Hi,

We have deposited A-Si by PECVD with Argon, is OK for 1um but any thicker we started having stress delamination.

Best regards

Peter

From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Carlos Gramajo
Sent: 26 May 2023 17:04
To: Zhu Yunxuan <yxzhuphys245 at gmail.com>
Cc: labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Question regarding amorphous Si deposition

This email was sent to you by someone outside the University.
You should only click on links or attachments if you are certain that the email is genuine and the content is safe.
Hi Zhu,

I have done 1 µm films of amorphous Si in an e-beam evaporation chamber.  I had some issues with delamination when I was growing it. What I had to do was to slow down the rate of deposition. Unfortunately that means you have longer deposition times, therefore more exposure to radiative heat from the source. In that chamber we did not have a thermocouple on the sample holder and we didn't know what was the max temperature reached.
It may be worth trying if you have a way to measure T and see if you can keep that in check by growing it in stages with cooling times in between.

I hope this helps,

Cheers,

Carlos

On Fri, May 26, 2023 at 6:57 AM Zhu Yunxuan <yxzhuphys245 at gmail.com<mailto:yxzhuphys245 at gmail.com>> wrote:
Hi everyone,

Is there a way to deposit thick amorphous Si on several micrometres scale(~5um or above) but with a low processing temperature (<150 deg C)? PECVD is one possibility but the typical operating substrate temperature is too high (>200 deg C) which will destroy our device.

Best regards,
Longji Cui

University of Colorado Boulder
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!iEEe9-9D49oHAlx2fmI8df9fC8KZJ0unnifnDxhvfLmZv-NC19Mba9aHRcQkySO78VTunKanJAnd3vt_IgVk_tWbYPqw$<https://urldefense.com/v3/__https:/mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!iEEe9-9D49oHAlx2fmI8df9fC8KZJ0unnifnDxhvfLmZv-NC19Mba9aHRcQkySO78VTunKanJAnd3vt_IgVk_tWbYPqw$>


--
Carlos Gramajo
Cleanroom Research Scientist
Shared Equipment Authority (SEA), Rice University
Cell: 713-743-8115; Office: 713-348-8243; cg70 at rice.edu<mailto:cg70 at rice.edu>
The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. Is e buidheann carthannais a th’ ann an Oilthigh Dhùn Èideann, clàraichte an Alba, àireamh clàraidh SC005336.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20230529/97d5703f/attachment.html>


More information about the labnetwork mailing list