[labnetwork] SPM spikes, bath life, (Sulphuric Peroxide Mixture)

Demis D. John demis at ucsb.edu
Thu Mar 6 20:34:18 EST 2025


Instead of Piranha, we have a heated bath of PureStrip.  From the MSDS:

Users heat the bath to 60°C when they want to use it, then cool it down
when they’re done.  This is because the datasheet and our experience finds
that keeping it at 60°C for >>24hrs makes it lose its potency.
Once in a while we have to refill the bath if we find (via SEM!) that PR
residues were not removed as expected.

We used to use Nanostrip instead, which basically never had this problem
and we rarely refilled.  I believe that the mfg stopped providing that and
recommended PureStrip.

-- Demis (contact info <https://wiki.nanotech.ucsb.edu/wiki/Demis_D._John>)
*Reminder*: The NanoFab has a publications policy
<https://wiki.nanotech.ucsb.edu/wiki/Frequently_Asked_Questions#Publications_acknowledging_the_Nanofab>


On Thu, Mar 6, 2025 at 16:59 Hope Fa-Kaji <hfakaji at lbl.gov> wrote:

> Hi Robert,
>
> This is an anecdotal answer based on my experience in industry.
>
> Bath life and spiking limits will depend greatly on the application for
> which you are using SPM. When H2SO4 and H2O2 react to form Caro's acid, the
> other liquid byproduct is H2O. As the H2O concentration of the bath
> increases, all cleans and etches become less effective.
>
> If you are using the SPM for bulk photoresist removal (i.e. your "dirty"
> SPM), I'd think the bath life limitation is driven much more by the
> particulate count than the spiking. Also, I assume in this case you have a
> much higher H2O2:H2SO4 ratio and are consuming H2SO4 much more quickly.
>
> For other applications like targeted etching or simple cleans, you should
> have more room to do more spiking. In particular, if you don't have a
> constant flow of wafers to process, I would consider pre-batch spiking
> instead of a time based interval. This ensures there is enough H2O2 present
> to react and form Caro's acid. Otherwise, the H2O2 could be consumed and/or
> evaporate before the wafers enter the bath.
>
> In my previous role (300mm wafer tools with inline filter/heater/pump and
> lid covers to minimize evaporation), we ran some studies on H2O2 spiking
> and could get 30 batches out of one bath (run at 120C) before the measured
> H2SO4 concentration went below 88%. We had previously run to 15 batches but
> we increased the limit by increasing the batch limit and running numerous
> particle wafer tests to demonstrate no decline in bath quality.
>
> I hope this helps!
>
> On Thu, Mar 6, 2025 at 2:22 PM Macdonald, Robert (GE Aerospace, US) <
> robert.macdonald at geaerospace.com> wrote:
>
>> Wondering who is operating SPM baths (Sulphuric Peroxide Mixture).
>>
>>
>>
>> We heat Sulphuric to 120C (in a quartz bath), and spike with H2O2, at 1%
>> by volume.
>>
>>
>>
>> We are doing 10 spikes, or 1 week, then dispose/aspirate (the cooled)
>> bath, whichever comes first.
>>
>>
>>
>> Does anyone here have opinions or data on the number of spikes they allow?
>>
>>
>>
>>
>> _______________________________________________
>> labnetwork mailing list
>> labnetwork at mtl.mit.edu
>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>>
>
>
> --
> *All the best,*
>
> Hope Fa-Kaji
> Semiconductor Equipment Engineer
> Microsystems Laboratory
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250306/959586cd/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: IMG_5393.jpeg
Type: image/jpeg
Size: 32167 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250306/959586cd/attachment.jpeg>


More information about the labnetwork mailing list