<html>
<head>
<meta http-equiv="Content-Type" content="text/html; charset=Windows-1252">
<style type="text/css" style="display:none;"> P {margin-top:0;margin-bottom:0;} </style>
</head>
<body dir="ltr">
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Hi Tim,</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Yes, this issue is endemic to these units, as all the deposition gases pass over the sealing surfaces of the valve under normal operating conditions.</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
One solution is to maintain a stock of the appropriate (but expensive) Kalrez-type orings, and do a valve sealing-face cleanup at intervals.  Eventually, however, you will find that deposition has encrusted the interior of the valve bellows, rendering it much
 less flexible than original equipment.</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
I am happy to share with you via PM some plumbing "rearrangements" that we have instituted which seem to have helped this problem greatly.</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Sincerely,</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Mac Hathaway</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Senior Process and Systems Engineer<br>
</div>
<div style="font-family: Calibri, Arial, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Harvard CNS<br>
</div>
<div id="appendonsend"></div>
<hr style="display:inline-block;width:98%" tabindex="-1">
<div id="divRplyFwdMsg" dir="ltr"><font face="Calibri, sans-serif" style="font-size:11pt" color="#000000"><b>From:</b> labnetwork <labnetwork-bounces@mtl.mit.edu> on behalf of Tim Gilheart <gilheart@rice.edu><br>
<b>Sent:</b> Tuesday, March 30, 2021 6:13 PM<br>
<b>To:</b> Fab Network <labnetwork@mtl.mit.edu><br>
<b>Subject:</b> [labnetwork] Cambridge/Ultratech/Veeco Savannah ALD - stop valve issues?</font>
<div> </div>
</div>
<div class="" style="word-wrap:break-word; line-break:after-white-space">Greetings,
<div class=""><br class="">
</div>
<div class="">I’m hoping that several teams on this list have relevant experience to share, since I know a number of you manage Savannah ALD tools.</div>
<div class=""><br class="">
</div>
<div class="">Our tool has presented with leakage issues through the stop valve twice in the past 6 months. We replaced the last valve at the end of Aug 2020, and have recently again experienced issues with failing to vent (due to the valve not closing completely
 and the vac pump still pulling on the chamber).</div>
<div class=""><br class="">
</div>
<div class="">We have a temporary workaround in place (an external manual valve on the pump line), but this leakage through the stop valve effects process outcomes and has proven costly to address, since valve replacement seems to be the recommended solution.</div>
<div class=""><br class="">
</div>
<div class="">Our tool typically runs Al2O3, HfO2, TiO2, and ZnO processes, sometimes ZrO2. Stop valve heating jacket is maintained at the recommended 150 C, and the heaters all seem to be working properly.</div>
<div class=""><br class="">
</div>
<div class="">Does anyone have suggestions about maintaining these stop valves for longer? Or, failing that, a referral for refurbishing them? I am willing to consider a solution that involves swapping out valves 1-2/year and sending them out for rebuild if
 that’s an option for a reasonable cost.</div>
<div class="">
<div class=""><br class="x_webkit-block-placeholder">
</div>
<div class="">Thanks in advance for any advice you can share,</div>
<div class="">
<div dir="auto" class="" style="color:rgb(0,0,0); letter-spacing:normal; text-align:start; text-indent:0px; text-transform:none; white-space:normal; word-spacing:0px; text-decoration:none; word-wrap:break-word; line-break:after-white-space">
<br class="">
-- <br class="">
Tim Gilheart, Ph.D.<br class="">