    Tom Britton,<br>
    <br>
    Thank you for your insight. <br>
    <br>
    The Marvell NanoLab is about to commission an HCl feed to an epi
    system. We are planning to add a purifier in series with the pigtail
    from the cylinder to the panel. When one installs such a purifier is
    it normal practice to plumb a shunt around it to use during
    trickle-purging/cylinder changes? Although a new term to me, we have
    made it a practice to reverse-feed N2 through pigtails, to the DISS,
    during cylinder changes, with good benefit.<br>
    <br>
    Also, if one install a reactive HCl purifier does this give an
    opportunity to step-down a level in gas purity for HCl, assuming it
    meets a < 1ppm H2O content?<br>
    <br>
    Bob Hamilton<br>
    <pre class="moz-signature" cols="72">-- 
Robert Hamilton
University of California at Berkeley
Marvell NanoLab
Equipment Eng. Mgr.
Room 520 Sutardja Dai Hall
Berkeley, CA 94720-1754
<a class="moz-txt-link-abbreviated" href="mailto:bob@eecs.berkeley.edu">bob@eecs.berkeley.edu</a>
Phone: 510-809-8600
Mobile: 510-325-7557 (Emergencies only)
e-mail preferred</pre>
    <br>
    <br>
    <br>
    <br>
    <div class="moz-cite-prefix"><br>
      <br>
      On 6/11/2014 9:20 AM, Tom Britton wrote:<br>
    </div>
    <blockquote
cite="mid:9820889A26AAC34EBEB01D62DFCEEB2E0115A5E0@P3PWEX2MB006.ex2.secureserver.net"
      type="cite">
      <meta http-equiv="Context-Type" content="text/html;
        charset=us-ascii">
      <meta name="Generator" content="Microsoft Word 14 (filtered
        medium)">
      <div class="WordSection1">
        <p class="MsoPlainText"><span>Hi John,</span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span>I'm glad that you found the source
            of the leak. Any time a system is compromised it causes huge
            issues with down time and cost of repair. As you are
            probably aware, your gas line is most likely trashed to the
            tool and needs to be replaced to the tool. Don't just
            replace the panel and run the gas again, as corrosion
            spreads once it starts. Your gas cabinet vendor should be
            able to get you a new panel, hopefully quickly, so you're
            not down for an excessive time.</span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span>In reading what you wrote about
            leak checking after install, this shouldn't be necessary.
            The major Fabs don't do this as the downtime and cost would
            be enormous. The initial installation by your mechanical
            contactor includes a helium leak check through the cabinet
            to the tool, and if you're performing a cylinder leak check
            after every bottle change, you should be fine from a leak
            check perspective. Instead focus on keeping the gas dry. It
            is well documented that HCL, and most other corrosives, in
            the gaseous form, can run for years in stainless steel lines
            with no issues, but as soon as there is the slightest trace
            of moisture, game over. It's my guess, that when you send
            this panel back to the gas cabinet supplier, you're probably
            going to hear that the valve corroded through where you
            found the leak.
          </span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span>The major points for using any
            corrosive gas are:</span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText">
          <span><span>1.<span>     
              </span></span></span><span>Helium leak check the
            connection after every cylinder change prior to the
            introduction of process gas. We advise using a "He/Inert
            Gas" blend for your purge gas, as you can use the purge gas
            to helium leak test the connection.</span></p>
        <p class="MsoPlainText">
          <span><span>2.<span>     
              </span></span></span><span>Always use DISS fittings for
            both the process gas and the purge gas. Not only do they
            provide a higher level of leak integrity and purity over a
            CGA, but one is less likely to introduce a lower purity
            purge gas by mistake than if the connection has a standard
            580 CGA on it.</span></p>
        <p class="MsoPlainText">
          <span><span>3.<span>     
              </span></span></span><span>Purge Gas Purifier - We highly
            recommend a purge gas purifier for all mineral acid gases.
