[labnetwork] Sputter conditions for XPS samples?
Oncel, Nuri
nuri.oncel at und.edu
Wed May 27 18:04:04 EDT 2026
Hi, Mario,
Ours is a fairly old system as well (PHI 5400). You can control the sputtering rate by adjusting the background pressure, beam voltage, and ion beam current.
We typically backfill the chamber with argon to around (10^{-7}) mbar. The beam voltage can range from 1 kV to 5 kV.
RBD Instruments used to provide free manuals and “how-to” documents, which may still be useful resources.
If you have any questions, feel free to send me a private email. We can also discuss it over Zoom, if you prefer.
Best
Nuri
Professor Nuri Oncel, Ph.D.
Director of UND-NanoFoundry<https://www.und.edu/Nanofoundry>
Department of Physics and Astrophysics
University of North Dakota
https://sites.google.com/view/oncel-research-group/group
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Mark Hofheins <mhofheins at unm.edu>
Sent: Wednesday, May 27, 2026 8:20 AM
To: Beaudoin, Mario <beaudoin at physics.ubc.ca>; 'labnetwork at mtl.mit.edu' <labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] Sputter conditions for XPS samples?
Good Morning Mario,
We will flow 13sccms to 20sccms of Ar during our sputter.
Optimum vacuum pressure for us is 3 millitorr.
Power will depend on what type of material the target is made of and
How much deposition you require/per time.
Best regards,
Mark Hofheins
mhofheins at unm.edu<mailto:mhofheins at unm.edu>
Cell 505-259-9278
Office 505-272-7506
Micro Electronics Technician &
Safety Coordinator MTTC
Manufacturing Engineering/
School of Engineering
Science and Technology park
University of New Mexico
800 Bradbury S.E. Suit 169
Albuquerque, New Mexico
87106-4346
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Beaudoin, Mario <beaudoin at physics.ubc.ca>
Sent: Tuesday, May 26, 2026 5:57 PM
To: 'labnetwork at mtl.mit.edu' <labnetwork at mtl.mit.edu>
Subject: [labnetwork] Sputter conditions for XPS samples?
[EXTERNAL]
Dear Network,
We have inherited an old XPS machine. The XPS part works fine and we know how to acquire data. For analysis, we will certainly consult experts before publishing as we are aware of the potential pitfalls.
We also have an Argon source and sputter power supply and we would like to sputter our sample surfaces to remove surface contaminants. The problem: we are neophytes at that part of the process besides operating the cylinder and leak valve.
My question: What are the typical Ar pressures and power settings generally used for sputtering?
Please respond privately or through the network.
Big THANK YOU.
Mario
--
[cid:f51feeb8-c7b8-440f-ae57-8b2c7d8947fe]
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20260527/782d02ff/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Outlook-crgezsts.jpg
Type: image/jpeg
Size: 22413 bytes
Desc: Outlook-crgezsts.jpg
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20260527/782d02ff/attachment.jpg>
More information about the labnetwork
mailing list