Hi Katharine,</div>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
At Utah Nanofab fume hoods, only solvents (no acids or bases) are handled, and we require the safety glasses as well as double nitrile gloves. The second set is a different color than the first set to give a visual confirmation the wearer chose to work with
a wet chemical and has a more limited sphere of operation outside the fume hood (touch nothing but the handles to the under-fume-hood storage) than when wearing the single layer of nitrile gloves.</div>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
At our wetbenches, where only acids and bases are used (no solvents), we require not only the safety glasses (already on and not ever removed in the lab) and second layer of nitrile gloves, but a chemical resistant apron and face shield, plus a third set of
gloves, Neoprene. The Neoprene gloves are washed and reused. Second set of nitrile gloves are removed and discarded at the end of operating in the wetbench. First set of gloves and safety glasses remain on while in the Cleanroom.</div>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
Kathy</div>
<div id="Signature" class="elementToProof">
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
<table cellspacing="0" cellpadding="0" border="0" style="text-align: left; background-color: rgb(255, 255, 255); margin: 30px 0px; width: 100%; color: rgb(71, 73, 74); box-sizing: border-box; border-collapse: collapse; border-spacing: 0px;">
<tbody>
<tr>
<td style="text-align: left; vertical-align: top; width: 110px;"><span style="font-family: myriad-pro, sans-serif; font-size: 17px; color: rgb(71, 73, 74);"><img alt="Block U Logo" width="110" style="width: 110px; height: auto; max-width: 100%; display: block;" src="https://digital.utah.edu/wp-content/uploads/sites/2/2024/08/block-u-signature.png"></span></td>
<td style="text-align: left; line-height: 1.2; vertical-align: top; color: rgb(128, 128, 128);">
<div class="elementToProof" style="text-align: left; line-height: 1.2; font-family: Arial;">
<span style="font-size: 14px; color: rgb(102, 102, 102);"><b>Kathy Anderson</b></span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(102, 102, 102);"><b><i>Chemical Engineer III</i></b></span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(128, 128, 128);">Utah Nanofab</span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(128, 128, 128);">The University of Utah</span><span style="font-size: 12px;"><br>
O: </span><span style="font-size: 12px; color: rgb(128, 128, 128);">801-213-6555</span><span style="font-size: 12px;"> | C:
</span><span style="font-size: 12px; color: rgb(128, 128, 128);">801-440-1465</span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(128, 128, 128);">https://www.nanofab.utah.edu</span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(128, 128, 128);">36 S Wasatch Drive Suite 2500</span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(128, 128, 128);">Salt Lake City, UT 84112</span><span style="font-size: 12px;"><br>
</span><span style="font-size: 12px; color: rgb(102, 102, 102);"><i>kathy.anderson@utah.edu</i></span></div>
</td>
</tr>
</tbody>
</table>
<div class="elementToProof" style="font-family: Aptos, Aptos_EmbeddedFont, Aptos_MSFontService, Calibri, Helvetica, sans-serif; font-size: 12pt; color: rgb(0, 0, 0);">
<br>
</div>
</div>
<div id="appendonsend"></div>
<hr style="display:inline-block;width:98%" tabindex="-1">
<div id="divRplyFwdMsg" dir="ltr"><font face="Calibri, sans-serif" style="font-size:11pt" color="#000000"><b>From:</b> labnetwork <labnetwork-bounces@mtl.mit.edu> on behalf of Mark Chiappa <mark.chiappa@ntnu.no><br>
<b>Sent:</b> Wednesday, August 12, 2026 12:38 AM<br>
<b>To:</b> Prashant Srinivasan <prashant@ucsb.edu>; Beaudoin, Mario <beaudoin@physics.ubc.ca>; labnetwork@mtl.mit.edu <labnetwork@mtl.mit.edu><br>
<b>Subject:</b> Re: [labnetwork] Solvent Bench PPE</font>
<div> </div>
</div>
<div>
