      I'd like to add another note of caution. I too have worked with H2
      for a considerable amount of time. Given most lab's volumes I do
      not agree a gas cabinet is "best practice" for 180 cu ft cylinder
      of H2. Having said this I'd like to stay of out that fray (note
      the LEL's for H2/air are ~17%, the LFL's ~ 5%; it doesn't take a
      big room to dilute H2 to below these values.  <br>
      <br>
      Worth adding to this discussion is the routing of H2 lines and the
      danger of H2 mixtures in closed chambers. <br>
      <br>
      H2, when routed through lines near a ceiling can pool H2 in
      dangerous amounts. For a mocvd installation, done some years ago
      in a room with a t-grid ceiling, an exhaust was added at the
      ceiling level to prevent the danger of pooling H2, a lighter gas
      than air. <br>
      <br>
      A mixture of H2 and air or another oxidizer, in a closed chamber
      such as a furnace tube or bell-jar, is a significant hazard. The
      brisance of H2 reactions can result in significant force. Proper
      mixtures of Ar/H2 or N2/H2 eliminate this risk. Ratios on the
      order of 20/1 are common. A few decades ago I routinely (daily)
      used forming gas (N2 80%/H2 20%) in confined spaces (mircowave
      tube processing) with open flames. I never witnessed that ratio
      burn. That was best known method at that time.<br>
      <br>
      Currently, the commonly accepted ratio H2 mixes is 5% in some
      other inert gas.<br>
      <br>
      I agree RFO's are a great line of defense. As per John
      Shott/Stanford we use the DISS standard for all of our dangerous
      gases and incorporated DISS + RFO's early on. <br>
      <br>
      Bob<br>
      <br>
      PS as an added note, the worst lab explosion I've witnessed (after
      the fact) was an iced-over LN Dewar<br>
      <br>
         <br>
      <pre class="moz-signature" cols="60">Robert M. Hamilton
Marvel NanoLab
University of CA at Berkeley
Rm 520 Sutardja Dai Hall
Berkeley, CA 94720-1754
<a class="moz-txt-link-abbreviated" href="mailto:bob@eecs.berkeley.edu">bob@eecs.berkeley.edu</a>
(e-mail preferred)
510-809-8600 
510-325-7557 (mobile - emergencies)
</pre>
      On 7/20/2012 7:34 AM, Weaver, John R wrote:<br>
    </div>
    <blockquote
cite="mid:18AD986E445FE847B2A80B53E65704EDC8B0F01807@VPEXCH02.purdue.lcl"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html;
        charset=ISO-8859-1">
      <meta name="Generator" content="Microsoft Word 12 (filtered
        medium)">
      <style><!--
/* Font Definitions */
@font-face
        {font-family:Wingdings;
        panose-1:5 0 0 0 0 0 0 0 0 0;}
@font-face
        {font-family:Wingdings;
        panose-1:5 0 0 0 0 0 0 0 0 0;}
@font-face
        {font-family:Calibri;
        panose-1:2 15 5 2 2 2 4 3 2 4;}
@font-face
        {font-family:Tahoma;
        panose-1:2 11 6 4 3 5 4 4 2 4;}
/* Style Definitions */
p.MsoNormal, li.MsoNormal, div.MsoNormal
        {margin:0in;
        margin-bottom:.0001pt;
        font-size:12.0pt;
        font-family:"Times New Roman","serif";
        color:black;}
a:link, span.MsoHyperlink
        {mso-style-priority:99;
        color:blue;
        text-decoration:underline;}
a:visited, span.MsoHyperlinkFollowed
        {mso-style-priority:99;
        color:purple;
        text-decoration:underline;}
span.EmailStyle17
        {mso-style-type:personal-reply;
        font-family:"Calibri","sans-serif";
        color:#1F497D;}
.MsoChpDefault
        {mso-style-type:export-only;
        font-size:10.0pt;}
@page WordSection1
        {size:8.5in 11.0in;
        margin:1.0in 1.0in 1.0in 1.0in;}
div.WordSection1
        {page:WordSection1;}
--></style><!--[if gte mso 9]><xml>
<o:shapedefaults v:ext="edit" spidmax="1026" />
</xml><![endif]--><!--[if gte mso 9]><xml>
<o:shapelayout v:ext="edit">
<o:idmap v:ext="edit" data="1" />
</o:shapelayout></xml><![endif]-->
      <div class="WordSection1">
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D">John
            –<o:p></o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D"><o:p> </o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D">My
            addled brain said it backward! Yes, the flow rate allowed by
            the fixed orifice should be higher than the trip point of
            the excess-flow sensor. The key is that you need to allow
            the sensor to trip in a downstream failure that causes
            excess flow while limiting the maximum allowable flow.<o:p></o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D"><o:p> </o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D">My
            apologies to the group for my confusing remark!<o:p></o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D"><o:p> </o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D">John
             <o:p></o:p></span></p>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D"><o:p> </o:p></span></p>
        <div>
          <p class="MsoNormal"><b><i><span
