    Dear Jim,<br>
    <br>
    I like the idea of  a single external Chiller. This should be fine
    for most of our tools. However a few equipments are more sensitive
    demanding higher flow and or pressure...<br>
    <br>
    The combination an external chiller and a few smaller ones line
    could be the best solution for us, having a small backup chiller in
    case of urgent need...<br>
    <br>
    I am curious to know the number of technicians you have in a
    facility with 40-50 tools... <br>
    Here, ww can afford only one  ;-/<br>
    <br>
    <br>
    best,<br>
    Loïk.<br>
    <br>
    <br>
    <br>
    <br>
    <div class="moz-cite-prefix">Le 16/03/2015 17:25, Beall, James A. a
      écrit :<br>
    </div>
    <blockquote cite="mid:614FCF45-FCFE-4B0B-B085-F2FCA4619E2E@nist.gov"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html;
        charset=windows-1252">
      Loïk,
      <div><br>
      </div>
      <div>I agree with the caution about accounting for the heat from
        the FFU; it is significant.</div>
      <div><br>
      </div>
      <div>In our fabs we use a large external Process Chilled Water
        unit which provides cooling for 40-50 tools. Some of the
        equipment (like cryopump compressors and larger pumps which
        typically have larger => 3/8” diameter), cooling lines can be
        directly cooled by the PCW loop. Many other tools like electron
        gun evaporators and sputter guns have smaller cooling lines and
        you may have trouble directly connecting them to the PCW loop
        and achieving the required flows and pressures. </div>
      <div><br>
      </div>
      <div>We end up with many small heat exchangers each set up for the
        particular needs of our tools. Some run distilled water, some DI
        water, some glycol and some high temperature thermal fluids. The
        pumps are the weak point in these units.</div>
      <div><br>
      </div>
      <div>When connecting tools across your PCW loop be careful to
        include balancing or flow setting valves as you will have many
        connections in parallel. </div>
      <div><br>
      </div>
      <div>Also I prefer to use a non-closed system, i.e., one with an
        atmospheric pressure return tank. This can be automatically
        refilled with treated water as needed to account for leaks and
        losses. Having your return water dump into a non-pressurized
        tank creates much less back pressure than the closed loop
        pressurized systems. The back pressure can be a killer with many
        parallel connected tools with different impedances.</div>
      <div><br>
      </div>
      <div><br>
        <div apple-content-edited="true">
          <div style="color: rgb(0, 0, 0); font-family: Helvetica;
            font-style: normal; font-variant: normal; font-weight:
            normal; letter-spacing: normal; line-height: normal;
            orphans: 2; text-align: -webkit-auto; text-indent: 0px;
            text-transform: none; white-space: normal; widows: 2;
            word-spacing: 0px; -webkit-text-size-adjust: auto;
            -webkit-text-stroke-width: 0px; word-wrap: break-word;
            -webkit-nbsp-mode: space; -webkit-line-break:
            after-white-space; ">
              - Jim</div>
          <br class="Apple-interchange-newline">
          <br class="Apple-interchange-newline">
        </div>
        <br>
        <div>
          <div>On Mar 16, 2015, at 11:41 AM, Iulian Codreanu <<a
              moz-do-not-send="true" href="mailto:codreanu@udel.edu">codreanu@udel.edu</a>>
            wrote:</div>
          <br class="Apple-interchange-newline">
          <blockquote type="cite">
            <div bgcolor="#FFFFFF" text="#000000">Loik,<br>
              <br>
              Another thing to keep in mind is the heat that is "taken
              away from the equipment" via exhaust.<br>
              <br>
              The heat generated by a piece of equipment can:<br>
              -Be dissipated into the room<br>
              -Be "taken away" by the cooling water.<br>
              -Be "taken way" by exhaust.<br>
              <br>
              If a piece of equipment is not cooled and does not have
              exhaust then all the heat gets dissipated into the room.<br>
              <br>
              If the piece of equipment is cooled and exhausted only
              part of the heat generated will end up in the room.<br>
              <br>
              If you know how much heat is taken away by the cooling
              water and/or exhaust that is great.  If you do not, I saw
              engineers using factor such as (PCW stands for process
              cooling water):<br>
              100% of the power into  the room if no PCW and no exhaust<br>
              50% of the power into  the room is no PCW but exhaust<br>
              40% of the power into  the room if PCW but no exhaust<br>
              20% of the power into  the room if both PCW and exhaust<br>
              <br>
              I prefer the "unique chiller (placed outside of the
              building)" approach for cooling the equipment.  If you can
              run the water from the unique chiller directly through the
              equipment, you do not have to deal with the heat generated
              by the "small chillers".<br>
              <br>
              Iulian<br>
              <br>
              <br>
              <br>
              <pre class="moz-signature" cols="72">iulian Codreanu, Ph.D.
Director of Operations, UD NanoFab
163 ISE Lab
221 Academy Street
Newark, DE 19716
302-831-2784
<a moz-do-not-send="true" class="moz-txt-link-freetext" href="http://udnf.udel.edu/">http://udnf.udel.edu</a></pre>
              <div class="moz-cite-prefix">On 3/16/2015 10:36 AM, Loïk
                GENCE wrote:<br>
              </div>
              <blockquote cite="mid:5506EA62.9000008@cetuc.puc-rio.br"
                type="cite">Hi Everyone,<br>
                <br>
                I am working on the design of a small cleanroom facility
                (70m2) ISO 5/6. Right now, I am dealing with the problem
                of thermal load,<br>
                Which is crucial here in Rio, where the average outside
                temperature is of 30 deg C with a relative humidity of 
                75-80 %.<br>
                <br>
                I would like to have your help and/or comments on
                several points:<br>
                <br>
                Most of the manufacturers do not provide a value for the
                thermal load (why?). Assuming 2 or max 3 people working
                simultaneously in the cleanrooms, and usual cleanroom
                value: 22 deg C - 55% humidity. The main issue is: how
                can I estimate the cooling power required for the Air
                Conditioning System (ACS)?<br>
                <br>
                I have the list of equipments we'll have and the
                corresponding electrical power. I assumed the heat
                generated by 2 users and added the total electrical
                power dissipated as heat inside of the cleanroom, using
                a simultaneity factor of 0.75 (all equipments won't be
                turned on at the same time) and neglected the heat
                produced by the filter fan and lightning.
                <br>
                    <b>    </b><b><br>
                </b><b>        - Is it reasonable for dimensioning the
                  ACS cooling power ? </b><br>
                <br>
                Also most of the equipments have a cooling water
                circuit. But the cooling power needed is well described
                in the installation manual provided with the equipments.
                <br>
                <br>
                <b>        - Should I subtract from the </b><b>ACS
                  cooling power</b><b> value, </b><b> the cooling po</b><b>wer
                </b><b>from the chillers?</b><br>
                <br>
                        - <b>Do you think a central and unique chiller
                  (placed outside of the building) is preferred to
                  several smaller chillers placed closer to the
                  equipment?</b><br>
                <br>
                <br>
                I thank you very much for your kind help or comments.<br>
                <br>
                <br>
                regards,<br>
                <br>
                Loïk.<br>
                  <br>
                <pre class="moz-signature" cols="72">-- 
__________________________________________
Dr. Loïk Gence      

