[labnetwork] experience with Germanium

J Provine jprovine at stanford.edu
Tue Jan 14 13:37:06 EST 2014


there is not much in the way of cross-contamination concern.  Ge will cool
much slower than the metals you mention, and if cool-ed rapidly can
"explode" in the vacuum making a real mess (and using up a lot of your
source).  establish an extra long cool down procedure for the users (5-10
minutes to ramp the temperature down) to avoid this.  then everything
should be fine.

j


On Tue, Jan 14, 2014 at 8:25 AM, Ferraguto, Thomas <Thomas_Ferraguto at uml.edu
> wrote:

>  Colleagues,
>
>
>
> We have a new CHA E-Beam / Thermal evaporator and there has been a some
> interest in depositing Germanium.
>
>
>
> What (if any) are the risks for allowing Germanium deposition in a system
> that does largely gold, titanium, aluminum and nickel?
>
>
>
> Thank you
>
>
>
> Thomas S. Ferraguto
>
> ETIC Nanofabrication Laboratory Director
>
> University of Massachusetts Lowell
>
> 1 University Avenue
>
> Lowell MA 01854-5120
>
> 978-934-1809 land
>
> 617-755-0910 mobile
>
> 978-934-1014 fax
>
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20140114/46582208/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 57996 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20140114/46582208/attachment.png>


More information about the labnetwork mailing list