[labnetwork] Contamination control in a UHV co-sputtering system (AJA Orion)
Gerhard Ulbricht
ulbrichtg at cp.dias.ie
Tue Aug 11 11:07:57 EDT 2026
Hello Filippo.
All sputter systems are different but here are my experiences:
I never had any issues with real cross-sputtering (material from a
neighboring, non active target being accidentally deposited). Typical
sputter targets have grounded shields / chimneys around them and a
shutter on top, together with the plasma forming magnets underneath your
active target this reliably prevents non-active targets from sputtering.
There are two risks though: If you have several densely package sputter
guns, you risk depositing some of material A during sputtering onto
inactive gun B. If you then sputter from gun B later you will get
material B + a bit of A. This is pressure dependent but especially at
higher Ar pressures some of your sputtered atoms will make a U-turn. To
prevent this make sure the gap between shutter and chimney of all guns
is as small as possible, and always run your target a bit against the
closed shutter for cleaning.
A similar issue can arise if you have two or more guns active
simultaneously: some material from A can deposit on B and will thus be
co-sputtered by gun B. You will not even notice that during your process
as you want to deposit both materials anyways. But after co-depositing
you should run all targets for some minutes of sputter time in an empty
chamber to make sure their targets are clean again.
Lithographic resist in a UHV sputter system: Depends on your definition
of UHV. If for you UHV means 10^-10 mbar or better, you need to bake out
anything you bring into the chamber, including your resist covered
samples. That will likely destroy / reflow /... the resist anyways. If
you are fine with 10^-8 mbar and therefore do not require strict
bake-outs, most lithographic resists I have used are totally fine to be
sputtered upon (there are many resist though, so there could be
exceptions). I am referring to resists that have seen post-exposure
bake-out before going into the vacuum chamber, reducing the risk of
out-gassing during sputtering.
But if very sensitive samples are run in a sputter system, resist is
usually banned, not necessarily because of the resist itself but because
a resist covered sample by definition is not as clean as possible (it
can't be sonicated / plasma-cleaned / ...) and therefore risks
contaminating the chamber with anything that accidentally sits on the
resist.
The use of Kapton tape in a 10^-8 mbar or better sputter system is not
recommended: The sample surface will likely get warm during sputtering,
risking the glue underneath the Kapton to evaporate and deposit anywhere
in the system, potentially contaminating future samples. I have never
tried Crystalbond myself though.
Hope that helps.
Gerhard.
On 11/08/2026 03:13, #MARTINELLI FILIPPO# wrote:
> Dear labnetwork community,
>
> I have recently been put in charge of a UHV magnetron sputtering
> system (AJA Orion) that I need to maintain and operate. The machine is
> dedicated to co-sputtering of two metals in a nitrogen environment.
>
> The previous operator strongly advised me to be careful about what
> goes into the chamber, as contamination can affect both the chamber
> condition and the properties of the deposited material. However, she
> was not very specific about what exactly causes contamination, and I
> am unsure how restrictive I should be with the materials I load.
>
> I have a few questions I would appreciate your opinion on:
>
> 1. I read that cross-sputtering can occur between nearby substrates,
> or between the substrate and the chamber. In your opinion, are there
> materials that are particularly prone to cross-sputtering and hard to
> remove with standard cleaning routines, and that should therefore be
> avoided in the chamber altogether?
>
> 2. Following on from that, what is your take on lithography resists?
>
> 3. How about mounting materials such as Kapton tape or Crystalbond?
>
> I am quite new to this, so any suggestion or recommendation is
> welcome. I look forward to hearing your opinions.
>
> Cheers,
> Filippo
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20260811/bed2b73b/attachment.html>
More information about the labnetwork
mailing list