<table border="0" cellspacing="0" cellpadding="0" width="100%" align="left" style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important; border:0!important; display:table!important; width:100%!important; table-layout:fixed!important; border-collapse:seperate!important; float:none!important; border-spacing:0px 0px!important">
<tbody style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important; display:block!important">
<tr style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important">
<td valign="middle" width="1px" bgcolor="#A6A6A6" cellpadding="7px 2px 7px 2px" style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important; padding:7px 2px 7px 2px!important; background-color:#A6A6A6!important; width:0px!important">
</td>
<td valign="middle" width="100%" bgcolor="#EAEAEA" cellpadding="7px 5px 7px 15px" color="#212121" style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important; width:100%!important; background-color:#EAEAEA!important; padding:7px 5px 7px 15px!important; font-family:wf_segoe-ui_normal,Segoe UI,Segoe WP,Tahoma,Arial,sans-serif!important; font-size:12px!important; font-weight:normal!important; color:#212121!important; text-align:left!important; word-wrap:break-word!important">
<div style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important">
You don't often get email from mark.chiappa@ntnu.no. <a href="https://aka.ms/LearnAboutSenderIdentification" style="background:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; opacity:revert!important; visibility:revert!important">
Learn why this is important</a> </div>
</td>
<td valign="middle" align="left" width="75px" bgcolor="#EAEAEA" cellpadding="7px 5px 7px 5px" color="#212121" style="background:revert!important; border:revert!important; bottom:revert!important; color:revert!important; direction:revert!important; display:revert!important; font-size:revert!important; height:revert!important; letter-spacing:revert!important; line-height:revert!important; margin:revert!important; opacity:revert!important; order:revert!important; outline:revert!important; overflow:revert!important; padding:revert!important; position:revert!important; tab-size:revert!important; table-layout:revert!important; text-align:revert!important; text-indent:revert!important; text-orientation:revert!important; text-overflow:revert!important; text-transform:revert!important; top:revert!important; vertical-align:revert!important; visibility:revert!important; white-space:revert!important; width:revert!important; word-break:revert!important; word-spacing:revert!important; writing-mode:revert!important; zoom:revert!important; width:75px!important; background-color:#EAEAEA!important; padding:7px 5px 7px 5px!important; font-family:wf_segoe-ui_normal,Segoe UI,Segoe WP,Tahoma,Arial,sans-serif!important; font-size:12px!important; font-weight:normal!important; color:#212121!important; text-align:left!important; word-wrap:break-word!important">
</td>
</tr>
</tbody>
</table>
<div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
Hi Katharine,</div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
<br>
</div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
Here at NTNU NanoLab we use Safety glasses and nitrile gloves for normal operations. When users clean Spinner bowl sets we insist they use an additional pair of thicker gloves
<a href="https://www.vwr.com/no/en/product/16702400/cleanroom-gloves-nitrile-advantech-519" data-outlook-id="6bc035bc-a6ba-4e38-be42-6782c24f48ca">
https://www.vwr.com/no/en/product/16702400/cleanroom-gloves-nitrile-advantech-519</a> </div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
<br>
</div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
Kind regards</div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
Mark </div>
<div style="direction:ltr; font-family:Aptos,Arial,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
<br>
</div>
<div id="x_ms-outlook-mobile-signature">
<div class="x_MsoNormal" style="direction:ltr">Sent from Outlook for Mac</div>
