[labnetwork] [EXTERNAL] Re: Automatic gas shutoff at the source
Ahad Syed
ahad.syed at kaust.edu.sa
Thu Apr 23 04:40:12 EDT 2026
Hi Kostis,
The response is same for KAUST-NCL. The fire alarm will trigger the TGM and that will turn off the valves, gas cabinets and VMBs. And vice-versa as well. Any TGM alarm from the NCL will trigger Building evacuation through the Fire System.
All the best,
Ahad
--
Nanofabrication Core Lab
4700 King Abdullah University of Science and Technology
0253 Building 3, Thuwal 23955-6900, Kingdom of Saudi Arabia
Tel:+966 12 808 2348
Visit The Core Labs<https://corelabs.kaust.edu.sa/visitor-information>
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Iulian Codreanu <codreanu at udel.edu>
Sent: Wednesday, April 22, 2026 2:51 PM
To: Michelakis, Kostis <kostis at slac.stanford.edu>
Cc: labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: [EXTERNAL] Re: [labnetwork] Automatic gas shutoff at the source
Hi Kostis,
Here at UDel it has been mandated that a fire alarm input to the TGMS closes all gas valves.
iulian Codreanu, Ph.D.
Director, Nanofabrication Facility
University of Delaware
Harker ISE Lab, Room 163
221 Academy Street
Newark, DE 19716
302-831-2784
https://udnf.udel.edu<https://urldefense.com/v3/__https://udnf.udel.edu__;!!Nmw4Hv0!y-HyrJbHKH6wuGBDErykMDLIVRAlsNGSR-6J0ORaDPxIjd2N9L6eqdNFho961l2du0Fg8MTplJRQWPVa9nhQySo$>
On 4/21/2026 3:37 PM, Michelakis, Kostis wrote:
Dear Labnetwork colleagues,
We are reviewing our emergency procedures and I would like to ask, specifically for your hazardous and toxic gases, in the case of a fire alarm being triggered, if you make use of the building's Fire Alarm Control Panel (FACP) respective signal to automatically shutoff gases at the source.
We have at our disposal a range of engineering controls to shutoff gases at the source, but I am specifically interested to learn if the method of using directly the fire alarm signal has been mandated or otherwise chosen in your setting.
Thank you in advance for your consideration,
Kostis
--
Kostis Michelakis, PhD
DMF Lab Manager
kostis at slac.stanford.edu<mailto:kostis at slac.stanford.edu>
SLAC National Accelerator Laboratory
slac.stanford.edu<https://urldefense.com/v3/__http://slac.stanford.edu__;!!Nmw4Hv0!y-HyrJbHKH6wuGBDErykMDLIVRAlsNGSR-6J0ORaDPxIjd2N9L6eqdNFho961l2du0Fg8MTplJRQWPVaJxoraLk$>
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>