    Matt:<br>
    <br>
    I think that this is a very good point of discussion and expect that
    a number of folks will weigh in.  I certainly expect to learn some
    things from many of my esteemed colleagues, but will also share with
    you what I believe to be an appropriate prioritization. Of course,
    my disclaimer is that I am not a registered PE and certainly would
    not claim to be an authority on fire and other relevant code ...
    plus, what I do know about code is based almost exclusively on
    California Fire Code and may not be applicable in other states or
    countries.<br>
    <br>
    I certainly believe that there are likely to be locale-to-locale
    code variations that have an impact plus probably an even greater
    variation in how your AHJ (Authority Having Jurisdiction) interprets
    a given set of code.  Plus, most of the relevant code of which I am
    aware tells you WHEN you need to have gas detection but often
    doesn't tell you either where those detectors need to be located or
    how many of them you need.  Of course, this discussion is all
    heavily influenced by the level of toxicity of the gas we are
    describing.  My assumption is that we are discussing either toxic or
    highly toxic materials where California code, at least, requires gas
    detection.  Finally, at the end of the day, regardless of what code
    requires, I believe that we all want/need to be comfortable that we
    have an appropriate collection of detectors that are able to detect
    leaks in an appropriate fashion so as to protect the people in and
    around our laboratories.<br>
    <br>
    Even though we have a facility that is about 30 years old, we
    completely replaced our aging gas monitoring system in January, 2012
    so that is the relevant date to think of in terms of when we (and
    the officials at Santa Clara County Fire) last considered these
    issues.<br>
    <br>
    I believe that California code requires us to have gas detectors for
    any toxic or highly toxic gas in the gas cabinet and in any location
    where there are non-welded connections.  That typically means that
    we have to have a detector in the gas box/vented enclosure of the
    tool and in the VMB if there is one.  California code, at least,
    only requires that those be set at 1/2 IDLH alarm points.  We,
    however, choose to set all of our detectors ... even in enclosed
    spaces ... to alarm at PEL, rather than 1/2 IDLH levels.  There are
    a few reasons for this:<br>
    <br>
    1. Small leaks tend to become big leaks over time.  I'd personally
    rather deal with a small leak today than a bigger leak tomorrow.<br>
    <br>
    2. In gas cabinets in particular, where exhaust flow is very high,
    even a good sized leak will be diluted by that air flow and may not
    reach 1/2 IDLH.<br>
    <br>
    3. Since exhausted spaces will be at negative pressure relative to
    their surroundings, the presence of gas outside the enclosure (which
    is generally an occupied breathing space) will be drawn into the
    enclosure and be detected there.  If detectors are set to alarm at
    PEL levels, you can often successfully argue that a detector
    monitoring the exhausted tool enclosure is also doing "double duty"
    and detecting that same gas close to, but outside of, the exhausted
    enclosure.<br>
    <br>
    However, we do not rely entirely on detectors in exhausted
    cabinets.  In the clean room, we do have a reasonable number of
    breathing air detectors.<br>
    <br>
    In general, we do not have detectors at the exhaust of pumps after
    the tools.  For the most part, we know that there will be nasty
    stuff in there.  We rely on our breathing air detectors to tell us
    when something (such as a flex line ...) in the pump exhaust system
    has failed.<br>
    <br>
    We do have one detector in the exhaust of an abatement system that
    is actually monitoring ammonia abatement in a GaN system.  Why
    monitor the ammonia abatement when we don't monitor the other
    abatement systems that often have more toxic gases?  Vendor-specific
    requirement ...<br>
    <br>
    To answer your question of prioritization I would include an
    appropriate number of sensors in the exhaust of gas cabinets, the
    exhaust of tools (and in a VMB, if any) plus the appropriate number
    of sensors in breathing air all in the High Priority category.<br>
    <br>
    I would include detectors in exhaust after abatement systems as a
    "Nice to Have" feature.<br>
    <br>
    I believe that having detectors in the pump exhaust of a tool is Low
    Priority and only confirms what you already know: if toxic stuff is
    going into the system, toxic stuff is coming out too.<br>
    <br>
    However, beyond that prioritization, the immediate follow up
    question is: how many detectors do you need.  Particularly for the
    breathing air sensors, the number and spatial density of them is an
    important consideration.<br>
    <br>
    Let me give you the number of sensors that we have for a 10,000 Sq
    Ft (1000 SqM) clean room facility:<br>
    <br>
    We have a total of about 120 gas sensors that directly
    support/monitor this facility and about 25-30 more that support
    other private laboratories.  Of those 120 sensors, approximately 30
    monitor gas cabinets and the 4 empty cabinets that we use for
    in-bunker storage.  Approximately 75 sensors are in the clean room
    itself with about 50 of them monitoring exhausted enclosures
    associated with individual tools and the remaining 25 monitor
    breathing air close to places where that gas might be found. 