</div>
<div id="x_mail-editor-reference-message-container">
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
</div>
<div style="padding:3pt 0in 0in; border-width:1pt medium medium; border-style:solid none none; border-color:rgb(181,196,223) currentcolor currentcolor">
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="text-align:left; font-family:Aptos; font-size:12pt; color:black">
<b>From: </b>labnetwork <labnetwork-bounces@mtl.mit.edu> on behalf of Prashant Srinivasan <prashant@ucsb.edu><br>
<b>Date: </b>Wednesday, 12 August 2026 at 02:30<br>
<b>To: </b>Beaudoin, Mario <beaudoin@physics.ubc.ca>; labnetwork@mtl.mit.edu <labnetwork@mtl.mit.edu><br>
<b>Subject: </b>Re: [labnetwork] Solvent Bench PPE<br>
<br>
</div>
</div>
<p class="x_ms-outlook-mobile-reference-message x_skipProofing">Katharine,</p>
<p class="x_ms-outlook-mobile-reference-message x_skipProofing">Our PPE is limited to full coverage splash googles and 1 layer of Nitrle gloves. We only use Acetone, 2-Propanol and Toluene in our teaching cleanroom.</p>
<p class="x_ms-outlook-mobile-reference-message x_skipProofing">Prashant</p>
<div class="x_moz-cite-prefix">On 8/11/2026 3:36 PM, Beaudoin, Mario wrote:</div>
<blockquote>
<p class="x_ms-outlook-mobile-reference-message x_skipProofing">Katharine,</p>
<p class="x_ms-outlook-mobile-reference-message x_skipProofing">We demand protection glasses and recommend splash goggles. And 1 layer of nitrile gloves.</p>
<p class="x_ms-outlook-mobile-reference-message x_skipProofing">Mario</p>
<div class="x_moz-cite-prefix">On 8/11/2026 12:12 PM, Martin, Michael wrote:</div>
<blockquote>
<table class="x_ms-outlook-mobile-reference-message x_skipProofing" style="box-sizing:border-box; border-collapse:collapse; border-spacing:0px">
<tbody>
<tr>
<td class="x_ms-outlook-mobile-reference-message x_skipProofing">
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="font-family:Arial,sans-serif; font-size:12px; color:rgb(0,0,0)">
<span style="background-color:rgb(255,236,179); line-height:1.6">[<b>CAUTION:</b> Non-UBC Email]</span></div>
</td>
</tr>
</tbody>
</table>
<div class="x_elementToProof" style="font-family:Aptos,Aptos_EmbeddedFont,Aptos_MSFontService,Calibri,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
Hi Katharine, </div>
<div class="x_elementToProof" style="font-family:Aptos,Aptos_EmbeddedFont,Aptos_MSFontService,Calibri,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
We have goggles (of course) and 1 layer of nitrile gloves. </div>
<div class="x_elementToProof" style="direction:ltr; font-family:Aptos,Aptos_EmbeddedFont,Aptos_MSFontService,Calibri,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
<br>
</div>
<div class="x_elementToProof" style="font-family:Aptos,Aptos_EmbeddedFont,Aptos_MSFontService,Calibri,Helvetica,sans-serif; font-size:12pt; color:rgb(0,0,0)">
-Michael </div>
<hr style="display:inline-block; width:98%">
<div id="x_divRplyFwdMsg" dir="ltr"><span style="font-family:Calibri,sans-serif; font-size:11pt; color:rgb(0,0,0)"><b>From:</b> labnetwork
<a href="mailto:labnetwork-bounces@mtl.mit.edu" class="x_moz-txt-link-rfc2396E" data-outlook-id="f8aa3459-6215-4e92-851f-1cc27ec412dc">
<labnetwork-bounces@mtl.mit.edu></a> on behalf of Katharine Beach <a href="mailto:beach@umich.edu" class="x_moz-txt-link-rfc2396E" data-outlook-id="bc4bffb4-f071-4d6d-b0c6-3c04caab50f8">
<beach@umich.edu></a><br>
<b>Sent:</b> Tuesday, August 11, 2026 8:28 AM<br>
<b>To:</b> <a href="mailto:labnetwork@mtl.mit.edu" class="x_moz-txt-link-abbreviated x_moz-txt-link-freetext" data-outlook-id="31d41ec0-ee35-4e73-8e6f-1b13db2fc334">
labnetwork@mtl.mit.edu</a> <a href="mailto:labnetwork@mtl.mit.edu" class="x_moz-txt-link-rfc2396E" data-outlook-id="cddffe88-68bf-4f84-b482-42c8c7ea00ea">
<labnetwork@mtl.mit.edu></a><br>
<b>Subject:</b> [labnetwork] Solvent Bench PPE</span>
<div> </div>
</div>
<p class="x_x_MsoNormal" style="line-height:12pt; background-color:rgb(255,77,77)">
