[labnetwork] question about Liner for Ni Ebeam Evaporation

Demis D. John demis at ucsb.edu
Wed Dec 4 20:39:06 EST 2024


I found out that we actually evap in the copper Hearth for at least one of
our systems - as others mentioned, just melting directly with no liner.

-- Demis (contact info <https://wiki.nanotech.ucsb.edu/wiki/Demis_D._John>)
*Reminder*: The NanoFab has a publications policy
<https://wiki.nanotech.ucsb.edu/wiki/Frequently_Asked_Questions#Publications_acknowledging_the_Nanofab>


On Wed, Dec 4, 2024 at 16:26 Jeff Salzmann <jks7 at buffalo.edu> wrote:

> Haghighi,
>
>
>
> This is my collection of failed crucibles used to evaporate Ni.
>
>
>
> We decided on graphite, because copper was worse. We limit our deposition
> rate to 0.2 Å/sec to help extend the lifespan of the crucibles. They
> still fail after 2-3 depositions.
>
>
>
> Regards,
>
> Jeff
>
>
>
>
>
> Jeff Salzmann
>
> Assistant Professor of Research
>
> Cleanroom Manager, Shared Instrumentation Laboratories
>
> School of Engineering and Applied Sciences
>
> University at Buffalo
>
> 114A Davis Hall
>
> Buffalo, NY 14260
>
> Tel: 716.645.2584
>
>
>
> *From: *labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Bernhard
> Reineke <bernhard.reineke at uni-paderborn.de>
> *Date: *Wednesday, December 4, 2024 at 11:46
> *To: *shokoofe haghighi <shokoofe_haghighi at yahoo.com>,
> labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
> *Subject: *Re: [labnetwork] question about Liner for Ni Ebeam Evaporation
>
> You don't often get email from bernhard.reineke at uni-paderborn.de. Learn
> why this is important <https://aka.ms/LearnAboutSenderIdentification>
>
> Hi Haghighi,
>
> we tried this once, same result. All liners will crack. Lesker has some
> recommendations on Ebeam evaporation of Ni: slow power ramp up and down
> (works well to prevent cracking for ~2-4 Runs, with fabmate). A Cu liner
> should work better but we never tested that.
>
>
> https://www.lesker.com/newweb/deposition_materials/depositionmaterials_evaporationmaterials_1.cfm?pgid=ni1
>
> Best,
>
> Bernhard
> ------------------------------
>
> *Von:* labnetwork <labnetwork-bounces at mtl.mit.edu> im Auftrag von
> shokoofe haghighi <shokoofe_haghighi at yahoo.com>
> *Gesendet:* Mittwoch, 4. Dezember 2024 10:35:19
> *An:* labnetwork at mtl.mit.edu
> *Betreff:* [labnetwork] question about Liner for Ni Ebeam Evaporation
>
>
>
> Hi, hope you are having good days,
>
>
>
> this is Haghighi, working on metallization by a PVD, using thermal and
> Ebeam evaporation.
>
>
>
> I have some issues choosing the right liner for Ni Ebeam evaporating, hope
> you find time to guide me through it.
>
> We have tested graphite, glassy carbon coated graphite, Al2O3, Mo, W and
> Cu liners so far.
>
>     Graphite liner just broke after first run from the border line of Ni
> surface. Glassy carbon coated graphite was more or less the same,
>
>     Al2O3 endured for 4-5 runs before breaking but there were cracks on
> it,
>
>     Mo liner totally melted and some kind of alloy was produced,
>
>     W and Cu endured but their internal walls were some how melted and we
> can't be sure about pureness of evaporated Ni and final produced thin films.
>
>
>
> do you have any other suggestions?
>
>
>
>
>
> high voltage: 6.1 KV
>
> emission current: 3 - 5 A
>
>
>
>
>
> Thanks for your time and attention in advance, I hope you kindly find time
> and consider replying me,
>
>
>
> Dr. Haghighi
>
>
> _______________________________________________