: 800-453-1340
     Email:  rob.breisch at trilliumus.com<https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Frob.breisch%40trilliumus.com%2F&data=04%7C01%7Cdoug.smith%40nxp.com%7C6a18a23fd52849ab6e2c08d9a2d36a20%7C686ea1d3bc2b4c6fa92cd99c5c301635%7C0%7C0%7C637719851320585298%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=28DKNqx19eklsozNWtmNaSym844LuMy0zetFDNZygeM%3D&reserved=0>
     WebSite: www.trilliumus.com<https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.trilliumus.com%2F&data=04%7C01%7Cdoug.smith%40nxp.com%7C6a18a23fd52849ab6e2c08d9a2d36a20%7C686ea1d3bc2b4c6fa92cd99c5c301635%7C0%7C0%7C637719851320595292%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=X54HcYftCFP6zKZ7vzk45e1k%2FRpAfSNZLp4GfLxSYVU%3D&reserved=0>
     Follow Trillium on LinkedIn<https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.linkedin.com%2Fcompany%2F3263466%3Ftrk%3Dtyah%26trkInfo%3Dtas%253Atrillium%2520us%252Cidx%253A1-2-2&data=04%7C01%7Cdoug.smith%40nxp.com%7C6a18a23fd52849ab6e2c08d9a2d36a20%7C686ea1d3bc2b4c6fa92cd99c5c301635%7C0%7C0%7C637719851320605287%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=AB6ZNk5uz%2Ffa%2Fux%2FFDCbaityo4aQYKar6%2F9rv%2FSc3hQ%3D&reserved=0>

This transmission is intended for the named addressee(s) only and may contain confidential and privileged information.  Reading, copying or disseminating this transmission by anyone other than the named addressee(s) is strictly prohibited.  If you have received this message in error, please notify us immediately by reply e-mail or phone at (503) 682-3837.





From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Joseph Losby
Sent: Friday, September 15, 2023 11:33 AM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Jupiter Callisto abatement system N2 usage

Hi everyone,

According to Critical Systems inc., 18 CFM (7 CFM emergency purge, 3 CFM normal purge, 8 CFM venturi pump) of N2 should be allocated to the Jupiter Callisto abatement system.  That seems like a lot to allocate to one piece of equipment.  Do any of you have any experience with them, and is this N2 usage realistic?  Is there a situation where the normal and emergency purges kick in at the same time?

Cheers,
Joe


Joe Losby, PhD
Manager, qLab Operations



[cid:image001.jpg at 01D9E7DD.582070D0]

Quantum City's qConnect 2023: Connecting quantum creators and users | Join us November 15-17, 2023 in Calgary, Alberta at Telus Convention Centre. Click here for more information.<https://research.ucalgary.ca/quantum-city/quantum-city/quantum-city/q-connect-2023>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20230915/76f7cac4/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.jpg
Type: image/jpeg
Size: 13901 bytes
Desc: image001.jpg
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20230915/76f7cac4/attachment.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 7265 bytes
Desc: image002.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20230915/76f7cac4/attachment.png>


More information about the labnetwork mailing list