PUC-RIO/CETUC-LABSEM 

End Rua Marques de São Vicente, 225-Gavea
CEP:22451-900   Rio de Janeiro, RJ-Brasil

      (Telefone)      +55 (021) 3527-2193

        (Mobile)     +55 (021) 99156-5558

<a moz-do-not-send="true" class="moz-txt-link-abbreviated" href="mailto:loik.gence@cetuc.puc-rio.br">loik.gence@cetuc.puc-rio.br</a>
__________________________________________</pre>
                <br>
                <fieldset class="mimeAttachmentHeader"></fieldset>
                <br>
                <pre wrap="">_______________________________________________
labnetwork mailing list
<a moz-do-not-send="true" class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a moz-do-not-send="true" class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</pre>
              </blockquote>
              <br>
            </div>
            _______________________________________________<br>
            labnetwork mailing list<br>
            <a moz-do-not-send="true"
              href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br>
            <a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
          </blockquote>
        </div>
        <br>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
labnetwork mailing list
<a class="moz-txt-link-abbreviated" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>
<a class="moz-txt-link-freetext" href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</pre>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
__________________________________________
Dr. Loïk Gence      

PUC-RIO/CETUC-LABSEM 

End Rua Marques de São Vicente, 225-Gavea
CEP:22451-900   Rio de Janeiro, RJ-Brasil

      (Telefone)      +55 (021) 3527-2193

        (Mobile)     +55 (021) 99156-5558

<a class="moz-txt-link-abbreviated" href="mailto:loik.gence@cetuc.puc-rio.br">loik.gence@cetuc.puc-rio.br</a>
__________________________________________</pre>
  </body>
</html>