<div>Hi,</div>
<div> </div>
<div>We have a new wet scrubber that has a similar issue.</div>
<div> </div>
<div>When ever we run a process that has HCl as by product the pH reduces drastically(pH< 4 from idle of 8)</div>
<div> </div>
<div>Maintenance</div>
<div>1. We drain and fill water once every week, and set the pH between 9 and 10 by adding few drops of NaOH solution</div>
<div>2. We monitor the pH periodically (morning and afternoon everyday) and when pH<4 , few drops of NaOH(Sodium Hydroxide) is added so that the pH > 9(pH= 10 is the upper limit)</div>
<div>3. Clean pH probe with water when any deposites on it</div>
<div> </div>
<div>Hope the above informations helps.</div>
<div> </div>
<div>Regards,</div>
<div>Siva Prasad Raju Penmetsa</div>
<div>Senior Facility Technologist</div>
<div>Nantional Nano Fabrication Center</div>
<div>Indian Institute of Science(IISc)</div>
<div>Bangalore, India<br><a href="mailto:siva@cense.iisc.ernet.in">siva@cense.iisc.ernet.in</a></div>
<div><br> </div>
<div class="gmail_quote">On Tue, Jul 24, 2012 at 1:29 PM, Michael Khbeis <span dir="ltr"><<a href="mailto:khbeis@uw.edu" target="_blank">khbeis@uw.edu</a>></span> wrote:<br>
<blockquote style="BORDER-LEFT:#ccc 1px solid;MARGIN:0px 0px 0px 0.8ex;PADDING-LEFT:1ex" class="gmail_quote">
<div style="WORD-WRAP:break-word">Dear Colleagues,
<div><br></div>
<div>I was wondering if any of you had experience in pH swings on fume scrubbers. We inherited the facility and associated scrubber that is attached to all our gas cabinets, GRC, and fume hoods. We have little documentation on the PM schedule and no Operating and Maintenance manuals. Basically, we are neutralizing or raising pH to 10, but then over the course of a few minutes up to several hours the system will drop to pH of <5. Acceptable range in our AOP is 5.5 - 11. We see this swing even when tools are idle - there does not appear to be an operational indicator that tracks the system response. </div>
<div><br></div>
<div>System description at a high level: Age of system is 15 years. Last media change - unknown, but plastic media looks mostly normal. Manual fill/drain, system recirculates with addition of caustic soda to balance out acidification - also manual dispense when pH is out of range. We started a bacterial kill that was recommended in the weekly PM schedule about 3 weeks ago and starting seeing this instability in pH a week later. We halted the PM schedule to see what impact there is, if any.</div>
<div><br></div>
<div>Remedies attempted: </div>
<div><span style="WHITE-SPACE:pre-wrap"></span>1) Added sodium bicarbonate to increase buffering of system. Minimal impact.</div>
<div><span style="WHITE-SPACE:pre-wrap"></span>2) Multiple drain/flush cycles - pH was normal (7) for few minutes then dropped to 4 shortly after. </div>
<div><span style="WHITE-SPACE:pre-wrap"></span>3) pH probes were recalibrated and verified to be operational independently.</div>
<div><br></div>
<div>Any recommendations or anecdotal experience would be appreciated. I thank you for participating in this community. </div>
<div><br></div>
<div>Gratefully,</div>
<div><br>
<div><span style="TEXT-TRANSFORM:none;TEXT-INDENT:0px;BORDER-COLLAPSE:separate;FONT:medium Helvetica;WHITE-SPACE:normal;LETTER-SPACING:normal;WORD-SPACING:0px"><span style="TEXT-TRANSFORM:none;TEXT-INDENT:0px;BORDER-COLLAPSE:separate;FONT:medium Helvetica;WHITE-SPACE:normal;LETTER-SPACING:normal;WORD-SPACING:0px">
<div style="WORD-WRAP:break-word"><span style="TEXT-TRANSFORM:none;TEXT-INDENT:0px;BORDER-COLLAPSE:separate;FONT:medium Helvetica;WHITE-SPACE:normal;LETTER-SPACING:normal;WORD-SPACING:0px">
<div style="WORD-WRAP:break-word"><span style="TEXT-TRANSFORM:none;TEXT-INDENT:0px;BORDER-COLLAPSE:separate;FONT:medium Helvetica;WHITE-SPACE:normal;LETTER-SPACING:normal;WORD-SPACING:0px">
<div style="WORD-WRAP:break-word">Dr. Michael Khbeis<br>Associate Director<br>Microfabrication Facility (MFF)<br>University of Washington</div>
<div style="WORD-WRAP:break-word">Fluke Hall, Box 352143<br>(O) 206.543.5101<br>(C) 443.254.5192<br><a href="mailto:khbeis@uw.edu" target="_blank">khbeis@uw.edu</a><br><br><br></div></span></div></span></div></span></span></div>
<br></div></div><br>_______________________________________________<br>labnetwork mailing list<br><a href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br><a href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork" target="_blank">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
<br></blockquote></div><br>