                  style="font-size:10.0pt;color:#1F497D">John R. Weaver<o:p></o:p></span></i></b></p>
          <p class="MsoNormal"><b><i><span
                  style="font-size:10.0pt;color:#1F497D">Facility
                  Manager<o:p></o:p></span></i></b></p>
          <p class="MsoNormal"><b><i><span
                  style="font-size:10.0pt;color:#1F497D">Birck
                  Nanotechnology Center<o:p></o:p></span></i></b></p>
          <p class="MsoNormal"><b><i><span
                  style="font-size:10.0pt;color:#1F497D">Purdue
                  University<o:p></o:p></span></i></b></p>
          <p class="MsoNormal"><b><i><span
                  style="font-size:10.0pt;color:#1F497D">(765) 494-5494<o:p></o:p></span></i></b></p>
          <p class="MsoNormal"><b><i><span
                  style="font-size:10.0pt;color:#1F497D"><a class="moz-txt-link-abbreviated" href="mailto:jrweaver@purdue.edu">jrweaver@purdue.edu</a><o:p></o:p></span></i></b></p>
        </div>
        <p class="MsoNormal"><span
style="font-size:11.0pt;font-family:"Calibri","sans-serif";color:#1F497D"><o:p> </o:p></span></p>
        <div>
          <div style="border:none;border-top:solid #B5C4DF
            1.0pt;padding:3.0pt 0in 0in 0in">
            <p class="MsoNormal" style="margin-left:.5in"><b><span
style="font-size:10.0pt;font-family:"Tahoma","sans-serif";color:windowtext">From:</span></b><span
style="font-size:10.0pt;font-family:"Tahoma","sans-serif";color:windowtext">
                John Shott [<a class="moz-txt-link-freetext" href="mailto:shott@stanford.edu">mailto:shott@stanford.edu</a>] <br>
                <b>Sent:</b> Friday, July 20, 2012 10:23 AM<br>
                <b>To:</b> <a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>; John Weaver<br>
                <b>Subject:</b> Re: [labnetwork] H2 Generators<o:p></o:p></span></p>
          </div>
        </div>
        <p class="MsoNormal" style="margin-left:.5in"><o:p> </o:p></p>
        <p class="MsoNormal" style="margin-left:.5in">John et al:<br>
          <br>
          I am also a big believer in excess-flow switches and RFOs in
          virtually all cylinders. We typically select a 0.010" RFO
          unless there is a good reason to go larger or smaller.  It is
          normally installed by the gas supplier in the cylinder valve
          before it ever shows up at our site ... although we do try to
          confirm that the proper RFO has been installed before we
          install it.  However, I'm a bit confused by the comment about
          setting the excess-flow sensor ABOVE the flow rate allowed by
          the orifice .... maybe I'm guilty of the mistake that you are
          describing.  If the excess flow switch is downstream of the
          RFO (which it is in all cases for us), doesn't that mean that
          the excess flow switch would never trip if it is set higher
          than the RFO-limited flow (at least under steady-flow
          conditions)?  We typically try to size our excess flow switch
          so that it will trip at a flow that is 2-3 times higher than
          the maximum expected flow but that is typically several times
          LOWER than the RFO-limited flow (at least at full cylinder
          pressure).  For most of our gases, we would likely have excess
          flow switches set to trip at 6-10 SLM.  Am I off the mark?<br>
          <br>
          More directly related to the hydrogen discussion, we've
          recently changed our hydrogen sensors from LEL to 0-1000 ppm. 
          While I believe that the legal requirement is to detect a leak
          that reaches 25% LEL, all of our hydrogen sensors are now set
          to give us a warning at 200 ppm and an alarm at 400 ppm.  Our
          thinking is that we are better off catching a leak as early as
          possible, since small leaks seem to have a way of becoming
          large leaks.<br>
          <br>
          My apologies to all for not staying on the topic of H2
          generators ... but whether coming from cylinders, generators,
          or a bulk cryogenic source, safe handling and distribution of
          H2 is important to us all.<br>
          <br>
          Thanks,<br>
          <br>
          John<br>
          <br>
          On 7/20/2012 5:51 AM, Weaver, John R wrote: <o:p></o:p></p>
        <p class="MsoNormal" style="margin-left:.5in"><span
            style="color:#1F497D">I’m also a believer in excess-flow
            sensors and fixed orifices. It seems obvious (but I have
            seen this mistake), but the excess-flow sensor must be set
            above the flow rate allowed by the orifice </span><span
            style="font-family:Wingdings;color:#1F497D">J</span><span
            style="color:#1F497D">.</span><o:p></o:p></p>
        <p class="MsoNormal" style="margin-left:.5in"><o:p> </o:p></p>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
labnetwork mailing list
<a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</pre>
    </blockquote>
    <br>
    <br>
  </body>
</html>