<font color='black' size='2' face='arial'>Hello Aamer,
<div><br>
</div>
<div>Out of curiosity, what type of high vacuum pump are you using? If you are using a cryo pump, I would be very worried about oxygen build up in the pump! Is the voltage adjustable on your E-gun power supply and how low can it be adjusted? </div>
<div><br>
</div>
<div>The 1st thing that comes to mind is arcing at higher pressures from the HV feed thru up to the gun. You would of course need to recenter the beam at lower voltage and you may need to run at a lower beam current......</div>
<div><br>
</div>
<div>Dennis J. Culver </div>
<div>Culver Consulting</div>
<div><br>
</div>
<div>5960 Valentine Road</div>
<div>Unit 14, Mail box 13</div>
<div>Ventura, CA 93003</div>
<div><br>
</div>
<div>Ph (805) 648-1184</div>
<div>Fx (805) 648-1485</div>
<div><br>
</div>
<div>Cell(805)680-7389</div>
<div><br>
</div>
<div>dennis@culverequipment.com <br>
<br>
<br>
<div style="font-family:arial,helvetica;font-size:10pt;color:black">-----Original Message-----<br>
From: Dr. Aamer Mahmood <amahmood@hbku.edu.qa><br>
To: labnetwork@mtl.mit.edu <labnetwork@mtl.mit.edu><br>
Sent: Thu, Feb 7, 2019 1:28 pm<br>
Subject: [labnetwork] Chamber pressure during reactive e-beam evaporation<br>
<br>
<div id="yiv7010235799">
<div>
<div style="direction:ltr;font-family:Tahoma;color:#000000;font-size:10pt;">Hello all,
<div>We have been doing reactive oxide depositions with e-beam evaporation of metals and a few sccm oxygen flowing through our system. We do not have a variable position gate valve and can only control the oxygen MFC settings. The corresponding chamber pressure
has been less than 2e-4 Torr. </div>
<div>I have recently received a request to deposit reactive metal oxide films at higher oxygen flow rates that'll correspond to higher chamber pressure during deposition.</div>
<div>I am hesitant to increase the deposition chamber pressure due to the increased flow rate because the manufacturer mentioned the aforementioned chamber pressure as an acceptable upper limit for the e-gun. </div>
<div>I'd appreciate any insights/experiences on this matter particularly 1) how does increased oxygen pressure in the chamber affect the system hardware, 2) any safety concerns etc.</div>
<div>Thanks in advance.</div>
<div><br>
</div>
<div>-Aamer Mahmood</div>
<div>Qatar Environment and Energy Research Institute</div>
<div>Doha, Qatar</div>
</div>
<br>
This email message, including any attachments is intended solely for the addressee(s), and may contain information that is confidential or legally protected. Any use, disclosure, copying, printing or distribution of the information contained herein by persons other than the designated addressee is unauthorized and may be unlawful. If you believe that you have received this email in error, please notify the sender immediately and delete this email and any attachments from your system and discard any printout thereof. Qatar Foundation makes no warranty that this email is error- or virus-free .The views, opinions, conclusions, and other information expressed in this email are not necessarily endorsed by Qatar Foundation unless otherwise indicated by an authorized representative. As such, Qatar Foundation does not accept any liability for any emails and attachments that may violate the relevant applicable laws.<br>
</div>
</div>
_______________________________________________<br>
labnetwork mailing list<br>
<a ymailto="mailto:labnetwork@mtl.mit.edu" rel="noopener noreferrer" href="mailto:labnetwork@mtl.mit.edu">labnetwork@mtl.mit.edu</a><br>
<a target="_blank" rel="noopener noreferrer" href="https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork">https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a><br>
</div>
</div>
</font>