    Thank you all for your messages. It helps!<br>
    <br>
    I'll introduced ligthning and FF units in the thermal load
    calculation.<br>
    I can see that the <i>a priori</i> estimation of the diversity
    factor seems to be difficult.<br>
    Overestimating the thermal load is comfortable, but, here,  I really
    have to minimize the cost of the HVAC system choosing a proper
    size/power of the HVAC  because of our limited budget...<br>
    <br>
    Thanks again!<br>
    <br>
    Loïk.<br>
    <br>
    <br>
    <div class="moz-cite-prefix">On 16/03/2015 12:55, Mike Young wrote:<br>
    </div>
    <blockquote cite="mid:A2C906E2-EE44-4331-AA45-FB87D4DD12F1@nd.edu"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html;
        charset=windows-1252">
      Hi, Loïk. Welcome to the group.
      <div class=""><br class="">
      </div>
      <div class="">I would caution you about neglecting the heat
        contribution of the cleanroom lighting and the fan filter units.
        Make sure they are really, truly, negligible, because it is not
        difficult for them to dominate the total cleanroom heat load
        under some conditions! In addition, we find that almost all of
        our equipment is in operation almost all of the time, so before
        you use the .75 diversity factor, make sure it is realistic.</div>
      <div class=""><br class="">
      </div>
      <div class="">The ACS has to be designed to meet the worst-case
        heat load, which will be worst-case weather conditions, plus
        everything turned on and running, plus maximum personnel heat
        load, under conditions of worst-case (maximum) makeup air flow.
        Getting all of this right is one of the most challenging
        problems in cleanroom design.</div>
      <div class=""><br class="">
      </div>
      <div class="">Good luck!</div>
      <div class=""><br class="">
      </div>
      <div class="">--Mike</div>
      <div class=""><br class="">
      </div>
      <div class=""><br class="">
        <div>
          <blockquote type="cite" class="">
            <div class="">On Mar 16, 2015, at 10:36 AM, Loïk GENCE <<a
                moz-do-not-send="true"
                href="mailto:Loik.gence@cetuc.puc-rio.br" class="">Loik.gence@cetuc.puc-rio.br</a>>
              wrote:</div>
            <br class="Apple-interchange-newline">
            <div class="">
              <meta content="text/html; charset=windows-1252"
                http-equiv="Content-Type" class="">
              <div text="#000000" bgcolor="#FFFFFF" class=""> Hi
                Everyone,<br class="">
                <br class="">
                I am working on the design of a small cleanroom facility
                (70m2) ISO 5/6. Right now, I am dealing with the problem
                of thermal load,<br class="">
                Which is crucial here in Rio, where the average outside
                temperature is of 30 deg C with a relative humidity of 
                75-80 %.<br class="">
                <br class="">
                I would like to have your help and/or comments on
                several points:<br class="">
                <br class="">
                Most of the manufacturers do not provide a value for the
                thermal load (why?). Assuming 2 or max 3 people working
                simultaneously in the cleanrooms, and usual cleanroom
                value: 22 deg C - 55% humidity. The main issue is: how
                can I estimate the cooling power required for the Air
                Conditioning System (ACS)?<br class="">
                <br class="">
                I have the list of equipments we'll have and the
                corresponding electrical power. I assumed the heat
                generated by 2 users and added the total electrical
                power dissipated as heat inside of the cleanroom, using
                a simultaneity factor of 0.75 (all equipments won't be
                turned on at the same time) and neglected the heat
                produced by the filter fan and lightning. <br class="">
                    <b class="">    </b><b class=""><br class="">
                </b><b class="">        - Is it reasonable for
                  dimensioning the ACS cooling power ? </b><br class="">
                <br class="">
                Also most of the equipments have a cooling water
                circuit. But the cooling power needed is well described
                in the installation manual provided with the equipments.
                <br class="">
                <br class="">
                <b class="">        - Should I subtract from the </b><b
                  class="">ACS cooling power</b><b class=""> value, </b><b
                  class=""> the cooling po</b><b class="">wer </b><b
                  class="">from the chillers?</b><br class="">
                <br class="">
                        - <b class="">Do you think a central and unique
                  chiller (placed outside of the building) is preferred
                  to several smaller chillers placed closer to the
                  equipment?</b><br class="">
                <br class="">
                <br class="">
                I thank you very much for your kind help or comments.<br
                  class="">
                <br class="">
                <br class="">
                regards,<br class="">
                <br class="">
                Loïk.<br class="">
                  <br class="">
                <pre class="moz-signature" cols="72">-- 
__________________________________________
Dr. Loïk Gence      

PUC-RIO/CETUC-LABSEM 

End Rua Marques de São Vicente, 225-Gavea
CEP:22451-900   Rio de Janeiro, RJ-Brasil

      (Telefone)      +55 (021) 3527-2193

        (Mobile)     +55 (021) 99156-5558

<a moz-do-not-send="true" class="moz-txt-link-abbreviated" href="mailto:loik.gence@cetuc.puc-rio.br">loik.gence@cetuc.puc-rio.br</a>
__________________________________________</pre>
              </div>
              _______________________________________________<br
                class="">
              labnetwork mailing list<br class="">
              <a moz-do-not-send="true"
                href="mailto:labnetwork@mtl.mit.edu" class="">labnetwork@mtl.mit.edu</a><br
                class="">
              <a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br
                class="">
            </div>
          </blockquote>
        </div>
        <br class="">
        <div apple-content-edited="true" class="">
          <div style="color: rgb(0, 0, 0); letter-spacing: normal;
            orphans: auto; text-align: start; text-indent: 0px;
            text-transform: none; white-space: normal; widows: auto;
            word-spacing: 0px; -webkit-text-stroke-width: 0px;
            word-wrap: break-word; -webkit-nbsp-mode: space;
            -webkit-line-break: after-white-space;" class="">-- <br
              class="">
            <font class="" face="Consolas"><span style="font-size:
                11px;" class="">Michael P. Young                       
                        (574) 631-3268 (office)<br class="">
                Nanofabrication Specialist                      (574)
                631-4393 (fax)<br class="">
                Department of Electrical Engineering            (765)
                637-6302 (cell)<br class="">
                University of Notre Dame                         <a
                  moz-do-not-send="true" href="mailto:mike.young@nd.edu"
                  class="">mike.young@nd.edu</a><br class="">
                B-38 Stinson-Remick Hall<br class="">
                Notre Dame, IN 46556-5637</span></font></div>
        </div>
        <br class="">
      </div>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
__________________________________________
Dr. Loïk Gence      

LABSEM – PUC-RIO

End Rua Marques de São Vicente, 225-Gávea
CEP:22451-900 – Rio de Janeiro, RJ-Brasil

      (Telefone)      +55 (021) 3527-2193

      (Mobile)       +55 (021) 99156 5558

              <a class="moz-txt-link-abbreviated" href="mailto:loik.gence@cetuc.puc-rio.br">loik.gence@cetuc.puc-rio.br</a>
__________________________________________</pre>
  </body>
</html>