    Finally, we have 17 detectors in the sub-fab.  Four of those monitor
    VMBs or the one BCl3 cabinet that lives outside the bunker, 9
    monitor breathing air, and four monitor for low oxygen (we don't
    have LN2 down there, but we do have 2" distribution lines that could
    deplete a lot of air if they ever ruptured.)<br>
    <br>
    Please let me know if you have any questions and I will looking
    forward to reading the responses from a number of other facilities.<br>
    <br>
    Note: we probably have more gases that need detection than some
    facilities.  Our list of detected gases includes the hydrides
    (silane, germane, diborane, arsine, and phosphine), DCS, chlorine,
    hydrogen bromide, boron trichloride, anhydrous hydrogen chloride,
    anhydrous hydogen fluoride, and ozone.  Also, I should add that our
    120 gas sensors probably includes 15-20 hydrogen detectors (we use
    0-1000 ppm rather than LEL hydrogen sensors).<br>
    <br>
    Thanks,<br>
    <br>
    John<br>
    <br>
    <div class="moz-cite-prefix">On 2/11/2015 9:39 AM, Matthieu Nannini,
      Dr. wrote:<br>
    </div>
    <blockquote
      cite="mid:84DF1F10-9D58-4A5B-963E-FF5E755B2C35@mcgill.ca"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html;
        charset=windows-1252">
      Colleagues, first thanks Vito for initiating this discussion. Very
      important points where made which led me to explore the labnetwork
      archives about sub-atmospheric setup and TGMS. Fore those
      interested I will save you the search:
      <div><a moz-do-not-send="true"
href="https://www-mtl.mit.edu/pipermail/labnetwork/2012-August/000541.html">https://www-mtl.mit.edu/pipermail/labnetwork/2012-August/000541.html</a></div>
      <div><a moz-do-not-send="true"
href="https://www-mtl.mit.edu/pipermail/labnetwork/2013-August/001004.html">https://www-mtl.mit.edu/pipermail/labnetwork/2013-August/001004.html</a></div>
      <div><a moz-do-not-send="true"
href="https://www-mtl.mit.edu/pipermail/labnetwork/2014-July/001346.html">https://www-mtl.mit.edu/pipermail/labnetwork/2014-July/001346.html</a></div>
      <div><a moz-do-not-send="true"
href="https://www-mtl.mit.edu/pipermail/labnetwork/2013-August/000998.html">https://www-mtl.mit.edu/pipermail/labnetwork/2013-August/000998.html</a></div>
      <div><br>
      </div>
      <div>Since we are in a gas discussion timing,</div>
      <div>
        <div><br>
        </div>
        <div>If you had to prioritize the following position of the
          sensors, what would you recommend ?</div>
        <div><br>
        </div>
        <div>- exhaust of gas cabinet</div>
        <div>- gas cabinet at the tool</div>
        <div>- VMB if any</div>
        <div>- exhaust of pump after the tool ?</div>
        <div>- exhaust after abatement system ?</div>
        <div>- free space sensors scattered around most sensitive areas:
          where human presence is usually high</div>
        <div><br>
        </div>
        <div>Thanks</div>
        <div><br>
        </div>
        <div>
          <div apple-content-edited="true"><span
              class="Apple-style-span" style="border-collapse: separate;
              border-spacing: 0px;"><span class="Apple-style-span"
                style="border-collapse: separate; color: rgb(0, 0, 0);
                font-family: Helvetica; font-style: normal;
                font-variant: normal; font-weight: normal;
                letter-spacing: normal; line-height: normal; orphans: 2;
                text-indent: 0px; text-transform: none; white-space:
                normal; widows: 2; word-spacing: 0px;
                -webkit-border-horizontal-spacing: 0px;
                -webkit-border-vertical-spacing: 0px;
                -webkit-text-decorations-in-effect: none;
                -webkit-text-size-adjust: auto;
                -webkit-text-stroke-width: 0px; ">
                <div style="word-wrap: break-word; -webkit-nbsp-mode:
                  space; -webkit-line-break: after-white-space; ">
                  <span class="Apple-style-span" style="border-collapse:
                    separate; color: rgb(0, 0, 0); font-family:
                    Helvetica; font-style: normal; font-variant: normal;
                    font-weight: normal; letter-spacing: normal;
                    line-height: normal; orphans: 2; text-indent: 0px;
                    text-transform: none; white-space: normal; widows:
                    2; word-spacing: 0px;
                    -webkit-border-horizontal-spacing: 0px;
                    -webkit-border-vertical-spacing: 0px;
                    -webkit-text-decorations-in-effect: none;
                    -webkit-text-size-adjust: auto;
                    -webkit-text-stroke-width: 0px; ">
                    <div style="word-wrap: break-word;
                      -webkit-nbsp-mode: space; -webkit-line-break:
                      after-white-space; ">
                      -----------------------------------
                      <div>Matthieu Nannini<br>
                        McGill Nanotools Microfab<br>
                        Manager<br>
                        t: 514 398 3310<br>
                        c: 514 758 3311<br>
                        f: 514 398 8434<br>
                        <a moz-do-not-send="true"
                          href="http://mnm.physics.mcgill.ca/">http://mnm.physics.mcgill.ca/</a><br>
