[labnetwork] [Non-DoD Source] Re: question about Liner for Ni Ebeam Evaporation

Robert Vandusen RobertVandusen at cunet.carleton.ca
Thu Dec 12 21:05:28 EST 2024


Hi all.
We have put Al and Ti directly into the copper hearth with no allowing at all.  We have done this for 30 + years now with an out issue.  The Al usually easily pops out for routine cleaning.  The Ti stays in a ball and also comes out ok.  All other materials use a fabmate liner.  We have the same issues with our Ni.  We use slow ramps but we live with the fact that we go through a liner every few runs.

Rob Vandusen
Carleton University
Ottawa, Canada

Get Outlook for Android<https://aka.ms/AAb9ysg>
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Martin, Michael <michael.martin at louisville.edu>
Sent: Thursday, December 12, 2024 12:41:25 PM
To: EDWARDS, JASON R CTR USAF AFMC AFRL/RYDD <jason.edwards.30.ctr at us.af.mil>; Beaudoin, Mario <beaudoin at physics.ubc.ca>; Graham Gibson <gibsong at queensu.ca>; Owain Clark <odc1n08 at soton.ac.uk>; shokoofe haghighi <shokoofe_haghighi at yahoo.com>; labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] [Non-DoD Source] Re: question about Liner for Ni Ebeam Evaporation

[External Email]
Personally, I'm terrified that without a liner we will punch a hole in the pocket.  This would be catastrophic since then coolant would spray into the chamber just a couple of cm away from both high voltage and high current electrodes.

Jason (and others who go liner-less), wouldn't you be concerned about alloying the target material to the copper liner, I guess not?  We have issues similar to Ni with aluminum.  Almost every liner material seems incompatible with Al in one way or another yet ebeam Al can be very low stress. BTW, I believe lesker suggests to use thermal evaporation for Al.  However, if you just put the Al in the pocket without a liner I would expect a copper-aluminum alloy to form. Didn't Al-ebeam come up not so long ago in this group with several mentioning they put the material directly in the pocket?
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of EDWARDS, JASON R CTR USAF AFMC AFRL/RYDD <jason.edwards.30.ctr at us.af.mil>
Sent: Thursday, December 12, 2024 6:57 AM
To: Beaudoin, Mario <beaudoin at physics.ubc.ca>; Graham Gibson <gibsong at queensu.ca>; Owain Clark <odc1n08 at soton.ac.uk>; shokoofe haghighi <shokoofe_haghighi at yahoo.com>; labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] [Non-DoD Source] Re: question about Liner for Ni Ebeam Evaporation

You don't often get email from jason.edwards.30.ctr at us.af.mil. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>

CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe.

We evaporate straight out of the pocket for most of our metals, it’s interesting to hear that this may not be standard procedure for most labs. Not to draw attention away from the original question in this thread but I would like to hear more about what benefits you are getting from using liners. Is it just for ease of swapping metals in the pockets? In the instance of a Cu liner for Ni are you able to make a proper melt slug with the Ni considering you are unable to get the beam too close to the Cu liner? Thanks in advance.













[cid:image002.png at 01DB4C63.2DC85AA0]

Jason Edwards

KBR  |  Clean Room Facilities Manager , Defense Systems Engineering

2241 Avionics Circle, Area B, Bldg.600|WPAFB, OH, 45433-7322 I

Office: +1 937.713.8741  | Mobile: +1 937.524.7701



         Jason.edwards.30.ctr at us.af.mil<http://Jason.edwards.30.ctr@us.af.mil/>



[cid:image003.png at 01DB4C63.2DC85AA0]<https://www.facebook.com/pages/KBR-Inc/153705171306696>       [cid:image004.png at 01DB4C63.2DC85AA0] <https://twitter.com/KBRincorporated>        [cid:image005.png at 01DB4C63.2DC85AA0] <https://www.linkedin.com/company/kbr-inc>       [cid:image006.png at 01DB4C63.2DC85AA0] <https://instagram.com/kbr.inc>











From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Beaudoin, Mario
Sent: Wednesday, December 11, 2024 5:07 PM
To: Graham Gibson <gibsong at queensu.ca>; Owain Clark <odc1n08 at soton.ac.uk>; shokoofe haghighi <shokoofe_haghighi at yahoo.com>; labnetwork at mtl.mit.edu
Subject: [Non-DoD Source] Re: [labnetwork] question about Liner for Ni Ebeam Evaporation



We use a Cu crucible (4.4cc in a 7cc pocket) with good results.

Mario



On 2024-12-11 9:11 a.m., Graham Gibson wrote:

[CAUTION: Non-UBC Email]

   It’s interesting to see such a wide variety of experiences. I’m too chicken to evaporate right out of the pocket, so we went through a lot of liners, too, before settling on something.



   We have had good luck with Cu liners for Ni e-beam evaporation, but with a 2.2 cc pocket, we have to be very careful to keep the beam away from the walls and limit the deposition rate to 1 Angstrom/s.



  Hope you find something that works, liners get expensive!





