    Kamal:<br>
    <br>
    Let me start by saying that I've never actually tried to calibrate
    meters of this type.  Why?  Because it is not easy.  Here is a good
    reference article from over 15 years ago that describes the process
    in great detail including the fact that the standard conductivity
    solutions only go down to about 5 uS/cm ... which isn't very close
    to the 0.06 uS/cm you are hoping to measure.  They also talk about
    separating the whole calibration process into the steps of
    calibrating the meter itself (easy), the temperature probe
    (reasonably easy), and the "cell constant" of the probe itself
    (hard).  A number of you will recognize that the author of this
    paper works for a company that makes and sells resistivity probes
    and monitors.  This is not intended to be an endorsement of that, or
    any other, company ... but, I think, indicates that detailed
    calibration of DI resistivity monitoring systems is typically found
    primarily in the companies that make and sell such instrumentation
    rather than by the folks that use such instrumentation.<br>
    <br>
    Their solution for high-precision calibration was to measure UHP
    water over a range of temperatures as a means of determining and/or
    calibrating the cell constant.  If you read this article, however,
    you will conclude that this is not a procedure for the faint of
    heart.<br>
    <br>
    In recirculating DI systems, I believe that it is more common to
    have continuous resistivity monitoring on both the supply side and
    return side of the system.  In our case, we typically see supply and
    return resistivity readings about 17.7 MOhm-cm or higher ... but
    that rarely, if ever, read the theoretically expected 18.2 MOhm-cm. 
    In fact, it is not uncommon to see a return resistivity that is
    slightly higher than the supply-side resistivity ... which would
    seem unlikely.<br>
    <br>
    Then, on an occasional basis ... probably not as frequently as we
    should ... we (well, a third-party analytical laboratory) collect
    samples and have them measured for particle content, bacteria grown,
    total oxidizable carbon, dissolved silica, and a 30-element
    mass-spec analysis for metal levels in the ppt range that is
    commonly used for DI systems.  In short, there are lots of things
    that CAN be wrong with DI water that are not seen by even an
    accurate resistivity measurement.  In other words, as long as our
    resistivity readings are on the order of 17.5 MOhm-cm or above on
    both supply and return lines, I, for one, don't worry about the
    resistivity aspects of our water.  In fact, earlier this week, I was
    comparing these DI analytical test results with another frequent
    contributor to this forum from the Bay Area institution with the
    longest history as a university laboratory in this field.<br>
    <br>
    Finally, when you say "periodic monitoring" do you mean that you
    have a probe in a continuously recirculating loop and you want to
    look at the resistivity of that periodically ... or that you
    occasionally collect a sample of water and are trying to measure
    it's resistivity?  If it is the latter, that can be tricky: when
    exposed to air, DI water absorbs CO2 which forms carbonic acid that
    can cause your resistivity numbers to degrade.<br>
    <br>
    My guess is some of the folks that run newer labs than ours will
    have more details about the way that they monitor the DI water in
    these newer operations.<br>
    <br>
    Let me know if you have any additional questions.<br>
    <br>
    John<br>
    <br>
    <div class="moz-cite-prefix">On 1/30/2015 2:01 AM, Kamal Yadav
      wrote:<br>
    </div>
    <blockquote
cite="mid:CAOfoCyiJ3FdsX3LDbhbkqD0A_ADdsfu234FRDPoCdiJrvqZzew@mail.gmail.com"
      type="cite">
      <div dir="ltr">Dear All,
        <div><br>
        </div>
        <div>What is the best way to calibrate conductivity meters for
          DI water resistivity periodic monitoring.</div>
        <div><br>
        </div>
        <div>Standard known conductivity solutions are available but
          which one is good and stable for this range of measurement.
          [18 MOhm-cm or ~ 0.06 uS/cm]</div>
        <div><br>
        </div>
        <div>Thanks a lot!<br clear="all">
          <div><br>
          </div>
          -- <br>
          <div class="gmail_signature">
            <div dir="ltr">
              <div>Thanks,<br>
              </div>
              <div>Kamal Yadav</div>
              <div>Sr. Process Technologist<br>
              </div>
              <div>IITBNF, EE Department, Annexe,</div>
              <div>IIT Bombay, Powai</div>
              <div>Mumbai 400076</div>
              <div>Internal: 4435</div>
              <div>Cell: 7506144798</div>
              <div>Email: <a moz-do-not-send="true"
                  href="mailto:kamal.yadav@gmail.com" target="_blank">kamal.yadav@gmail.com</a>,
                <a moz-do-not-send="true"
                  href="mailto:kamalyadav@ee.iitb.ac.in" target="_blank">kamalyadav@ee.iitb.ac.in</a></div>
            </div>
          </div>
        </div>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
labnetwork mailing list
<a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</pre>
    </blockquote>
    <br>
  </body>
</html>