<span style="font-size:10pt; color:black"><b>CAUTION:</b> This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe.</span></p>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
Greetings,</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
At the LNF we are evaluating our Solvent Bench PPE requirements. I've been tasked with comparing our current PPE requirements with that of our peers'. If you would be so kind as to share links to this information, or just your requirements that would be greatly
appreciated.</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
Our current requirements are to wear your safety glasses, these should alway be on in the Cleanroom, and a pair of polychloroprene gloves in addition to the base nitrile gloves.</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
Thanks you,</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
Katharine Beach</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
--</div>
<table class="x_x_gmail_signature" cellspacing="0" cellpadding="0" border="0" style="direction:ltr; color:rgb(34,34,34)">
<tbody>
<tr>
<td colspan="3" class="x_x_gmail_signature" style="direction:ltr; padding-top:8px; padding-right:1px; vertical-align:middle">
<table class="x_x_gmail_signature" cellspacing="0" cellpadding="0" border="0" style="direction:ltr">
<tbody>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-right:9px; padding-left:1px; vertical-align:top">
<table class="x_x_gmail_signature" cellspacing="0" cellpadding="0" border="0" style="direction:ltr">
<tbody>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:17px; white-space:nowrap; padding-right:1px">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:17px; white-space:nowrap; margin:1px; font-weight:700; color:rgb(0,0,0)">
<span style="font-family:Arial,sans-serif; font-size:13px">Katharine Beach</span></p>
<p class="x_x_gmail_signature" style="direction:ltr; line-height:18px; white-space:nowrap; margin:1px; color:rgb(136,136,136)">
<span style="font-family:Arial,sans-serif; font-size:13px">Lead Research Engineer</span></p>
<p class="x_x_gmail_signature" style="direction:ltr; line-height:18px; white-space:nowrap; margin:1px; color:rgb(136,136,136)">
<span style="font-family:Arial,sans-serif; font-size:13px">Lurie Nanofabrication Facility</span></p>
<p class="x_x_gmail_signature" style="direction:ltr; line-height:18px; white-space:nowrap; margin:1px; color:rgb(136,136,136)">
<span style="font-family:Arial,sans-serif; font-size:13px">University of Michigan</span></p>
</td>
</tr>
</tbody>
</table>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; border-right:2px solid rgb(0,123,255); padding-top:1px">
<div class="x_x_gmail_signature" style="direction:ltr; color:rgb(34,34,34)"><br>
</div>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; padding-right:1px; padding-left:9px; vertical-align:top">
<table class="x_x_gmail_signature" cellspacing="0" cellpadding="0" border="0" style="direction:ltr">
<tbody>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-right:1px">
<table class="x_x_gmail_signature" cellspacing="0" cellpadding="0" border="0" style="direction:ltr">
<tbody>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-top:1px; padding-right:5px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; margin:1px"><span style="color:rgb(34,34,34)"><img width="18" height="18" style="width:18px; height:18px; max-width:100%; margin-top:0px; margin-bottom:0px; display:block" src="https://signaturehound.com/api/v1/png/email/default/007bff.png"></span></p>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:17px; padding-top:1px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:17px; margin:1px">
<span style="font-family:Arial,sans-serif; font-size:13px; color:rgb(136,136,136); line-height:17px"><a href="mailto:vpray@umich.edu" target="_blank" data-outlook-id="99a8dfac-6e9e-4fd0-90ab-28dd9da47a63" style="color:rgb(136,136,136); margin-top:0px; margin-bottom:0px">beach@umich.edu</a></span></p>
</td>
