<html>
<head>
<meta http-equiv="Content-Type" content="text/html; charset=us-ascii">
</head>
<body dir="auto">
<div>Hi Kevin,</div>
<div id="AppleMailSignature"><br>
</div>
<div id="AppleMailSignature">Thanks for the reply. Our system is 3 years old and uses 6 inch wafers. We set our lur spec at 170 for 15 torr and 100 for 10 torr. I keep an eye on the flow at 15 torr and I like to see less than 9 sccm of helium. You are consistent
 with what I observe. Our lur are higher than when the system was new but this is to be expected. We do not see any burned resist and our etches are consistent. I Was asked by a fellow engineer in another lab and wanted to hear of other observations from people
 using esc's to confirm I wasn't giving bad information. Thanks for your reply.</div>
<div id="AppleMailSignature">Dave<br>
<br>
Sent from my iPhone</div>
<div><br>
On Jun 13, 2016, at 4:17 PM, Kevin Owen <<a href="mailto:kjvowen@lnf.umich.edu">kjvowen@lnf.umich.edu</a>> wrote:<br>
<br>
</div>
<blockquote type="cite">
<div>
<div dir="ltr">
<div>We have 2 Pegasus systems, one 4" and one 6", and a 6" APS, all with 6kV esc's. Assuming the same software and recipe format, it's in the "leak check" step defined for each recipe. There should be a warning level and a fault level. Ours is set for 125
 mTorr/min warning, 150 mTorr/min fault. On the APS, it's higher (200 mTorr/min fault), since that chuck is older and more worn and we don't get as good leak checks on it. It doesn't really affect process as long as the chuck and wafer are clean so you are
 getting consistent results wafer-to-wafer.</div>
<div><br>
</div>
We typically run 10 Torr backside pressure, most recipes are either 10C or 20C (although temperature doesn't really seem to matter for this). With brand new, shiny ESCs, we've seen anywhere from 10-40 mTorr/min. Leak rates then slowly increase over time as
 the chucks wear. With proper care, they seem to last a few years before it gets uncomfortably high (to me, uncomfortable is ~150 mTorr/min) and we replace them. Currently, both Pegasus tools are somewhere in the 40-70 mTorr/min range, and the APS is at 100-120.
<div><br>
</div>
<div>I'd say the biggest factor that will affect this is the cleanliness of the back of the wafer. Many years ago, when our tools were still new, we would have very regular (probably once per day) leak check failures due to particulate on the back of their
 samples. What sort of leak rates are you seeing? What's your chamber leak rate?<br>
<div>
<div><br>
</div>
<div>-Kevin</div>
</div>
</div>
</div>
<div class="gmail_extra"><br>
<div class="gmail_quote">On Mon, Jun 13, 2016 at 2:03 PM, LaFleur, David W <span dir="ltr">
<<a href="mailto:dlafleur@cns.fas.harvard.edu" target="_blank">dlafleur@cns.fas.harvard.edu</a>></span> wrote:<br>
<blockquote class="gmail_quote" style="margin:0 0 0 .8ex;border-left:1px #ccc solid;padding-left:1ex">
<div style="word-wrap:break-word;color:rgb(0,0,0);font-size:14px;font-family:Calibri,sans-serif">
<div>To Engineers who maintain Pegasus – Rapier Systems with Esc</div>
<div><br>
</div>
<div>Where do  you set your 15 Torr leak up rate pass/fail limit for the Esc. Does anyone have success in achieving the factory spec of 50mt/min on a 150mm chuck for 15 torr @15degC ?</div>
<div>What is your max flow you like to see for 15 Torr?</div>
<div>Thanks in advance.</div>
<div><br>
</div>
<div>David LaFleur</div>
<div>Equipment Engineer</div>
<div>Harvard University</div>
</div>
<br>
_______________________________________________<br>
labnetwork mailing list<br>
<a href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br>
<a href="https://urldefense.proofpoint.com/v2/url?u=https-3A__www-2Dmtl.mit.edu_mailman_listinfo.cgi_labnetwork&d=CwMFaQ&c=WO-RGvefibhHBZq3fL85hQ&r=6eLRBFWstHA_77uCRuGQiAtV9vZWeQjYS4oSMBfEHZE&m=098yfwLpJf6UPqbDciuqLNOmpnUdkfb6BZ29gLZuSsc&s=5e8tLMB3K9Ayee8HxBCndE_5XA4YJx48J1UL2zRTxME&e=" rel="noreferrer" target="_blank"><font color="red"><b>The MTL Mail Server has detected a possible fraud attempt from "urldefense.proofpoint.com" claiming to be</b></font> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
<br>
</blockquote>
</div>
<br>
<br clear="all">
<div><br>
</div>
-- <br>
<div class="gmail_signature" data-smartmail="gmail_signature">
<div dir="ltr">
<div>
<div dir="ltr">
<div>Kevin Owen
<div>Senior Engineer in Research</div>
<div>Operations Group, Lurie Nanofabrication Facility</div>
<div>University of Michigan</div>
<div>(734) 545-4014</div>
</div>
</div>
</div>
</div>
</div>
</div>
</div>
</blockquote>
</body>
</html>