    <div class="moz-cite-prefix">I have a different opinion about
      chillers: A single central unit creates a single point of failure
      for the entire facility, and maintenance requires everything to be
      shut down at once. I would rather have the (several small)
      chillers where I can see them, and I don't trust facilities people
      to do the maintenance properly. Same advice for UPS. Out of sight,
      out of mind. Also, if you have numerous chillers then you should
      keep a spare handy, just as you would for vacuum pumps.<br>
      <div class="moz-signature"> <br>
        <font size="-2">--------------------------------<br>
          Michael Rooks <br>
          Yale Institute of Nanoscience and Quantum Engineering <br>
          <a href="http://nano.yale.edu">nano.yale.edu</a>
        </font>
        <br>
        <br>
        <br>
        <br>
      </div>
      On 03/16/2015 12:26 PM, Michael Khbeis wrote:<br>
    </div>
    <blockquote cite="mid:58156CB3-D8D3-4F19-A46C-BE921FEAFC64@uw.edu"
      type="cite">
      <meta http-equiv="content-type" content="text/html;
        charset=windows-1252">
      <div>Loik </div>
      <div><br>
      </div>
      <div>I think that having the electrical load converted to heat
        load and having the diversity factor is sufficient. You can
        subtract the heat load dissipated through chilled water but that
        is presumably also being fed by your primary chiller. I would
        err on the side of being conservative since cleanrooms always
        end up being packed to a maximum density of equipment and people
        during their lifespans - give yourself a 20% margin at the
        least. I also would not discount the impact of the filter fan
        unit heat dissipation.  Since you are recycling most of the air
        several times the joule heating from the FFUs will cause a
        noticeable temperature rise. Given the class of your Cleanroom
        the FFU density will not be as high but I would add them into
        the load calc. </div>
      <div><br>
      </div>
      <div>Finally, regarding central chiller vs multiple point of use
        units - much easier to keep spare parts on hand and do PMs on a
        central chiller unit as these are usually handled by the
        building facilities folks. In my opinion, multiple point of use
        chillers are a maintenance nightmare and you often end up buying
        something quickly to replace or augment a failing unit. It is
        nearly impossible to keep multiple point of use units running
        consistently if you don't have dedicated maintenance staff. The
        exception to this is if your tools all required different or
        special temperatures and you end up needing multiple
        water-to-water heat exchangers then point of use chillers may
        make sense. </div>
      <div><br>
      </div>
      <div>Best,<br>
        <br>
        <div><span style="background-color: rgba(255, 255, 255, 0);">Dr.
            Michael Khbeis</span>
          <div><span style="background-color: rgba(255, 255, 255, 0);">Washington
              Nanofab Facility<br>
              University of Washington<br>
              Fluke Hall, Box 352143<br>
              (O) <a moz-do-not-send="true" href="tel:206.543.5101"
                x-apple-data-detectors="true"
                x-apple-data-detectors-type="telephone"
                x-apple-data-detectors-result="0/0">206.543.5101</a><br>
              (C) <a moz-do-not-send="true" href="tel:443.254.5192"
                x-apple-data-detectors="true"
                x-apple-data-detectors-type="telephone"
                x-apple-data-detectors-result="0/1">443.254.5192</a><br>
              <a moz-do-not-send="true" href="mailto:khbeis@uw.edu"
                x-apple-data-detectors="true"
                x-apple-data-detectors-type="link"
                x-apple-data-detectors-result="0/2">khbeis@uw.edu</a></span></div>
        </div>
      </div>
      <div><br>
        On Mar 16, 2015, at 7:36 AM, Loïk GENCE <<a
          moz-do-not-send="true"
          href="mailto:Loik.gence@cetuc.puc-rio.br">Loik.gence@cetuc.puc-rio.br</a>>
        wrote:<br>
        <br>
      </div>
      <blockquote type="cite">
        <div>
          <meta content="text/html; charset=windows-1252"
            http-equiv="Content-Type">
          Hi Everyone,<br>
          <br>
          I am working on the design of a small cleanroom facility
          (70m2) ISO 5/6. Right now, I am dealing with the problem of
          thermal load,<br>
          Which is crucial here in Rio, where the average outside
          temperature is of 30 deg C with a relative humidity of  75-80
          %.<br>
          <br>
          I would like to have your help and/or comments on several
          points:<br>
          <br>
          Most of the manufacturers do not provide a value for the
          thermal load (why?). Assuming 2 or max 3 people working
          simultaneously in the cleanrooms, and usual cleanroom value:
          22 deg C - 55% humidity. The main issue is: how can I estimate
          the cooling power required for the Air Conditioning System
          (ACS)?<br>
          <br>
          I have the list of equipments we'll have and the corresponding
          electrical power. I assumed the heat generated by 2 users and
          added the total electrical power dissipated as heat inside of
          the cleanroom, using a simultaneity factor of 0.75 (all
          equipments won't be turned on at the same time) and neglected
          the heat produced by the filter fan and lightning. <br>
              <b>    </b><b><br>
          </b><b>        - Is it reasonable for dimensioning the ACS
            cooling power ? </b><br>
          <br>
          Also most of the equipments have a cooling water circuit. But
          the cooling power needed is well described in the installation
          manual provided with the equipments. <br>
          <br>
          <b>        - Should I subtract from the </b><b>ACS cooling
            power</b><b> value, </b><b> the cooling po</b><b>wer </b><b>from
            the chillers?</b><br>
          <br>
                  - <b>Do you think a central and unique chiller
            (placed outside of the building) is preferred to several
            smaller chillers placed closer to the equipment?</b><br>
          <br>
          <br>
          I thank you very much for your kind help or comments.<br>
          <br>
          <br>
          regards,<br>
          <br>
          Loïk.<br>
            <br>
          <pre class="moz-signature" cols="72">-- 
__________________________________________
Dr. Loïk Gence      