                        ------------------------------------</div>
                    </div>
                  </span></div>
              </span></span></div>
          <br>
          <div>
            <div>Le 2015-02-11 à 10:53, Vito Logiudice <<a
                moz-do-not-send="true"
                href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>>
              a écrit :</div>
            <br class="Apple-interchange-newline">
            <blockquote type="cite">
              <div style="word-wrap: break-word; -webkit-nbsp-mode:
                space; -webkit-line-break: after-white-space; font-size:
                15px; font-family: Calibri, sans-serif;">
                <div>
                  <div>Hi Dennis,</div>
                  <div><br>
                  </div>
                  <div>Great insights – thanks very much for sharing.</div>
                  <div><br>
                  </div>
                  <div>Our aim is to avoid cold spots and keep the
                    entire system at 19C to 20C, especially since the
                    DCS line traverses a loading dock between the gas
                    bunker and the fab. The two roll-up dock doors are
                    equipped with heated air curtains but we wanted the
                    added insurance of a heated/insulated line.  </div>
                  <div><br>
                  </div>
                  <div>In our particular case, we've got a single 120
                    foot line between the gas cabinet and the point of
                    use (no VMB's) and we did our best to stay true to
                    the use of large radius bends all along the run. The
                    DCS panel design was kept as simple as possible (no
                    regulator) and the entire cabinet is located in a
                    heated bunker in which temperature trends are
                    monitored.</div>
                  <div><br>
                  </div>
                  <div>Good point about the possible risk of fire. While
                    the heat trace controller is capable of outputting a
                    limited amount of power, we did see some odd "burn"
                    marks at some locations which lead us to conclude
                    that the Armaflex insulation's upper use limit of
                    105C may have been exceeded at some of the void
                    locations. In light of these findings we've decided
                    to use fiberglass insulation instead of Armaflex for
                    the repair.</div>
                  <div><br>
                  </div>
                  <div>Best,</div>
                  <div>Vito</div>
                  <div>
                    <div><br>
                    </div>
                  </div>
                </div>
                <span id="OLK_SRC_BODY_SECTION">
                  <div style="font-family: Calibri; font-size: 11pt;
                    text-align: left; border-width: 1pt medium medium;
                    border-style: solid none none; padding: 3pt 0in 0in;
                    border-top-color: rgb(181, 196, 223);">
                    <span style="font-weight:bold">From: </span>Dennis
                    Grimard <<a moz-do-not-send="true"
                      href="mailto:dgrimard@umich.edu">dgrimard@umich.edu</a>><br>
                    <span style="font-weight:bold">Date: </span>Tuesday,
                    10 February, 2015 10:32 PM<br>
                    <span style="font-weight:bold">To: </span>Vito
                    Logiudice <<a moz-do-not-send="true"
                      href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>><br>
                    <span style="font-weight:bold">Cc: </span>Labnetwork
                    <<a moz-do-not-send="true"
                      href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>><br>
                    <span style="font-weight:bold">Subject: </span>Re:
                    [labnetwork] Conclusion: Heat trace issues on DCS
                    gas lines<br>
                  </div>
                  <div><br>
                  </div>
                  <div>
                    <div dir="auto">
                      <div>Vito:</div>
                      <div><br>
                      </div>
                      <div>I have watched with great pleasure the
                        discussion on this topic.  I too agree that much
                        good info has been discussed ... Great feedback
                        from some very knowledgable people indeed.</div>
                      <div><br>
                      </div>
                      <div>I need to throw a wrench in the discussion
                        (or prove my ignorance).  I have always resisted
                        heat taping for the following reasons:  1) when
                        the tube enters a VMB or any ventilated
                        enclosure there is a significant temperature
                        drop due to the large purging flow rate within
                        the enclosure ... Tending to cool the line at
                        the worst possible point, 2) the VMB type
                        enclosures tend to have many right angle welds
                        and valves which promote condensation ....
