[labnetwork] [EXTERNAL] Re: Issue Striking Plasma for NiFe Sputtering on Angstrom Nexdep PVD System – Need Guidance

Alex Dixon agdixon at mines.edu
Tue Oct 14 19:30:49 EDT 2025


For magnetic materials the standard recommendation is 0.125 thick target with a spacer ring, and mount directly to the gun without a backing plate or Cu mesh, however I have been able to sputter 1/4" thick magnetic targets as long as they are directly mounted.

Also, did you remove the center magnet stack from the gun and replace with a Fe slug prior to loading your target?

-Alex

__________________________________________________________


Alex Dixon, Ph.D. (He/Him)
Lab Manager
Nanofabrication and Thin Film Deposition
Shared Instrument Facility
Colorado School of Mines

________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Jing Guo <jg78 at rice.edu>
Sent: Monday, October 13, 2025 9:18 PM
To: Mohammad Mohtasim Hamid Pial <mpial001 at fiu.edu>
Cc: labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: [EXTERNAL] Re: [labnetwork] Issue Striking Plasma for NiFe Sputtering on Angstrom Nexdep PVD System – Need Guidance

CAUTION: This email originated from outside of the Colorado School of Mines organization. Do not click on links or open attachments unless you recognize the sender and know the content is safe.

Hi Pial,

Usually for magnetic target, it cannot be a thick one (typically 1-3mm). Otherwise the magnetic field is shunting which will result in plasma strike failure. When I install the magnetic target on AJA tool, there is no bonding Cu backing plate, the Fe or Ni target will be installed directly onto the sputter gun. I think your target thickness is fine, but the backing plate is a problem. You should check with Angstrom sputter engineers about their system setup.
 Another concern is the striking pressure, usually I use a high pressure to strike (20-30mTorr), then reduce the pressure to 3-5mTorr for deposition.

I feel the bonded target thickness is the biggest problem.



Best,
Jing

---------------------------------------
Jing Guo Ph.D.
Cleanroom Manager/Research Scientist
SEA Cleanroom (SST 017)
Rice University
Houston, TX
jeanne.guo at rice.edu
713-348-8227

On Oct 13, 2025, at 3:32 PM, Mohammad Mohtasim Hamid Pial <mpial001 at fiu.edu> wrote:

Dear All,

I am attempting to deposit NiFe using a magnetron sputtering setup (Angstrom Engineering – Nexdep PVD platform). Below are my target specifications and sputtering parameters:

Target: NiFe (2" dia × 1 mm, 4N purity) bonded to a 1/8" Cu backing plate
Sputtering Method: DC Magnetron Sputtering
DC Power: 100–150 W
Base Pressure: ~10⁻⁶ Torr
Ar Gas Flow Rate: 20–30 sccm
Working Pressure: 3–5 mTorr

However, I am unable to ignite the plasma. The DC power supply setpoint keeps blinking and the power stalls around 15 W before dropping

Could this issue be due to insufficient magnetic field strength (not penetrating the thick target/Cu backing), or is the target too thick for effective plasma ignition? Do I need a thinner target or additional magnetic confinement?

I would greatly appreciate any insights or recommendations from anyone who has sputtered NiFe or similar high-permeability alloys on this platform.

Thank you in advance!
Pial


_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20251014/111f289a/attachment.html>


More information about the labnetwork mailing list