[labnetwork] Anneal temp for aluminum on silicon with forming gas
Henk van Zeijl - EWI
H.W.vanZeijl at tudelft.nl
Wed Nov 14 08:12:08 EST 2012
Dear Julia Aebersold
This is an Al/Si(1%) alloy recipe in use at Delft university, the Netherlands
PROCESS
TEMPERATURE
(in °C)
GASSES & FLOWS
(in liter/min)
TIME
(in minutes)
boat in
400
nitrogen: 3.0
5
stabilize 1
400
nitrogen: 3.0
5
stabilize2
400
nitrogen: 3.0
10
alloying
400
nitrogen: 3.0
hydrogen: 0.3
20
purge
400
nitrogen: 3.0
5
boat out
400
nitrogen: 3.0
5
We use Al(Si1%) to avoid junction spiking on shallow n+ p junctions. This recipe also improves the low current characteristics of bipolar npn transistors, see below (copy of lecture slides).
[cid:image001.png at 01CDC272.0D7368E0][cid:image002.png at 01CDC272.0D7368E0]
Kind regars, Henk van Zeijl
Delft University of Technology / DIMES
POBox 5053
NL-2600 GB DELFT
The Netherlands
phone: ++31 15 2781092
Mobile: ++31 6 24373305
fax: ++31 15 2787369
Email: h.w.vanzeijl at tudelft.nl
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Aebersold,Julia W.
Sent: 13 November 2012 21:12
To: Lab Network (labnetwork at mtl.mit.edu)
Subject: [labnetwork] Anneal temp for aluminum on silicon with forming gas
I wanted to gather some information of aluminum anneal on a silicon wafer performed by other cleanrooms. We are using 5% forming gas with an anneal temp of 400C for ½ hour with a 20C/min ramp up and ramp down However, I have also seen higher temps up to 475C, but was concerned about the flammability of 100% hydrogen per its MSDS.
Cheers!
Julia Aebersold, Ph.D.
MNTC Cleanroom Manager
Shumaker Research Building, Room 233
2210 South Brook Street
University of Louisville
Louisville, KY 40292
502-852-1572
http://louisville.edu/micronano/
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20121114/63d72ab4/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 104149 bytes
Desc: image001.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20121114/63d72ab4/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 93791 bytes
Desc: image002.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20121114/63d72ab4/attachment-0001.png>
More information about the labnetwork
mailing list