[labnetwork] Bubble formation when e-beam evaporating Gold on PMMA
Philipp Altpeter
philipp.altpeter at lmu.de
Tue Jul 22 04:23:27 EDT 2025
Dear all,
In recent months, we installed a new UHV e-beam evaporator and have
since encountered significant issues with our standard lift-off process
using 3 nm Chromium followed by 50 nm Gold. Specifically, we are
observing the formation of large bubbles beneath the PMMA, which
severely damage the PMMA layer (see image below). The rate is decently
low, 0.6 A/s at around 40mA emission current, 10 kV. After around 30 nm
of thickness, bubbles become clearly visible.
Interestingly, deposition on bare silicon, silicon dioxide, or other
(photo)resists appears to proceed without problems.
We have already tried adjusting various parameters — including cooling
conditions, beam wobbling, throw distance, and acceleration voltage. We
also modified the PMMA baking protocol and tested PMMA dissolved in
different solvents — all without success.
If anyone has experienced similar issues or has suggestions for
troubleshooting, your input would be greatly appreciated.
Thank you in advance for your help!
Best regards,
Philipp
--
Philipp Altpeter
Fakultät für Physik der LMU
LS Prof. Efetov
Geschwister-Scholl-Platz 1
D-80539 München
T. +49 (0)89 2180-3733
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250722/de2e98b5/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: c8InTNtYBI4xtxk4.png
Type: image/png
Size: 383866 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250722/de2e98b5/attachment.png>
More information about the labnetwork
mailing list