    Also,<br>
    <a class="moz-txt-link-freetext" href="http://www.bay-technologies.net/cryo_pump_and_compressor_repair">http://www.bay-technologies.net/cryo_pump_and_compressor_repair</a><br>
    <br>
    <pre class="moz-signature" cols="72">-- 
Robert Hamilton
University of California at Berkeley
Marvell NanoLab
Equipment Eng. Mgr.
Room 520 Sutardja Dai Hall
Berkeley, CA 94720-1754
<a class="moz-txt-link-abbreviated" href="mailto:bob@eecs.berkeley.edu">bob@eecs.berkeley.edu</a>
Phone: 510-809-8600
Mobile: 510-325-7557
e-mail preferred</pre>
    <br>
    <br>
    <br>
    <div class="moz-cite-prefix">On 2/7/2014 9:09 AM, Olona, Leonard E.
      wrote:<br>
    </div>
    <blockquote cite="mid:309A5424-5446-4EA2-9005-EDDBCC36E72B@ou.edu"
      type="cite">
      <meta http-equiv="Context-Type" content="text/html;
        charset=us-ascii">
      <div>Austin scientific.   <br>
        <br>
        <div>
          <div><b>Leonard E. Olona</b></div>
          <div><i>University Cleanroom Manager</i></div>
          <div><span>University of Oklahoma</span></div>
          <div><span><a moz-do-not-send="true"
                href="x-apple-data-detectors://2">110 West Boyd Street
                Rm</a> 550</span></div>
          <div><a moz-do-not-send="true"
              href="x-apple-data-detectors://3/0">Norman,
              Oklahoma  73019</a></div>
          <div><span>D: <a moz-do-not-send="true"
                href="tel:%2B1-%20405%20325-4374" value="+14053254374"
                target="_blank">+1- 405 325-4374</a></span></div>
          <div><span>C: <a moz-do-not-send="true"
                href="tel:%2B1-%20405%20630-9068" value="+14056309068"
                target="_blank">+1- 405 630-9068</a></span></div>
          <div><span>F: <a moz-do-not-send="true"
                href="tel:%2B1-%20405%20325-7066" value="+14053257066"
                target="_blank">+1- 405 325-7066</a></span></div>
        </div>
      </div>
      <div><br>
        On Feb 7, 2014, at 10:59 AM, "Wang,Xiaojin" <<a
          moz-do-not-send="true"
          href="mailto:xiaojin.wang@louisville.edu">xiaojin.wang@louisville.edu</a>>
        wrote:<br>
        <br>
      </div>
      <blockquote type="cite">
        <div>
          <meta name="Generator" content="Microsoft Word 15 (filtered
            medium)">
          <div class="WordSection1">
            <p class="MsoNormal"><span>Dear all:</span></p>
            <p class="MsoNormal"><span></span></p>
            <p class="MsoNormal"><span>We are looking for a refurbished
                cryopump on e-beam evaporator. Could you recommend me a
                company that can provide good refurbished cryopump with
                a reasonable price ($2000~$3000). Thank you.</span></p>
            <p class="MsoNormal"><span> </span></p>
            <p class="MsoNormal"><span>Best wishes</span></p>
            <p class="MsoNormal"><span>Xiaojin Wang</span></p>
            <p class="MsoNormal"><span>Senior process engineer</span></p>
            <p class="MsoNormal"><span>Micro/nano technology center</span></p>
            <p class="MsoNormal"><span>University of Louisville</span></p>
            <p class="MsoNormal"><span> </span></p>
            <div>
              <div>
                <p class="MsoNormal"><b><span>From:</span></b><span>
                    <a moz-do-not-send="true"
                      href="mailto:labnetwork-bounces@mtl.mit.edu">labnetwork-bounces@mtl.mit.edu</a>
                    [<a moz-do-not-send="true"
                      href="mailto:labnetwork-bounces@mtl.mit.edu">mailto:labnetwork-bounces@mtl.mit.edu</a>]
                    <b>On Behalf Of </b>Vito Logiudice<br>
                    <b>Sent:</b> Wednesday, February 05, 2014 8:44 AM<br>
                    <b>To:</b> Bill Flounders<br>
                    <b>Cc:</b> <a moz-do-not-send="true"
                      href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br>
                    <b>Subject:</b> Re: [labnetwork] Monitoring quality
                    of CDA</span></p>
              </div>
            </div>
            <p class="MsoNormal"> </p>
            <div>
              <div>
                <p class="MsoNormal"><span>Bill,</span></p>
              </div>
              <div>
                <p class="MsoNormal"><span> </span></p>
              </div>
              <div>
                <p class="MsoNormal"><span>Thank you much for this.