Graham Gibson



Operations Manager, NanoFabrication Kingston

Queen’s University

945 Princess Street

Kingston, Ontario, Canada

K7L 0E9

gibsong at queensu.ca<mailto:gibsong at queensu.ca>

www.nanofabkingston.ca<http://www.nanofabkingston.ca/>







From: labnetwork <labnetwork-bounces at mtl.mit.edu><mailto:labnetwork-bounces at mtl.mit.edu> On Behalf Of Owain Clark
Sent: December 5, 2024 3:48 AM
To: Jeff Salzmann <jks7 at buffalo.edu><mailto:jks7 at buffalo.edu>; Bernhard Reineke <bernhard.reineke at uni-paderborn.de><mailto:bernhard.reineke at uni-paderborn.de>; shokoofe haghighi <shokoofe_haghighi at yahoo.com><mailto:shokoofe_haghighi at yahoo.com>; labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] question about Liner for Ni Ebeam Evaporation



Interestingly we have no trouble with Ni, it is one of our most stable materials and the crucible has not been changed in years. I believe it is W.



I suspect reading these comments it is because we have larger 40cc crucibles and we only melt a spot in the center with a small beam deflection. The crucible walls never see melted Ni.



BR, Owain



From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Jeff Salzmann
Sent: 04 December 2024 17:15
To: Bernhard Reineke <bernhard.reineke at uni-paderborn.de<mailto:bernhard.reineke at uni-paderborn.de>>; shokoofe haghighi <shokoofe_haghighi at yahoo.com<mailto:shokoofe_haghighi at yahoo.com>>; labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] question about Liner for Ni Ebeam Evaporation



You don't often get email from jks7 at buffalo.edu<mailto:jks7 at buffalo.edu>. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>

CAUTION: This e-mail originated outside the University of Southampton.

Haghighi,



This is my collection of failed crucibles used to evaporate Ni.



We decided on graphite, because copper was worse. We limit our deposition rate to 0.2 Å/sec to help extend the lifespan of the crucibles. They still fail after 2-3 depositions.



Regards,

Jeff





Jeff Salzmann

Assistant Professor of Research

Cleanroom Manager, Shared Instrumentation Laboratories

School of Engineering and Applied Sciences

University at Buffalo

114A Davis Hall

Buffalo, NY 14260

Tel: 716.645.2584



From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> on behalf of Bernhard Reineke <bernhard.reineke at uni-paderborn.de<mailto:bernhard.reineke at uni-paderborn.de>>
Date: Wednesday, December 4, 2024 at 11:46
To: shokoofe haghighi <shokoofe_haghighi at yahoo.com<mailto:shokoofe_haghighi at yahoo.com>>, labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu> <labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>>
Subject: Re: [labnetwork] question about Liner for Ni Ebeam Evaporation

You don't often get email from bernhard.reineke at uni-paderborn.de<mailto:bernhard.reineke at uni-paderborn.de>. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>

Hi Haghighi,

we tried this once, same result. All liners will crack. Lesker has some recommendations on Ebeam evaporation of Ni: slow power ramp up and down (works well to prevent cracking for ~2-4 Runs, with fabmate). A Cu liner should work better but we never tested that.

https://www.lesker.com/newweb/deposition_materials/depositionmaterials_evaporationmaterials_1.cfm?pgid=ni1

Best,

Bernhard

________________________________

Von: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> im Auftrag von shokoofe haghighi <shokoofe_haghighi at yahoo.com<mailto:shokoofe_haghighi at yahoo.com>>
Gesendet: Mittwoch, 4. Dezember 2024 10:35:19
An: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Betreff: [labnetwork] question about Liner for Ni Ebeam Evaporation



Hi, hope you are having good days,



this is Haghighi, working on metallization by a PVD, using thermal and Ebeam evaporation.



I have some issues choosing the right liner for Ni Ebeam evaporating, hope you find time to guide me through it.

We have tested graphite, glassy carbon coated graphite, Al2O3, Mo, W and Cu liners so far.

    Graphite liner just broke after first run from the border line of Ni surface. Glassy carbon coated graphite was more or less the same,

    Al2O3 endured for 4-5 runs before breaking but there were cracks on it,

    Mo liner totally melted and some kind of alloy was produced,

    W and Cu endured but their internal walls were some how melted and we can't be sure about pureness of evaporated Ni and final produced thin films.



do you have any other suggestions?





high voltage: 6.1 KV

emission current: 3 - 5 A





Thanks for your time and attention in advance, I hope you kindly find time and consider replying me,



Dr. Haghighi





_______________________________________________

labnetwork mailing list

labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>

https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork

--
[cid:image001.jpg at 01DB4C61.3F2CF050]

This email contains links to content or websites. Always be cautious when opening external links or attachments.

Please visit https://carleton.ca/its/help-centre/report-phishing/ for information on reporting phishing messages.

When in doubt, the ITS Service Desk can provide assistance. https://carleton.ca/its/chat

-----End of Disclaimer-----

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.jpg
Type: image/jpeg
Size: 22413 bytes
Desc: image001.jpg
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 5577 bytes
Desc: image002.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.png
Type: image/png
Size: 584 bytes
Desc: image003.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image004.png
Type: image/png
Size: 660 bytes
Desc: image004.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment-0002.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image005.png
Type: image/png
Size: 715 bytes
Desc: image005.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment-0003.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image006.png
Type: image/png
Size: 901 bytes
Desc: image006.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241213/609a888a/attachment-0004.png>


More information about the labnetwork mailing list