                        Rather than long graceful bends typically used
                        external to the enclosure, 3) SS is a horrible
                        heat conductor ... As is n2 gas ... So if I heat
                        trace a double wall tube how much heat actually
                        gets to the inner tube?  how consistent is that
                        heat?  What is the temperature gradient?, and 4)
                        the actual cold to hot temperature gradient
                        (desired) is difficult to institute along the
                        length of line ... A good feedback loop is
                        required.  Also, heat tape gives me the district
                        impression that it can contribute to an out of
                        control heating failure with a possible fire as
                        a result.</div>
                      <div><br>
                      </div>
                      <div>So, not that it solves your problem but here
                        is what I have tried to always implement: 1)
                        short runs (home runs not a distribution), 2)
                        minimum short radius right angles, 3) minimize
                        VMB's ... Mini gas cabinets with multiple
                        outputs in the cabinet, 4) chilled bottles, 5)
                        vacuum delivery, and 6) large radius bends.</div>
                      <div><br>
                      </div>
                      <div>Just food for thought ...<br>
                        <br>
                        Dennis S Grimard, Ph.D.
                        <div>Associate Director of Operations, MIT.nano</div>
                        <div><br>
                        </div>
                        <div>Massachusetts Institute of Technology</div>
                        <div>60 Vassor Street, Bldg 39-556</div>
                        <div>Cambridge, MA 02149</div>
                        <div><br>
                        </div>
                        <div>C:     (734) 368-7172</div>
                        <div>EM:  <a moz-do-not-send="true"
                            href="mailto:dgrimard@mit.edu">dgrimard@mit.edu</a></div>
                      </div>
                      <div><br>
                        On Feb 10, 2015, at 1:30 PM, Vito Logiudice <<a
                          moz-do-not-send="true"
                          href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>>
                        wrote:<br>
                        <br>
                      </div>
                      <blockquote type="cite">
                        <div>
                          <div>
                            <div>Dear Colleagues,</div>
                            <div><br>
                            </div>
                            <div>Thank you very much to everyone whom
                              took the time to write in with their
                              insights on this issue. Special thanks to
                              John Shott and Tom Britton for the photos
                              and reference documents provided.</div>
                            <div><br>
                            </div>
                            <div>So that others may perhaps benefit from
                              our experience, we've concluded that the
                              cause of the premature failure appears to
                              have been the presence of several "voids"
                              where the heat trace was not in intimate
                              contact with the SS tubing. This occurred
                              even though the trace had been taped every
                              12 inches per the manufacturer's
                              recommendations. We also noted voids at
                              some elbows where maintaining contact
                              was/is difficult. </div>
                            <div><br>
                            </div>
                            <div>To keep the issue from repeating itself
                              in the future, our plan is to reinstall
                              two new heat traces along the length of
                              the tubing, one on the bottom and one on
                              the top. One of these will remain active
                              while the backup trace will be kept off
                              and act as an insurance policy should the
                              primary unit fail in the future. If anyone
                              sees a problem with this particular
                              approach, I would be glad to hear from
                              you.</div>
                            <div><br>
                            </div>
                            <div>In the new installation, conductive
                              putty will be used to fill any voids
                              before aluminum tape is applied along the
                              entire length of the line much like John
                              showed in his attached photo. The entire
                              assembly will then be re-insulated per the
                              original design specification.