            Granted, your cylinder costs $1200, but it's cheap insurance
            and it will last a very long time.</span></p>
        <p class="MsoPlainText">
          <span><span>4.<span>     
              </span></span></span><span>Trickle Purge Valve - check to
            make you have a Trickle Purge Valve in your purge line and
            that it's open during bottle changes. This does a good job
            of keeping atmosphere from getting into the gas line during
            the physical changing of the bottle.</span></p>
        <p class="MsoPlainText">
          <span><span>5.<span>     
              </span></span></span><span>Cycle Purges - 30 is not
            enough. Set your purge cycles at 60 for both pre and post
            purge. The pre-purge is for the safety of the operators, and
            does a better job of scrubbing the inside of the tubing and
            components of the HCl, so any atmospheric moisture that gets
            past the trickle purge doesn't react. The post purge is to
            dry the inside of the gas line of the tube and components
            before the HCl is reintroduced into the system. Any less and
            there is a chance that residual moisture will remain, and
            once the HCl gas hits it, the corrosion will start and it's
            a matter of time before you see the effects.</span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span>I was talking with Troy on this
            and he ran his HCl at ON Semi for over 10 years using the
            above practice, and some of their systems sat idle for quite
            some time between quartz cleaning steps on seldom used
            tools. I also talk with a lot of gas systems managers in my
            travels and this is a fairly common way that the production
            Fabs keep their corrosive systems running and the problems
            at bay.</span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span>Hope this helps. See you on
            Sunday.</span></p>
        <p class="MsoNormal"> </p>
        <p class="MsoNormal"><span>Tom</span></p>
        <p class="MsoNormal"><span> </span></p>
        <p class="MsoNormal"><span>Tom Britton</span><span></span></p>
        <p class="MsoNormal"><span>Director of Sales</span><span></span></p>
        <p class="MsoNormal"><i><span>Critical Systems, Inc.</span></i><span></span></p>
        <p class="MsoNormal"><span>Direct: 208-890-1417</span></p>
        <p class="MsoNormal"><span>Office: 877-572-5515</span></p>
        <p class="MsoNormal"><span><a moz-do-not-send="true"
              href="http://www.criticalsystemsinc.com/"><i><span>www.CriticalSystemsInc.com</span></i></a></span><i><u><span></span></u></i></p>
        <p class="MsoNormal"><i><u><span><span> </span></span></u></i></p>
        <p class="MsoNormal"><span><img id="Picture_x0020_1"
              src="cid:part2.07040506.00040004@berkeley.edu" alt="logo
              for email signature png" height="48" width="245"></span></p>
        <p class="MsoNormal"> </p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText"><span> </span></p>
        <p class="MsoPlainText">-----Original Message-----<br>
          From: <a class="moz-txt-link-abbreviated" href="mailto:labnetwork-bounces@mtl.mit.edu">labnetwork-bounces@mtl.mit.edu</a>
          [<a class="moz-txt-link-freetext" href="mailto:labnetwork-bounces@mtl.mit.edu">mailto:labnetwork-bounces@mtl.mit.edu</a>] On Behalf Of John
          Shott<br>
          Sent: Tuesday, June 10, 2014 2:20 PM<br>
          To: Weaver, John R; <a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br>
          Subject: Re: [labnetwork] Anhydrous HCl after it escapes ...</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">Labnetwork Community:</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">First, I'd like to thank Chung Wing
          Leong, Dennis Grimard, Craig Cheney, Tom Britton and John
          Weaver for taking the time to respond to my query. 
        </p>
        <p class="MsoPlainText">I appreciate your collective expertise
          and input and am taking a number of things you suggested under
          careful consideration as we move forward.</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">In the meantime, we have learned a bit
          more about this leak and I thought that I would share with you
          a bit more about what we have learned:</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">First, the leak seems to have occurred
          at the joint between the valve body and the tube stub ... not,
          as we had originally feared and suspected at one of the VCR
          connection points.  Nor did it occur at one of the orbital
          welds between the tube stub on the valve and a VCR gland. </p>
        <p class="MsoPlainText">To me it is unusual to see a leak at
          that point in a valve ...
        </p>
        <p class="MsoPlainText">particularly one that has not been in
          service very long.</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">Second, this failure occurred downstream
          (on the tool side) of the pneumatically-controlled,
          low-pressure outlet valve on the panel.
        </p>
        <p class="MsoPlainText">Certainly on our panel, and in most
          auto-purge panels of this type, this valve is typically closed
          during bottle changes and similar leak checking activities. 
          To be candid, this has been a bit of a blind spot for me: 
          while I told you that we leak check the panel each time that
          we change a cylinder (which is true), now that I think about
          it more carefully, we don't have a mechanism for leak checking
          anything on the downstream side of this valve on a regular
          basis.</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">In fact what we normally leak check is
          that portion of the panel that is upstream of the
          pneumatically-controlled, low-pressure outlet valve. 
        </p>
        <p class="MsoPlainText">Since this is a manifolded cabinet that
          has two bottles feeding a single output line, there are
          actually a number of potential failure points that we do not
          regularly check that includes the output side of each
          pneumatically-controlled output valve, the manual quarter-turn
          valve in series with and downstream from the
          pneumatically-controlled output valve and the fittings
          associated with the tee in the manifold.  As this cabinet has
          not been service for an extended period of time, those
          downstream fittings have likely been checked only twice:  once
          by the cabinet manufacturer at their facility and once by us
          during the cabinet install.  While our cabinet was helium leak
          checked and certified by a third-party to be less than 10^-9
          SCCM leakage, is there a chance that we missed something
          then?  I'm certainly wondering ...</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">My guess is that I am not alone in
          having thought that we leak check the entire panel and all
          fittings each time we change a gas cylinder ...
