[labnetwork] PECVD dilute gas question(s)
Nicholas Menounos
menounos at mit.edu
Wed Apr 9 17:01:53 EDT 2025
Hi Nathan,
Reduced code considerations can be significant benefit.
If you can get <1.37% SiH4 the gas is not governed by CGA G-13.
If you can get below the flammability mixture ranges in CGA P-23 you are not flammable.
Assuming you had good results with 2%, you might find a better use for the process gas hardware and associated volume (i.e. MAQ) with an additional hazardous gas.
I can't speak to the process benefits or research results.
Best,
Nicholas Menounos, PE, LEED AP
Associate Director of Infrastructure
MIT.nano
Massachusetts Institute of Technology
77 Massachusetts Ave, Bldg 12-5007
Cambridge, MA 02139
Cell: (508) 932-0938<tel:+15089320938>
Office: (617) 253-7234<tel:+16172537234>
Email: Menounos at mit.edu<mailto:Menounos at mit.edu>
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Martin, Michael
Sent: Tuesday, April 8, 2025 12:52 PM
To: Nathan Aultman <naultman at creol.ucf.edu>; labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] PECVD dilute gas question(s)
Hi Nathan,
We are using 5% SiH4 in Argon. We switched to this from 100% SiH4 some years ago after we had a fire while swapping out a tank. Our stress and other data for nitrides can be found here https://louisville.edu/micronano/files/documents/characterization-data/PECVDSiliconNitrideStressControlandRateData.pdf
In essence, it works for us.
-Michael
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> on behalf of Nathan Aultman <naultman at creol.ucf.edu<mailto:naultman at creol.ucf.edu>>
Sent: Monday, April 7, 2025 9:49 AM
To: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu> <labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>>
Subject: [labnetwork] PECVD dilute gas question(s)
CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe.
Hello Labnetwork,
We have a PECVD system using dilute silane (2%)/N2. Some labs with similar systems use dilute silane (5%)/helium. We are nearing the end of our current cylinder and will need a replacement soon. What are the advantages and disadvantages when choosing the dilute gas? Does choosing a helium mixture produce more desirable films and/or quality? In other words, is it worth it to purchase a cylinder with helium? Also, what would be the reason(s) for using helium as an additional process gas? E.g., a separate MFC and gas line.
Thank you all in advance.
Nathan C. Aultman
University of Central Florida
407-823-6852
naultman at creol.ucf.edu<mailto:naultman at creol.ucf.edu>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20250409/9240d4f6/attachment.html>
More information about the labnetwork
mailing list