Hi Joseph,
In my experience the pressure of the gaseous N2 entering the building is
of the order of 100 psig.
You can estimate the required flow from the utility matrix prepared by
your design team. If other labs in the building need N2, please add
their estimated usage to the flow.
LN2 tanks come in various volumes. The one I had at UPenn was 6,000
gallons and come with two vaporizers that alternate on their roles of
generating gaseous N2. At UDel I have two 3,000 gallons tanks, each with
two alternating vaporizers. At both UPenn and UDel, the tank(s) feed
both the cleanroom and the research labs in the building.
To avoid major issues with ice on the vaporizers, please try to avoid
having the LN2 tank(s) and vaporizer(s) on the North side of the building.
Cheers,
Iulian
iulian Codreanu, Ph.D.
Director, Nanofabrication Facility
University of Delaware
Harker ISE Lab, Room 163
221 Academy Street
Newark, DE 19716
302-831-2784
http://udnf.udel.edu
On 7/26/2022 3:41 PM, Joseph Losby wrote:
> Hi Everyone, we are in the early stages of building a cleanroom
> facility and looking into nitrogen production as well. I am
> interested in any options. The vendors that I have spoken with
> require N2 purity, flow, volume, and pressure. Any advice on how one
> would roughly calculate the latter three parameters, or have values
> based on usage by your facility? I have been using the latter option
> and scaling our requirements accordingly based on tools we plan on
> acquiring, and roughly what our facility usage would be.
>
> Cheers,
>
> Joseph Losby
> Institute for Quantum Science and Technology
> University of Calgary
> ------------------------------------------------------------------------
> *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Peder
> Lenvik <plenvik at ucsb.edu>
> *Sent:* July 26, 2022 11:02 AM
> *To:* Ferraguto, Thomas S <Thomas_Ferraguto at uml.edu>; Labnetwork
> <labnetwork at mtl.mit.edu>
> *Cc:* thomas.ferraguto at baesystems.com <thomas.ferraguto at baesystems.com>
> *Subject:* Re: [labnetwork] Nitrogen Generation
> [△EXTERNAL]
>
>
>
> We are considering it.
>
> Interested to hear anyone else’s experiences.
>
> *Peder Lenvik*
>
> Sr. Facilities Lead
>
> Electrical and Computer Engineering
>
> E.S.B. Bldg. #225, Room 1109F
>
> University California Santa Barbara, Santa Barbara, Ca. 93106
>
> Cell: (805)698-7461
>
> plenvik at ucsb.edu <mailto:plenvik at ucsb.edu>
>
> http://www.nanotech.ucsb.edu <http://www.nanotech.ucsb.edu/>
>
> *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> *On Behalf Of
> *Ferraguto, Thomas S
> *Sent:* Tuesday, July 26, 2022 5:12 AM
> *To:* Labnetwork (labnetwork at mtl.mit.edu) <labnetwork at mtl.mit.edu>
> *Cc:* thomas.ferraguto at baesystems.com
> *Subject:* [labnetwork] Nitrogen Generation
>
> Colleagues,
>
> Has anyone taken the leap moved to Nitrogen Generation or at least
> supplemented with Nitrogen Generation (for 3 9's house Nitrogen)?
>
> FYI... I'm no longer at UMass but there is nothing better for fast
> feed back than the Labnetwork!
>
> Tom
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220727/8959e4bd/attachment.html>
More information about the labnetwork
mailing list