<div dir="ltr">Iulian, <div>We installed 0.01 um filters down stream of our aluminum evaporator. The install was recent so I do not yet have long term data on replacement intervals. If 0.003 um is necessary for your lab, you might consider a larger pore pre-filter upstream. </div>
<div>Items to consider:</div><div>The cost of the filter housing and installation will be much more than the cost of the replacement filters.</div><div>The filters need to be plumbed to enable filter change without interrupting supply to the labs. (we accomplished this with 2 filters in parallel). </div>
<div>There needs to be a method of purging the filters and manifold after filter change. </div><div>Make sure that the filters' CFM rating is equal to or greater than the capacity of the evaporator so you don't limit the supply capacity. </div>
<div>Install pressure gauges both sides of each filter to allow for differential pressure readings. (this will allow you to determine when filter change is necessary). </div><div><br></div><div>Kevin Hensley</div><div><span style="font-size:13px;font-family:arial,sans-serif;border-collapse:collapse"><span style="border-collapse:collapse"><span style="border-collapse:collapse"><span style="border-collapse:collapse">--<font color="#888888"><font color="#888888"><br>
Cleanroom Infrastructure Specialist <br>Utah Nanofab<br><a href="http://www.nanofab.utah.edu/" target="_blank" style="color:rgb(66,99,171)">www.nanofab.utah.edu</a><br>Office: <a href="tel:801%2F587-0658" value="+18015870658" target="_blank">801/587-0658</a><br>
Cell: <a href="tel:801%2F403-6038" value="+18014036038" target="_blank">801/403-6038</a><br>Email: <a href="mailto:khensley@eng.utah.edu" target="_blank" style="color:rgb(66,99,171)">khensley@eng.utah.edu</a><br></font></font></span></span></span></span></div>
<div><span style="font-size:13px;font-family:arial,sans-serif;border-collapse:collapse"><span style="border-collapse:collapse"><span style="border-collapse:collapse"><span style="border-collapse:collapse"><font color="#888888"><br>
</font></span></span></span></span></div></div><div class="gmail_extra"><br><br><div class="gmail_quote">On Wed, Nov 6, 2013 at 4:08 PM, Iulian Codreanu <span dir="ltr"><<a href="mailto:codreanu@udel.edu" target="_blank">codreanu@udel.edu</a>></span> wrote:<br>
<blockquote class="gmail_quote" style="margin:0 0 0 .8ex;border-left:1px #ccc solid;padding-left:1ex">Esteemed Colleagues,<br>
<br>
What is the micron rating of the filters you use after the vaporizers on your bulk N2 system? Would a 0.003 um filter clog too fast in a system with brazed Cu piping and aluminum vaporizers?<br>
<br>
Thank you very much.<br>
<br>
Iulian<br>
<br>
-- <br>
iulian Codreanu, Ph.D.<br>
Director of Operations, UD NanoFab<br>
University of Delaware<br>
149 Evans Hall<br>
Newark, DE 19716<br>
<a href="tel:302-831-2784" value="+13028312784" target="_blank">302-831-2784</a><br>
<br>
<br>
______________________________<u></u>_________________<br>
labnetwork mailing list<br>
<a href="mailto:labnetwork@mtl.mit.edu" target="_blank">labnetwork@mtl.mit.edu</a><br>
<a href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork" target="_blank">https://www-mtl.mit.edu/<u></u>mailman/listinfo.cgi/<u></u>labnetwork</a><br>
</blockquote></div><br></div>