We found the same thing in a very similar setup and identical ES&H guidance. Typically, we see 70-75 dB in the absence of personnel. The chase is closer to 75 dBa, and the bays are closer to 70 dBa. There were a few addressable offenders, like a loud spot where disrupted airflow created significant turbulence, increasing sound levels to 75 dBa in the cleanroom, but this was easily resolved by removing the obstruction. Because none of these exceed the 85 dBA for 8h limit, ES&H will not get involved.
Users concerned about sound levels were tentative about hearing protection. What they actually wanted in raising the concern was (a) to address the handful of loud spots, and (b) approval to use their noise cancelling earbuds in the cleanroom. To the latter, we’re in a trial period – if it doesn’t impact our ability to get someone’s attention and talk with them when needed (e.g., can switch into transparency mode for a conversation), then neither I nor ES&H mind terribly if they wear their earbuds. Quality of life counts for something. We only cautioned that if they play music, they set volume limits to not damage their hearing.
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of N Shane Patrick
Sent: Tuesday, March 18, 2025 5:41 PM
To: Joseph Losby <joseph.losby at ucalgary.ca>
Cc: labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Cleanroom dB levels
Our facility is bay and chase, no subfab, so all the pumps and chillers and all are right next door to the equipment. After a recent request by one of our staff members, I did an informal survey using NIOSH’s Sound Level Monitor app (NIOSH SLM) and an uncalibrated microphone. The only time I got above 80db was when people were talking or dropped things, even in our loudest chases.
Sustained levels at 85 or more are the levels of concern for hearing loss according to US agencies. Our EHS team agreed with me that even with an uncalibrated system we were very unlikely to reach sustained levels of concern. People are still welcome to use hearing protection for their own comfort, of course, and I’d likely provide it for my staff if asked, but it takes a lot to get to that sustained 85db point.
Now, 80db acoustics and the associated mechanical vibrations from all of that is certainly enough to mess with sensitive instruments, and we have had to relocate certain tools or just accept our environmental noise translates to measurement noise. That’s more a matter of design and space limitations for us than anything else.
-N. Shane Patrick
Manager, Lab Operations and Safety
Washington Nanofabrication Facility
Sent from my iPhone
On Mar 18, 2025, at 5:26 PM, Joseph Losby <joseph.losby at ucalgary.ca<mailto:joseph.losby at ucalgary.ca>> wrote:
Hi everyone,
Have any of you taken acoustic noise measurements of your cleanroom facilities, or have a basis of design for acceptable levels?
If you can share your facility ACH with noise level ( dB ) that would be great. Any other input will be much appreciated as well.
Cheers,
Joe
Joseph Losby, PhD
Manager, qLab Operations
joseph.losby at ucalgary.ca<mailto:joseph.losby at ucalgary.ca>
<https://urldefense.us/v3/__https:/urldefense.com/v3/__https:/*io.ucalgary.ca/quantum-city__;!!K-Hz7m0Vt54!iHLndPt-5Otg8wVsZQXZxk-3iaKCEbMpPvt5GPN-MChuQyulVrIaqzrRJ0q_RieM8khiBQeSZ9L7gXCkYXcid4wIyg$__;Lw!!G2kpM7uM-TzIFchu!waiizu6-NrHKSNYCAeodwwzn2tkvyywQPadBM_6aIAQ07Ac016jlYSSD1IujWBz1mMgfFKOzu6kWGk4sg1A$>
<Outlook-3jbzrxwq.jpg><https://urldefense.us/v3/__https:/urldefense.com/v3/__https:/*io.ucalgary.ca/quantum-city__;!!K-Hz7m0Vt54!iHLndPt-5Otg8wVsZQXZxk-3iaKCEbMpPvt5GPN-MChuQyulVrIaqzrRJ0q_RieM8khiBQeSZ9L7gXCkYXcid4wIyg$__;Lw!!G2kpM7uM-TzIFchu!waiizu6-NrHKSNYCAeodwwzn2tkvyywQPadBM_6aIAQ07Ac016jlYSSD1IujWBz1mMgfFKOzu6kWGk4sg1A$>
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!K-Hz7m0Vt54!iHLndPt-5Otg8wVsZQXZxk-3iaKCEbMpPvt5GPN-MChuQyulVrIaqzrRJ0q_RieM8khiBQeSZ9L7gXCkYXfllrKreA$<https://urldefense.us/v3/__https:/urldefense.com/v3/__https:/*mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!K-Hz7m0Vt54!iHLndPt-5Otg8wVsZQXZxk-3iaKCEbMpPvt5GPN-MChuQyulVrIaqzrRJ0q_RieM8khiBQeSZ9L7gXCkYXfllrKreA$__;Lw!!G2kpM7uM-TzIFchu!waiizu6-NrHKSNYCAeodwwzn2tkvyywQPadBM_6aIAQ07Ac016jlYSSD1IujWBz1mMgfFKOzu6kWLGLr4ac$>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250319/95b31592/attachment.html>
More information about the labnetwork
mailing list