                              Fortunately, the problem occurred under
                              warranty so our only out-of-pocket cost
                              will be limited to the cost of the backup
                              heat trace (a few hundred dollars).</div>
                            <div><br>
                            </div>
                            <div>Regards,</div>
                            <div>Vito</div>
                          </div>
                          <div><br>
                          </div>
                          <span id="OLK_SRC_BODY_SECTION">
                            <div style="font-family: Calibri; font-size:
                              11pt; text-align: left; border-width: 1pt
                              medium medium; border-style: solid none
                              none; padding: 3pt 0in 0in;
                              border-top-color: rgb(181, 196, 223);">
                              <span style="font-weight:bold">From: </span>Vito
                              Logiudice <<a moz-do-not-send="true"
                                href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>><br>
                              <span style="font-weight:bold">Date: </span>Wednesday,
                              21 January, 2015 12:23 PM<br>
                              <span style="font-weight:bold">To: </span>Labnetwork
                              <<a moz-do-not-send="true"
                                href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>><br>
                              <span style="font-weight:bold">Subject: </span>[labnetwork]
                              Heat trace issues on DCS gas lines<br>
                            </div>
                            <div><br>
                            </div>
                            <div>
                              <div style="word-wrap: break-word;
                                -webkit-nbsp-mode: space;
                                -webkit-line-break: after-white-space;
                                font-size: 15px; font-family: Calibri,
                                sans-serif;">
                                <div>Dear Colleagues,</div>
                                <div><br>
                                </div>
                                <div>We are experiencing an issue with
                                  the heat trace on our Dichlorosilane
                                  gas line. The all-welded 1/4" SS line
                                  is encapsulated with a 1/2" SS outer
                                  containment  line which is itself heat
                                  traced with a single strand of heat
                                  trace that runs the entire length of
                                  the coax assembly. The 120 foot line
                                  is insulated as shown in the attached
                                  photo. A portion of the heat-trace
                                  appears to have failed prematurely (it
                                  was installed less than one year ago)
                                  and we are wondering if the method of
                                  installation may be the cause.</div>
                                <div><br>
                                </div>
                                <div>The heat trace was not installed in
                                  a spiral fashion around the outer 1/2"
                                  tube. Rather it was installed in a
                                  straight fashion along its entire
                                  length with "heat trace fastening
                                  tape" located every four feet or so. A
                                  member of my team has suggested that
                                  such a straight rather than spiral
                                  installation may have caused hot spots
                                  (at the fastening locations) which may
                                  have in turn caused the failure.</div>
                                <div><br>
                                </div>
                                <div>I would appreciate hearing from the
                                  community on this point: Are the heat
                                  traces around your low pressure gas
                                  lines spiral-wound around the lines or
                                  are they installed in a straight
                                  fashion and somehow fastened along the
                                  entire length?</div>
                                <div><br>
                                </div>
                                <div>Other insights/suggestions on the
                                  proper heat tracing of gas lines by
                                  experts in the field as well as
                                  comments on possible causes of
                                  premature heat trace failure are very
                                  much welcome and appreciated. Thank
                                  you.</div>
                                <div><br>
                                </div>
                                <div>Regards,</div>
                                <div>Vito</div>
                                <div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        style="font-size: 14px; "
                                        face="Calibri"><font
                                          face="Calibri,sans-serif">--</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        style="font-size: 14px; "
                                        face="Calibri"><font
                                          face="Calibri,sans-serif">Vito
                                          Logiudice  </font>P.<font
                                          face="Calibri,sans-serif">Eng.</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        face="Calibri"><font
                                          style="font-size: 14px; "
                                          face="Calibri,sans-serif">Director
                                          of Operations, Quantum NanoFab</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        face="Calibri"><font
                                          style="font-size: 14px; "
                                          face="Calibri,sans-serif">University
                                          of Waterloo</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        face="Calibri"><font
                                          style="font-size: 14px; "
                                          face="Calibri,sans-serif">Lazaridis
                                          QNC 1207</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        face="Calibri"><font
                                          style="font-size: 14px; "
                                          face="Calibri,sans-serif">200
                                          University Avenue West</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        face="Calibri"><font
                                          style="font-size: 14px; "
                                          face="Calibri,sans-serif">Waterloo,
                                          ON           Canada N2L 3G1</font></font></font></div>
                                  <div><font class="Apple-style-span"><font
                                        class="Apple-style-span"
                                        face="Calibri"><font
                                          style="font-size: 14px; "
                                          face="Calibri,sans-serif">Tel.:
                                          (519) 888-4567  ext. 38703</font></font></font></div>
                                  <div><span style="font-size: 14px; ">Email: <a
                                        moz-do-not-send="true"
                                        href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a></span></div>
                                  <div><span style="font-size: 14px; ">Website: </span><a
                                      moz-do-not-send="true"
                                      href="https://fab.qnc.uwaterloo.ca/">https://fab.qnc.uwaterloo.ca</a></div>
                                  <div><br>
                                  </div>
                                </div>
                              </div>
                            </div>
                          </span></div>
                      </blockquote>
                      <blockquote type="cite">
                        <div><span>_______________________________________________</span><br>
                          <span>labnetwork mailing list</span><br>
                          <span><a moz-do-not-send="true"
                              href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a></span><br>
                          <span><a moz-do-not-send="true"
                              href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a></span><br>
                        </div>
                      </blockquote>
                    </div>
                  </div>
                </span></div>
              _______________________________________________<br>
              labnetwork mailing list<br>
              <a moz-do-not-send="true"
                href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br>
              <a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
            </blockquote>
          </div>
          <br>
        </div>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>