Haghighi,

This is my collection of failed crucibles used to evaporate Ni.

We decided on graphite, because copper was worse. We limit our deposition rate to 0.2 Å/sec to help extend the lifespan of the crucibles. They still fail after 2-3 depositions.

Regards,
Jeff


Jeff Salzmann
Assistant Professor of Research
Cleanroom Manager, Shared Instrumentation Laboratories
School of Engineering and Applied Sciences
University at Buffalo
114A Davis Hall
Buffalo, NY 14260
Tel: 716.645.2584

From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Bernhard Reineke <bernhard.reineke at uni-paderborn.de>
Date: Wednesday, December 4, 2024 at 11:46
To: shokoofe haghighi <shokoofe_haghighi at yahoo.com>, labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] question about Liner for Ni Ebeam Evaporation
You don't often get email from bernhard.reineke at uni-paderborn.de. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>
Hi Haghighi,

we tried this once, same result. All liners will crack. Lesker has some recommendations on Ebeam evaporation of Ni: slow power ramp up and down (works well to prevent cracking for ~2-4 Runs, with fabmate). A Cu liner should work better but we never tested that.

https://www.lesker.com/newweb/deposition_materials/depositionmaterials_evaporationmaterials_1.cfm?pgid=ni1

Best,

Bernhard
________________________________
Von: labnetwork <labnetwork-bounces at mtl.mit.edu> im Auftrag von shokoofe haghighi <shokoofe_haghighi at yahoo.com>
Gesendet: Mittwoch, 4. Dezember 2024 10:35:19
An: labnetwork at mtl.mit.edu
Betreff: [labnetwork] question about Liner for Ni Ebeam Evaporation

Hi, hope you are having good days,

this is Haghighi, working on metallization by a PVD, using thermal and Ebeam evaporation.

I have some issues choosing the right liner for Ni Ebeam evaporating, hope you find time to guide me through it.
We have tested graphite, glassy carbon coated graphite, Al2O3, Mo, W and Cu liners so far.
    Graphite liner just broke after first run from the border line of Ni surface. Glassy carbon coated graphite was more or less the same,
    Al2O3 endured for 4-5 runs before breaking but there were cracks on it,
    Mo liner totally melted and some kind of alloy was produced,
    W and Cu endured but their internal walls were some how melted and we can't be sure about pureness of evaporated Ni and final produced thin films.

do you have any other suggestions?


high voltage: 6.1 KV
emission current: 3 - 5 A


Thanks for your time and attention in advance, I hope you kindly find time and consider replying me,

Dr. Haghighi

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241204/de8c6d0e/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Ni Graveyard.png
Type: image/png
Size: 2682529 bytes
Desc: Ni Graveyard.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20241204/de8c6d0e/attachment.png>


More information about the labnetwork mailing list