<div dir="ltr">Iain,<div><br></div><div>one thing I forgot to mention, both of our systems are fabricated from CPVC materials, so we don't have much in the way of metal ions in the stream.  The pumps are stainless.</div></div><div class="gmail_extra"><br clear="all"><div><div class="gmail_signature"><div>Dennis</div><div> </div><div>734.647.2055 Ofc</div><div> </div><div>"People can be divided into 3 groups - those that make things happen, those that watch things happen, and those that wonder what happened."  Within which group do you belong?</div></div></div>
<br><div class="gmail_quote">On Tue, Sep 15, 2015 at 2:16 AM, Anteney I.M. <span dir="ltr"><<a href="mailto:I.M.Anteney@soton.ac.uk" target="_blank">I.M.Anteney@soton.ac.uk</a>></span> wrote:<br><blockquote class="gmail_quote" style="margin:0 0 0 .8ex;border-left:1px #ccc solid;padding-left:1ex">Denis,<br>
<br>
I'm particularly interested in your comment about polishing the loops to get 1 mega-ohm resistance as we have struggled to do this especially when we dose the system with biocide and corrosion inhibitors. What are you using to polish the system, what inhibitors/biocides do you use and how do these affect the conductivity.<br>
<br>
Regards<br>
<br>
Iain<br>
<br>
Cleanroom Manager<br>
University of Southampton<br>
<br>
Sent from my iPhone<br>
<br>
On 15 Sep 2015, at 00:36, Dennis Schweiger <<a href="mailto:schweig@umich.edu">schweig@umich.edu</a><mailto:<a href="mailto:schweig@umich.edu">schweig@umich.edu</a>>> wrote:<br>
<br>
Vito,<br>
<br>
here at UofMichigan, one of our PCW systems runs about 62.5F (16.9C) with a daily deviation of about 1F. The second runs about 59-60F (15C).  It too has a daily deviation of about 1F.  For both systems we may see a seasonal deviation of 3-4F as the main chiller plant transitions from "free cooling" to the use of chillers to create cooling water for the campus wide loop.  We also "polish" both loops to 1 meg-ohm so that we have a "high resistance" in the RF cooling circuits.<br>
<br>
Dennis Schweiger<br>
University of Mighigan/LNF<br>
<br>
734.647.2055 Ofc<br>
<br>
On Fri, Sep 11, 2015 at 3:47 PM, Vito Logiudice <<a href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a><mailto:<a href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>>> wrote:<br>
Dear Colleagues,<br>
<br>
The process chilled water loop in our cleanroom has consistently oscillated between ~15.4C and ~17.2C over a span of roughly 10 minutes. Our Plant Operations group is looking into the possibility of tightening this up for us by tuning the control sequence. They have asked for guidance on an acceptable specification.<br>
<br>
In light of this I would appreciate hearing from you as to your facility’s PCW temperature target and tolerance. Thank you much.<br>
<br>
Best regards,<br>
Vito<br>
--<br>
Vito Logiudice  P.Eng.<br>
Director of Operations, Quantum NanoFab<br>
University of Waterloo<br>
Lazaridis QNC 1207<br>
200 University Avenue West<br>
Waterloo, ON           Canada N2L 3G1<br>
Tel.: (519) 888-4567  ext. 38703<tel:%28519%29%20888-4567%20%C2%A0ext.%2038703><br>
Email: <a href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a><mailto:<a href="mailto:vito.logiudice@uwaterloo.ca">vito.logiudice@uwaterloo.ca</a>><br>
Website: <a href="https://fab.qnc.uwaterloo.ca" rel="noreferrer" target="_blank">https://fab.qnc.uwaterloo.ca</a><br>
<br>
<br>
_______________________________________________<br>
labnetwork mailing list<br>
<a href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><mailto:<a href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>><br>
<a href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork" rel="noreferrer" target="_blank">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
<br>
<br>
_______________________________________________<br>
labnetwork mailing list<br>
<a href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><mailto:<a href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a>><br>
<a href="https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork" rel="noreferrer" target="_blank">https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
</blockquote></div><br></div>