[labnetwork] Bubble formation when e-beam evaporating Gold on PMMA

Rob Belan robb at lesker.com
Tue Jul 22 10:47:48 EDT 2025


Hi Philipp,

For electron beam evaporation, the slower the rate the hotter the substrate will get.  Also, the more power you use, gold usually requires a fair amount of power to evaporate, the hotter the substrate gets.  PMMA will bubble at temps of 80C or higher.  One way to prove this out is to double the rate (which decreases the exposure time in half) and see if that cures the bubble issue.  There are alternate ways to run at lower powers and keep the same rate through the use of spacers under the crucible liners.  But I would for now just see if increasing the rate a bit will keep the PMMA from bubbling.

Let me know if you have further questions.

Rob


Rob Belan
Technical Director PVD Products
Tel: +1 (412) 387-9200 x7378| Cell: +1 (412) 398-1275| Fax: +1 (412) 233-4375
| Email: robb at lesker.com 
If this email is transmitted to you in error, please notify me, and then delete this email and all copies. If this email is sent for legal or business purposes, (a) this email and its content or attachments may include confidential and/or proprietary information of Kurt J. Lesker Company, in which case all rights to that information are reserved by the Company, and (b) your right and authorization to retain, and use, this information is limited to the purpose(s) expressly stated in, or reasonably implied by, this email. Any questions should be directed to me. Thank you.
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Philipp Altpeter
Sent: Tuesday, July 22, 2025 4:23 AM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Bubble formation when e-beam evaporating Gold on PMMA

You don't often get email from philipp.altpeter at lmu.de<mailto:philipp.altpeter at lmu.de>. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>

Dear all,

In recent months, we installed a new UHV e-beam evaporator and have since encountered significant issues with our standard lift-off process using 3 nm Chromium followed by 50 nm Gold. Specifically, we are observing the formation of large bubbles beneath the PMMA, which severely damage the PMMA layer (see image below). The rate is decently low, 0.6 A/s at around 40mA emission current, 10 kV. After around 30 nm of thickness, bubbles become clearly visible.

Interestingly, deposition on bare silicon, silicon dioxide, or other (photo)resists appears to proceed without problems.

We have already tried adjusting various parameters — including cooling conditions, beam wobbling, throw distance, and acceleration voltage. We also modified the PMMA baking protocol and tested PMMA dissolved in different solvents — all without success.

If anyone has experienced similar issues or has suggestions for troubleshooting, your input would be greatly appreciated.

Thank you in advance for your help!

Best regards,
Philipp



[cid:image001.png at 01DBFAF5.FA722E00]

--

Philipp Altpeter

Fakultät für Physik der LMU

LS Prof. Efetov

Geschwister-Scholl-Platz 1

D-80539 München

T. +49 (0)89 2180-3733
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250722/45d031aa/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 383866 bytes
Desc: image001.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250722/45d031aa/attachment.png>


More information about the labnetwork mailing list