We had the same issue on our board, but I was able to replace the transistor and get our PECVD up and running.
Cheers!
Julia Aebersold, Ph.D.
Manager, Micro/Nano Technology Center
University of Louisville
Shumaker Research Building, Room 233
2210 South Brook Street
Louisville, KY 40292
(502) 852-1572
http://louisville.edu/micronano/
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Football
Sent: Thursday, June 23, 2022 11:13 AM
To: labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Oxford Wafer Jam/Break
CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe.
Hi Long,
We had a very similar issue on an Oxford PECVD tool some time back. And it was a ZTX 450 transistor on a board under the loadlock chamber that was burnt out. And as others have mentioned, it can be identified by the discolored place on the board. The service engineer from Oxford replaced the board though.
[cid:image001.png at 01D887BA.F7463570]
Fubo Rao, Ph.D.,
Nanofabrication Cleanroom Manager,
Center for Nanoscale Materials,
Argonne National Laboratory
9700 S. Cass Ave, Lemont, IL 60439
Phone: 630-252-5708
Email: frao at anl.gov<https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Ffrao%40anl.gov%2F&data=05%7C01%7Cjulia.aebersold%40louisville.EDU%7Cf866969679974a32d66908da555a0cf4%7Cdd246e4a54344e158ae391ad9797b209%7C0%7C0%7C637916144256104262%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=9C0Pkd3ltGSxQqbPOBfwalXl5XofWUtrAi1BGkmaBWM%3D&reserved=0>
On Thu, Jun 23, 2022 at 8:50 AM Michael Perry <M.G.Perry at soton.ac.uk<mailto:M.G.Perry at soton.ac.uk>> wrote:
Hi Long,
Justin is likely to be right, as steve says, he knows his stuff. If its been failing for some time though it can also start to burn out one of the other transistors. Any that are discolored replace. Its sometimes easier to just snip the legs off and solder a new one to the legs to minimise risk of damage to pcb. Or if you want to future proof it, stick some sockets in there. Really common problem on oipt LL's. keep a batch on 450's and 550's as spares. They can fail as well.
mike
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> on behalf of Paolini, Steven <spaolini at cns.fas.harvard.edu<mailto:spaolini at cns.fas.harvard.edu>>
Sent: 22 June 2022 17:06
To: Moreau, Justin <Justin_Moreau at uml.edu<mailto:Justin_Moreau at uml.edu>>; Chang, Long <lvchang at Central.UH.EDU<mailto:lvchang at Central.UH.EDU>>; labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu> <labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>>
Subject: Re: [labnetwork] Oxford Wafer Jam/Break
CAUTION: This e-mail originated outside the University of Southampton.
Long,
I misinterpreted your message yesterday. It would be wise to take Justin's advice, he is the top dog when it comes to Oxford equipment.
Steve
Steve Paolini
Principal Equipment Engineer
Harvard University Center for Nanoscale Systems
11 Oxford St.
Cambridge, MA 02138
617- 496- 9816
spaolini at cns.fas.harvard.edu<mailto:spaolini at cns.fas.harvard.edu>
www.cns.fas.harvard.edu<https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.cns.fas.harvard.edu%2F&data=05%7C01%7Cjulia.aebersold%40louisville.EDU%7Cf866969679974a32d66908da555a0cf4%7Cdd246e4a54344e158ae391ad9797b209%7C0%7C0%7C637916144256104262%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=Rx2fLmsIEQ%2ByWLthUZ%2FX8oVE%2BzyPPs9zZQSiPF2JEKQ%3D&reserved=0>
From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Moreau, Justin
Sent: Wednesday, June 22, 2022 7:59 AM
To: Chang, Long <lvchang at Central.UH.EDU<mailto:lvchang at Central.UH.EDU>>; labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] Oxford Wafer Jam/Break
Hi Long
By jamming does the handler just stall?
I don't think adding more time or removing PR from your wafer edge will help.
I suspect one of the handling sensors isn't made, you have a mechanical bind (ff feedthrough, gearbox, linear bearing) or you have a failing drive transistor.
Looking at your image- Slit valve is open, clamp is up, lift is up, arm is extended
Is your "arm out"sensor made? If no you should adjust the arm out sensor.
If the arm out sensor is made. Check for a mechanical bind- Remove the mounting plate that holds the motor and gearbox assy. You can then try and rotate the ferrofluid feedthrough shaft by hand.
If you can rotate, you don't have a bind.
This looks like a Blue PLC control system so it would likely be the Q4 transistor on the LL control card that's the problem. Its mounted just below the LL chamber. It can overheat and intermittently work. Q4 controls the drive back into the LL.
It will also discolor so it should be easy to spot. You can replace it with a ZTX450 NPN.
Hope this helps
[University of Massachusetts Lowell]
Justin Moreau
Sr. Lab Manager, Nanofabrication Core
CORE RESEARCH FACILITIES
ETIC
E: Justin_Moreau at uml.edu<mailto:Justin_Moreau at uml.edu>
T: 978-934-3615
From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Chang, Long
Sent: Tuesday, June 21, 2022 2:51 PM
To: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: [labnetwork] Oxford Wafer Jam/Break
This e-mail originated from outside the UMass Lowell network.
________________________________
Hi Guys,
After etching through 380um of Silicon, the loading arm would get jammed during unloading, see photo. The problem was fixed by adding a 15min step to do nothing step to allow the clamp to cool down before unloading. Now the problem is back and the plan is to increase the cool down time until it stops jamming. Is there a better way to avoid this problem? Should the PR on the wafer edge be removed?
Thanks,
Long
[cid:image003.jpg at 01D887BA.F7463570]
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork<https://nam11.safelinks.protection.outlook.com/?url=https%3A%2F%2Fmtl.mit.edu%2Fmailman%2Flistinfo.cgi%2Flabnetwork&data=05%7C01%7Cjulia.aebersold%40louisville.EDU%7Cf866969679974a32d66908da555a0cf4%7Cdd246e4a54344e158ae391ad9797b209%7C0%7C0%7C637916144256104262%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=S7gL6Oy3rjou%2BRVAyHDbDZPfworZKr2oT6DpvNEOhEs%3D&reserved=0>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220624/ffb1f891/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 451970 bytes
Desc: image001.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220624/ffb1f891/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 2650 bytes
Desc: image002.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220624/ffb1f891/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.jpg
Type: image/jpeg
Size: 300976 bytes
Desc: image003.jpg
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220624/ffb1f891/attachment.jpg>
More information about the labnetwork
mailing list