                    Might you be able to share some info as to the make
                    and model of the dew point sensor in use at your
                    facility? I'd like to stick with a tried and true
                    unit if at all possible. Thank you.</span></p>
              </div>
              <div>
                <p class="MsoNormal"><span> </span></p>
              </div>
              <div>
                <p class="MsoNormal"><span>Best regards,</span></p>
              </div>
              <div>
                <p class="MsoNormal"><span>Vito</span></p>
              </div>
            </div>
            <div>
              <p class="MsoNormal"><span> </span></p>
            </div>
            <div>
              <p class="MsoNormal"><b><span>From:
                  </span></b><span>Bill Flounders <<a
                    moz-do-not-send="true"
                    href="mailto:bill@eecs.berkeley.edu">bill@eecs.berkeley.edu</a>><br>
                  <b>Date: </b>Tuesday, 4 February, 2014 8:19 PM<br>
                  <b>To: </b>Vito Logiudice <<a
                    moz-do-not-send="true"
                    href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>>,
                  "<a moz-do-not-send="true"
                    href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>"
                  <<a moz-do-not-send="true"
                    href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>><br>
                  <b>Subject: </b>Re: [labnetwork] Monitoring quality
                  of CDA</span></p>
            </div>
            <div>
              <p class="MsoNormal"><span> </span></p>
            </div>
            <div>
              <div>
                <div>
                  <p class="MsoNormal"><span><br>
                      Install dew point sensor on outlet side of air
                      dryer with 4-20mA output signal<br>
                      Common set points are  -100 F, -75 F and - 40 F<br>
                      corresponding to approx 1ppm, 10ppm and 100ppm H2O
                      content.<br>
                      <br>
                      We deliver at 1-10ppm; system alarms at 100ppm
                      (-40 F)<br>
                      <br>
                      Useful reference: <a moz-do-not-send="true"
                        href="http://www.asge-online.com/pdf/ASGEpg185.pdf">http://www.asge-online.com/pdf/ASGEpg185.pdf</a><br>
                      <br>
                      Bill Flounders<br>
                      UC Berkeley<br>
                      <br>
                      <br>
                      Vito Logiudice wrote:</span></p>
                </div>
                <blockquote>
                  <div>
                    <p class="MsoNormal"><span>Dear Colleagues,</span><span></span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span> </span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span>We are in the process of
                        implementing a standalone system for monitoring
                        cleanroom environmental parameters such as
                        temperature, humidity, differential pressures,
                        etc. An additional variable we would like to
                        monitor is moisture content of our clean, dry
                        air (compressed air) supply.</span><span></span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span> </span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span>I would be grateful for
                        any insights on how to best monitor this
                        variable in a reliable fashion. Thank you.</span><span></span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span> </span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span>Regards,</span><span></span></p>
                  </div>
                  <div>
                    <p class="MsoNormal"><span>Vito Logiudice</span><span></span></p>
                  </div>
                  <div>
                    <div>
                      <p class="MsoNormal"><span>-- </span><span></span></p>
                    </div>
                    <div>
                      <p class="MsoNormal"><span>Vito Logiudice
                           M.A.Sc., P.Eng.</span><span></span></p>
                      <p class="MsoNormal"><span>Director of Operations,
                          Quantum NanoFab</span><span></span></p>
                      <p class="MsoNormal"><span>University of Waterloo</span><span></span></p>
                      <p class="MsoNormal"><span>200 University Avenue
                          West</span><span></span></p>
                      <p class="MsoNormal"><span>Waterloo, ON     
                          Canada   N2L 3G1</span><span></span></p>
                      <p class="MsoNormal"><span>Tel:  1-519-888-4567 
                          ext. 38703</span><span></span></p>
                      <p class="MsoNormal"><span>Email: <a
                            moz-do-not-send="true"
                            href="mailto:vlogiudi@uwaterloo.ca">vlogiudi@uwaterloo.ca</a></span><span></span></p>
                      <p class="MsoNormal"><span>Website: <a
                            moz-do-not-send="true"
                            href="https://qncfab.uwaterloo.ca">
                            https://qncfab.uwaterloo.ca</a>/</span><span></span></p>
                    </div>
                  </div>
                  <p class="MsoNormal"><span><br>
                      <br>
                      <br>
                    </span></p>
                  <pre><span>_______________________________________________</span></pre>
                  <pre><span>labnetwork mailing list</span></pre>
                  <pre><span><a moz-do-not-send="true" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><a moz-do-not-send="true" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a></span></pre>
                </blockquote>
                <p class="MsoNormal"><span> </span></p>
              </div>
            </div>
          </div>
        </div>
      </blockquote>
      <blockquote type="cite">
        <div><span>_______________________________________________</span><br>
          <span>labnetwork mailing list</span><br>
          <span><a moz-do-not-send="true"
              href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a></span><br>
          <span><a moz-do-not-send="true"
              href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a></span><br>
        </div>
      </blockquote>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
labnetwork mailing list
<a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</pre>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
Robert Hamilton
University of California at Berkeley
Marvell NanoLab
Equipment Eng. Mgr.
Room 520 Sutardja Dai Hall
Berkeley, CA 94720-1754
<a class="moz-txt-link-abbreviated" href="mailto:bob@eecs.berkeley.edu">bob@eecs.berkeley.edu</a>
Phone: 510-809-8600
Mobile: 510-325-7557
e-mail preferred