</tr>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-top:1px; padding-right:5px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; margin:1px"><span style="color:rgb(34,34,34)"><img style="max-width:100%; margin-top:0px; margin-bottom:0px; display:block" src="https://signaturehound.com/api/v1/png/phone/default/007bff.png"></span></p>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:17px; padding-top:1px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:17px; margin:1px">
<span style="font-family:Arial,sans-serif; font-size:13px; color:rgb(136,136,136); line-height:17px"><a href="tel:7347638105" target="_blank" data-outlook-id="1b3208b1-42b0-4da2-bf38-d0e68b7e0c37" style="color:rgb(136,136,136); margin-top:0px; margin-bottom:0px">(734)
255-9487</a></span></p>
</td>
</tr>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-top:1px; padding-right:5px; padding-bottom:1px; vertical-align:middle">
<div class="x_x_gmail_signature" style="direction:ltr; color:rgb(34,34,34)"><br>
</div>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:17px; padding-top:1px; padding-bottom:1px; vertical-align:middle">
<div class="x_x_gmail_signature" style="direction:ltr; line-height:17px; font-family:Arial,sans-serif; font-size:13px; color:rgb(0,123,255)">
<span style="line-height:17px; font-weight:700"><a href="https://calendly.com/beachlnf" target="_blank" originalsrc="https://calendly.com/beachlnf" data-outlook-id="0b2e9693-91ca-4d32-b238-331fdfc9622a" style="color:rgb(0,123,255)">Calendly</a></span></div>
</td>
</tr>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-top:1px; padding-right:5px; padding-bottom:1px; vertical-align:top">
<span style="color:rgb(34,34,34)"><img width="18" height="18" style="width:18px; height:18px; max-width:100%; display:block" src="https://signaturehound.com/api/v1/png/map/default/007bff.png"></span></td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:17px; padding-top:1px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:17px; margin:1px">
<span style="font-family:Arial,sans-serif; font-size:13px; color:rgb(136,136,136); line-height:17px">1301 Beal Avenue, 1241 EECS Building<br>
Ann Arbor, MI 48109-2122</span></p>
</td>
</tr>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; padding-top:1px; padding-right:5px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; margin:1px"><span style="color:rgb(34,34,34)"><img width="18" height="18" style="width:18px; height:18px; max-width:100%; margin-top:0px; margin-bottom:0px; display:block" src="https://signaturehound.com/api/v1/png/website/default/007bff.png"></span></p>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:17px; padding-top:1px; padding-bottom:1px; vertical-align:middle">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:17px; margin:1px">
<span style="font-family:Arial,sans-serif; font-size:13px; color:rgb(0,123,255); line-height:17px; font-weight:700"><a href="http://www.lnf.umich.edu/" target="_blank" originalsrc="http://www.lnf.umich.edu/" data-outlook-id="09f436e7-0f98-4ff1-94c5-b0525e913022" style="color:rgb(0,123,255); margin-top:0px; margin-bottom:0px">LNF.umich.edu</a></span></p>
</td>
</tr>
</tbody>
</table>
</td>
</tr>
</tbody>
</table>
</td>
</tr>
<tr>
<td colspan="3" class="x_x_gmail_signature" style="direction:ltr; border-bottom:2px solid rgb(0,123,255); padding-right:1px; padding-bottom:8px">
<div class="x_x_gmail_signature" style="direction:ltr; color:rgb(34,34,34)"><br>
</div>
</td>
</tr>
<tr>
<td colspan="3" class="x_x_gmail_signature" style="direction:ltr; padding-top:8px; padding-right:1px; vertical-align:middle">
<table class="x_x_gmail_signature" cellspacing="0" cellpadding="0" border="0" style="direction:ltr">
<tbody>
<tr>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:0px; padding-right:1px; width:30px">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:0px; margin:1px">