PUC-RIO/CETUC-LABSEM 

End Rua Marques de São Vicente, 225-Gavea
CEP:22451-900   Rio de Janeiro, RJ-Brasil

      (Telefone)      +55 (021) 3527-2193

        (Mobile)     +55 (021) 99156-5558

<a moz-do-not-send="true" class="moz-txt-link-abbreviated" href="mailto:loik.gence@cetuc.puc-rio.br">loik.gence@cetuc.puc-rio.br</a>
__________________________________________</pre>
        </div>
      </blockquote>
      <blockquote type="cite">
        <div><span>_______________________________________________</span><br>
          <span>labnetwork mailing list</span><br>
          <span><a moz-do-not-send="true"
              href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a></span><br>
          <span><a moz-do-not-send="true"
href="https://urldefense.proofpoint.com/v2/url?u=https-3A__www-2Dmtl.mit.edu_mailman_listinfo.cgi_labnetwork&d=AwMFaQ&c=-dg2m7zWuuDZ0MUcV7Sdqw&r=apnDUg1OD9ejswcjrIvVgS28NpQ7-FGy7Sl7_YPlupc&m=hBL35UTQaOURuTrkXiNi1rZCy7Gy_rS075IrmahPX18&s=ZF7WKY75a9SR06b7Ulb31qeww2Hm1AcHRtLpSb5eXJA&e="><font color="red"><b>The MTL Mail Server has detected a possible fraud attempt from "urldefense.proofpoint.com" claiming to be</b></font> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a></span><br>
        </div>
      </blockquote>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
labnetwork mailing list
<a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a class="moz-txt-link-freetext" href="https://urldefense.proofpoint.com/v2/url?u=https-3A__www-2Dmtl.mit.edu_mailman_listinfo.cgi_labnetwork&d=AwICAg&c=-dg2m7zWuuDZ0MUcV7Sdqw&r=apnDUg1OD9ejswcjrIvVgS28NpQ7-FGy7Sl7_YPlupc&m=hBL35UTQaOURuTrkXiNi1rZCy7Gy_rS075IrmahPX18&s=ZF7WKY75a9SR06b7Ulb31qeww2Hm1AcHRtLpSb5eXJA&e=">https://urldefense.proofpoint.com/v2/url?u=https-3A__www-2Dmtl.mit.edu_mailman_listinfo.cgi_labnetwork&d=AwICAg&c=-dg2m7zWuuDZ0MUcV7Sdqw&r=apnDUg1OD9ejswcjrIvVgS28NpQ7-FGy7Sl7_YPlupc&m=hBL35UTQaOURuTrkXiNi1rZCy7Gy_rS075IrmahPX18&s=ZF7WKY75a9SR06b7Ulb31qeww2Hm1AcHRtLpSb5eXJA&e=</a> 
</pre>
    </blockquote>
    <br>
  </body>
</html>