[labnetwork] Dry Plasma Etch Cleaning Recipes and Evaluation
Nadia Pervez
np65 at rice.edu
Thu Mar 20 10:14:56 EDT 2025
Hi Emma,
Most of our users run very short processes and Ar/O2 clean before
conditioning. We're in the unfortunate situation of having both F and Cl in
the same tool.
This our mandatory clean after BCl3 processes (after discussion with Craig
Ward at Oxford):
- step 1: Cl2 clean
- Cl2: 50 sccm
- Pressure: 20 mTorr
- ICP: 1500 W
- RF: 150 W
- Temperature: 20C
- Backside He pressure: 5 Torr
- Time: 20 min (or the length of the BCl3 process, whichever is
longer)
- step 2: O2/SF6 clean
- O2: 40 sccm
- SF6: 10 sccm
- Pressure: 20 mTorr
- ICP: 1500 W
- RF: 150 W
- Temperature: 20C
- Backside He pressure: 5 Torr
- Time: 20 min (or the length of the BCl3 process, whichever is
longer)
The intention of step 1 is to remove BCl3, and step 2 is to remove Cl2.
Thanks,
Nadia
On Mon, Mar 17, 2025 at 8:36 PM Emma Anquillare <eanquillare at gc.cuny.edu>
wrote:
> Dear Labnetwork Hivemind,
>
> I am looking to assemble what we know collectively about dry plasma etch
> cleaning recipes so that I can summarize, compare, and ultimately
> experimentally evaluate them.
>
> I am interested in:
> -Do your facility's etchers (ICP/ RIE) have custom clean recipes specific
> to certain etch recipes or materials? If so, what are they and what are
> both the recipe and clean parameters?
> -Do you use an emission spectrometer (or any other method) to develop or
> evaluate these cleans?
> -How did you come up with these cleans in the first place?
> -Have you seen any good talks, literature, or other resources relevant to
> this subject?
>
>
> Anything you can share is appreciated and (I hope) will be worth your
> efforts to provide, as I am seeking to compile this for public benefit.
>
> Thanks!
> Emma
>
> *_____________________________________________*
> *Emma Anquillare, PhD*
> Research Scientist
> ASRC Nanofabrication Facility
> City University of New York
>
> *Catalyzing Change, Celebrating Gains: *
> *A decade of visionary science for the public good.*
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
--
Nadia Pervez, PhD
Cleanroom Research Scientist
Shared Equipment Authority (SEA)
Rice University
6100 Main St, MS-686
Houston, TX 77005
np65 at rice.edu
Office: SST-017, 713-348-5069
Cell: 805-284-1193
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250320/a5fa3ace/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Outlook-mzn1fp0b.jpg
Type: image/jpeg
Size: 141606 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250320/a5fa3ace/attachment.jpg>
More information about the labnetwork
mailing list