<span style="font-size:0px; color:rgb(17,85,204)"><a href="https://www.facebook.com/LurieNanofab" target="_blank" originalsrc="https://www.facebook.com/LurieNanofab" data-outlook-id="5e4cb0b0-267c-4564-bf86-251b4ff43050" style="color:rgb(17,85,204); margin-top:0px; margin-bottom:0px"><img width="30" height="30" style="width:30px; height:30px; max-width:100%; margin-top:0px; margin-bottom:0px; display:block" src="https://signaturehound.com/api/v1/png/facebook/default/007bff.png"></a></span></p>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; padding-bottom:1px; width:3px">
<div class="x_x_gmail_signature" style="direction:ltr; color:rgb(34,34,34)"><br>
</div>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:0px; padding-right:1px; width:30px">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:0px; margin:1px">
<span style="font-size:0px; color:rgb(17,85,204)"><a href="https://twitter.com/LurieNanoFab" target="_blank" originalsrc="https://twitter.com/LurieNanoFab" data-outlook-id="afa66731-f938-454c-ba94-40b823bbcce2" style="color:rgb(17,85,204); margin-top:0px; margin-bottom:0px"><img width="30" height="30" style="width:30px; height:30px; max-width:100%; margin-top:0px; margin-bottom:0px; display:block" src="https://signaturehound.com/api/v1/png/twitter/default/007bff.png"></a></span></p>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; padding-bottom:1px; width:3px">
<div class="x_x_gmail_signature" style="direction:ltr; color:rgb(34,34,34)"><br>
</div>
</td>
<td class="x_x_gmail_signature" style="direction:ltr; line-height:0px; padding-right:1px; width:30px">
<p class="x_x_gmail_signature" style="direction:ltr; line-height:0px; margin:1px">
<span style="font-size:0px; color:rgb(17,85,204)"><a href="https://www.linkedin.com/company/lurie-nanofabrication-facility" target="_blank" originalsrc="https://www.linkedin.com/company/lurie-nanofabrication-facility" data-outlook-id="75e0bc13-a72c-4626-b59b-376f82771a27" style="color:rgb(17,85,204); margin-top:0px; margin-bottom:0px"><img width="30" height="30" style="width:30px; height:30px; max-width:100%; margin-top:0px; margin-bottom:0px; display:block" src="https://signaturehound.com/api/v1/png/linkedin/default/007bff.png"></a></span></p>
</td>
</tr>
</tbody>
</table>
</td>
</tr>
</tbody>
</table>
</td>
</tr>
</tbody>
</table>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<pre><div class="x_moz-quote-pre">_______________________________________________
labnetwork mailing list
<a href="mailto:labnetwork@mtl.mit.edu" class="x_moz-txt-link-abbreviated x_moz-txt-link-freetext" data-outlook-id="69c57a6d-f19e-4696-a50e-2868fd380961">labnetwork@mtl.mit.edu</a>
<a href="https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork" class="x_moz-txt-link-freetext" data-outlook-id="9da01ac8-7208-418b-8ad1-10a716919198">https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</div></pre>
</blockquote>
<div class="x_moz-signature">--<br>
<img style="max-width:100%" data-outlook-trace="F:1|T:1" src="cid:part1.qBZ6qkQM.VbIOSXa0@ucsb.edu"></div>
<div class="x_ms-outlook-mobile-reference-message x_skipProofing" style="direction:ltr">
<br>
</div>
<pre><div class="x_moz-quote-pre">_______________________________________________
labnetwork mailing list
<a href="mailto:labnetwork@mtl.mit.edu" class="x_moz-txt-link-abbreviated" data-outlook-id="8c8622dd-1c4f-4903-a7a1-f18bbc536497">labnetwork@mtl.mit.edu</a>
<a href="https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork" class="x_moz-txt-link-freetext" data-outlook-id="b5ddf94f-463f-40b3-9c80-c38101ffc1bc">https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork</a>
</div></pre>
</blockquote>
<pre><div class="x_moz-signature">--
Prashant Srinivasan
Teaching Cleanroom Engineer
Electrical & Computer Engineering
Univ. of California Santa Barbara
<a href="https://www.ece.ucsb.edu/department-resources/electronics-shop/tcr" class="x_moz-txt-link-freetext" data-outlook-id="2ab6d357-d1a1-4cb0-80c4-19a8e91d95c2">https://www.ece.ucsb.edu/department-resources/electronics-shop/tcr</a></div></pre>
</div>
</div>
</div>
</body>
</html>