        </p>
        <p class="MsoPlainText">whereas there are likely several
          fittings and connection points that are not actually checked
          each time.  I will certainly think about this in a new light
          ...</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">Thanks,</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">John</p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">On 6/9/2014 5:43 AM, Weaver, John R
          wrote:</p>
        <p class="MsoPlainText">> John -</p>
        <p class="MsoPlainText">> I've had experiences (certainly
          plural :)) with both external and internal leaks. For an
          internal leak, where moisture has gained access to the piping,
          I have always replaced the piping components up to the point
          where the leak was blocked - a valve, etc. I use valves with
          Hastelloy wetted surfaces (sometimes called Hastelloy trim),
          which is not impervious to HCl corrosion but certainly holds
          up better.</p>
        <p class="MsoPlainText">> For external leaks, the key is to
          neutralize the surfaces as quickly as possible. In one
          instance we had to replace many components because we didn't
          neutralize the surfaces more quickly. A mild base works well
          for neutralization - something with a pH in the 9 range. Then
          a thorough rinse to remove any salts is necessary.</p>
        <p class="MsoPlainText">> We run all of the gas cabinet
          exhausts for chloride gas cabinets through our plastic exhaust
          system, so we don't have to worry about exhaust. If you use a
          metal exhaust on the gas cabinet where the leak occurred, I'd
          carefully inspect the exhaust for a month or two after the
          incident.</p>
        <p class="MsoPlainText">> I hope this helps.</p>
        <p class="MsoPlainText">> John</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> John R. Weaver</p>
        <p class="MsoPlainText">> Facility Manager</p>
        <p class="MsoPlainText">> Birck Nanotechnology Center</p>
        <p class="MsoPlainText">> Purdue University</p>
        <p class="MsoPlainText">> <a moz-do-not-send="true"
            href="mailto:jrweaver@purdue.edu"><span>jrweaver@purdue.edu</span></a></p>
        <p class="MsoPlainText">> nano.purdue.edu</p>
        <p class="MsoPlainText">>
          ________________________________________</p>
        <p class="MsoPlainText">> From: <a moz-do-not-send="true"
            href="mailto:labnetwork-bounces@mtl.mit.edu"><span>labnetwork-bounces@mtl.mit.edu</span></a>
          [<a class="moz-txt-link-abbreviated" href="mailto:labnetwork-bounces@mtl.mit.edu">labnetwork-bounces@mtl.mit.edu</a>]
        </p>
        <p class="MsoPlainText">> on behalf of John Shott
          [<a class="moz-txt-link-abbreviated" href="mailto:shott@stanford.edu">shott@stanford.edu</a>]</p>
        <p class="MsoPlainText">> Sent: Thursday, June 05, 2014 3:08
          PM</p>
        <p class="MsoPlainText">> To: <a moz-do-not-send="true"
            href="mailto:labnetwork@mtl.mit.edu"><span>labnetwork@mtl.mit.edu</span></a></p>
        <p class="MsoPlainText">> Subject: [labnetwork] Anhydrous HCl
          after it escapes ...</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> Labnetwork Community:</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> Sooner or later, it seems,
          anhydrous hydrogen chloride (or a number of
        </p>
        <p class="MsoPlainText">> equally corrosive materials) is
          going to escape.  At least that is our
        </p>
        <p class="MsoPlainText">> experience at the Stanford
          Nanofabrication Facility.  Despite our best
        </p>
        <p class="MsoPlainText">> efforts to use quality components,
          do proper leak checking, etc., it
        </p>
        <p class="MsoPlainText">> seems as if we end up with a leak
          either in a gas cabinet or in a tool
        </p>
        <p class="MsoPlainText">> near a mass flow controller.  At
          that point, anything near the site of
        </p>
        <p class="MsoPlainText">> the original leak has been covered
          with now moisture-laden hydrogen
        </p>
        <p class="MsoPlainText">> chloride ... which, I believe, is
          far more corrosive than the original
        </p>
        <p class="MsoPlainText">> anhydrous material.</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> For those of you who have
          encountered similar situations, how do you
        </p>
        <p class="MsoPlainText">> recover or what to you replace?  Do
          you have effective means of
        </p>
        <p class="MsoPlainText">> neutralizing those metal surfaces? 
          Do your replace VCR gaskets with
        </p>
        <p class="MsoPlainText">> thicker-than-normal or grooved
          "super gaskets"?  Do you replace the
        </p>
        <p class="MsoPlainText">> entire assembly?  Do you leak
          check, put back in service, and pray?</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> Thanks for sharing your experience
          and insights,</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> John</p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">> </p>
        <p class="MsoPlainText">>
          _______________________________________________</p>
        <p class="MsoPlainText">> labnetwork mailing list</p>
        <p class="MsoPlainText">> <a moz-do-not-send="true"
            href="mailto:labnetwork@mtl.mit.edu"><span>labnetwork@mtl.mit.edu</span></a></p>
        <p class="MsoPlainText">> <a moz-do-not-send="true"
            href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">
            <span>https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</span></a></p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText"> </p>
        <p class="MsoPlainText">_______________________________________________</p>
        <p class="MsoPlainText">labnetwork mailing list</p>
        <p class="MsoPlainText"><a moz-do-not-send="true"
            href="mailto:labnetwork@mtl.mit.edu"><span>labnetwork@mtl.mit.edu</span></a></p>
        <p class="MsoPlainText"><a moz-do-not-send="true"
            href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork"><span>https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</span></a></p>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
labnetwork mailing list
<a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</pre>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">

</pre>
  </body>
</html>